Products > Test Equipment
Possible GW Instek GDS-1000B hack
TK:
I tested the BW after applying the 300MHz hack.
Test setup:
Signal Generator (Marconi 2024) --> SMA cable --> SMA_BNC --> 50ohm passthrough --> Scope input Channel 1 (1X attenuation)
Measured Frequency and Vpp
10MHz 1.2Vpp
100MHz 900mVpp (-3dB)
200MHz 280mVpp
250MHz 110mVpp
300MHz 48mVpp
350MHz 20mVpp (5mV vertical division setting)
Cannot measure reliable frequency at 400MHz and up. I think officially it is a 100MHz front end.
Mortymore:
@TK
What are the bandwidth limit options with the 300MHz BW hack?
For the 200MHz, options are 20,100,Full.
Note that GW-Instek has the 1000B with 50/70/100MHz bandwidth selections.
TK:
--- Quote from: Mortymore on April 30, 2019, 03:41:45 pm ---@TK
What are the bandwidth limit options with the 300MHz BW hack?
For the 200MHz, options are 20,100,Full.
Note that GW-Instek has the 1000B with 50/70/100MHz bandwidth selections.
--- End quote ---
For 300MHz, BW limitation options are 20, 100, 200, Full.
TK:
I am not liking the refresh rate of the GDS1054B. I am feeding an AM modulated signal and on both the Keysight 1000X and Rigol DS1054Z I can see a nice AM modulated waveform but on the GDS1054B I see single traces or at most 3-4 traces, but not the nice area filled trace like the other scopes.
The scope UI is very responsive like the Keysight, but I suspect the refresh rate is not near 50,000 wfm/s
nctnico:
The update rate depends on the memory depth setting. But don't expect a very graded signal. The GW Instek goes from like 50% to 100% in order not to hide any rare signal occurence.