Author Topic: Reverse engineering Tektronix TDS3GV module for TDS3000 series oscilloscopes?  (Read 28677 times)
Offline JacquesBBB
Thanks a lot, Sicco
I use normally 3.3V TTL . This is  the setting I put on my TTL to usb  adapter (foca v2.2)https://wiki.iteadstudio.com/Foca
I have now verified that  the data I type is properly received by the TDS by looking to the debug zone
If I type : *IDN?
I see in the 30 last car
*IDN?\015
so this is correct as I configured with  CR (\015)
ON the opposite, No communication Module is detected, but  reading the past posts  from pmercier
Little update. Received the connnectors today. They're a perfect match.
I soldered dupond wires for the 3.3v serial/usb module directly on the provided pins and ... worked like a charm once i removed the hardware control flow from the scope config.
Sended : *IDN?
Got : TEKTRONIX,TDS 3032C,0,CF:91.1CT FV:v4.05 TDS3FFT:v1.00 TDS3TRG:v1.00
I  understand that the 100 pins connectors + TTL to USB  was sufficient.
Offline sicco
Some have tried pulling low the relevant databus lines through eight 1k resistors. Not the best of all ways to do it, but that will make it detect a Communications Card also.
Offline JacquesBBB
Succeed !
I have  carefully followed the instructions of pmercier
https://hackaday.io/project/172242-extension-card-for-tds3000-scopes/log/179520-the-actual-informations
and build the mock communication card based on a SN74HC245N (which I had in stock in PDIP form).
Communication with the TDS3032 was then immediate
*IDN?
TEKTRONIX,TDS 3032,0,CF:91.1CT FV:v3.39 TDS3GM:v1.00 TDS3ENG:v1.00 TDS3FFT:v1.00 TDS3TRG:v1.00
PASSWORD PITBULL
MCONFIG TDS3052
*IDN?
TEKTRONIX,TDS 3052,0,CF:91.1CT FV:v3.39 TDS3GM:v1.00 TDS3ENG:v1.00 TDS3FFT:v1.00 TDS3TRG:v1.00
and it seems the scope was converted to TDS3052
Thanks to all  who helped by direct interaction or by putting usable information online.
« Last Edit: July 29, 2022, 05:50:08 pm by JacquesBBB »
Offline JacquesBBB
A cleaner realisation of  the mock  TDS3GM,
the strange list of communication modules with both  "None" and "TDS3GM"
Isnt that contradictory ?
The rise time after the upgrade using Leo Bodnar  device.
The following users thanked this post: Icchan