Author Topic: Power supply for home lab - do I really need a R&S?  (Read 11385 times)
Offline thm_w
  • Non-expert
Re: Power supply for home lab - do I really need a R&S?
« Reply #50 on: October 03, 2024, 06:01:47 pm »
I think the control loop is in software though. I have an Agilent 66311B which has the same 'problem' (delayed current limit). If the current limit is done in an analog circuit, then I don't see how you can screw up a power supply circuit that bad. OTOH, it could be done to deal with circuits which have a rush-in current which otherwise wouldn't start up properly. Some DC-DC converter modules are prone to not starting when the current limit is set too low. But I can't see this as a good point on a general purpose lab PSU though.
With the 66311B "Mobile Communications DC Source" you could make the argument its not a general purpose PSU, so there will be trade-offs to optimize it for high speed response. I would argue a general purpose PSU should prioritize safety and stability over speed.
Here is an old R&S DC source PSU (ngmo), same issue but much faster response:
Offline mhsprang
« Reply #51 on: October 03, 2024, 06:44:05 pm »
That's totally unacceptable behavior from any supply, it's a disaster in the waiting for just about any sensitive electronics use we can think of.
I don't disagree with that statement but this behavior is quite common for a number of supplies from reputable manufacturers.
I just tested the supplies I currently have on the bench. All set to 8 V, 10 mA, loaded with 50 ohms. The attached pictures show the output voltage when the load is connected to the supply.
HMP4040-E3646A.png: green trace: R&S HMP4040, white trace: Agilent E3646A.
6632B.png: Agilent 6632B, with rear panel "Fast/Normal Operation" switch set to "normal" (notice the time scaling).
The Agilent 6632B hangs at 8 V for about 150 ms.
If the output is enabled before you connect the load, you will obviously see a spike due to the capacitor on the output of the PSU. The idea is that you connect the load first, then enable the output. That is what I have been doing. And in that case, I think no overshoot should be tolerated.
The following users thanked this post: nctnico, thm_w, Kean, Jacon, mawyatt
Offline rf-messkopf
  • Mario H.
    • Homepage
« Reply #52 on: October 03, 2024, 06:56:46 pm »
Okay, I thought you are after the transition from voltage to current regulation. Attached see the test with the output being switched on (green trace: HMP4040, white trace: Agilent E3646A), load is 50 ohms again.
« Last Edit: October 03, 2024, 07:10:30 pm by rf-messkopf »
Offline mhsprang
« Reply #53 on: October 03, 2024, 07:43:02 pm »
Here are the results from the power supplies I have at home. The Delta is a pure analog supply, the TTi is at least digitally controlled, has a keyboard and all, but I don't know of the control loop is analog or digital. The Tenma is also digitally controlled but cheap: €99. Setup is 8V, 10 mA in 50 Ohm. Load connected, then outputs enabled. Two screen shots each with different time bases, to see short term spikes and "long term" stabilisation of the set current.
Offline mawyatt
« Reply #54 on: October 03, 2024, 08:00:51 pm »
That's what we've shown (with Blue LED as load), and agree on acceptable overshoot!!
Best
« Last Edit: October 03, 2024, 08:14:14 pm by mawyatt »
~Wyatt Labs by Mike~
Offline nctnico
« Reply #55 on: October 03, 2024, 08:11:48 pm »
Just to add another data point: The Keysight E36313A behaves OK. Set to 20V, 8mA with a 47 Ohm resistor attached.
I had to use 500mV/div other wise the scope wouldn't trigger.
Offline rf-messkopf
  • Mario H.
    • Homepage
« Reply #56 on: October 03, 2024, 08:12:26 pm »
That's what we've shown, and agree on acceptable overshoot!!
Sure, but still the settling time when transiting from voltage to current regulation varies among different supplies and can be rather long. The Agilent 6632B takes over 150 ms with a 50 ohms load, for example, as I show in the post above. I find this rather disappointing. That will fry a LED instantly when connected to the supply when in CV mode with the output on. I should test that supply with the rear panel switch set to "fast".
Offline 2N3055
  • Country: hr
« Reply #57 on: October 03, 2024, 08:51:24 pm »
Well, that is mostly because of large capacitors on the output. To make it worse, internal impedance of capacitors is in milliohms...
Current sources for testing sensitive devices usually have shorted outputs when you are connecting DUT, and you remove short to ramp up voltage/current in controlled manner. Current sources also can have voltage limiters, where you can set max voltage, and if current source is unloaded, limiters work like parallel regulators. Basically adjustable zener in paralel to DUT.
Actually, that kind of device could be easily constructed and added as additional device in parallel to PSU output to "clamp" any misbehaviour....
Dr. Richard W. Hamming
The following users thanked this post: egonotto, rf-messkopf
Offline rf-messkopf
  • Mario H.
    • Homepage
« Reply #58 on: October 03, 2024, 09:04:56 pm »
The output voltage curve doesn't look like a discharging capacitor, though. Rather, the supply sits there for milliseconds before the voltage ramps down, same with the R&S, but much shorter. That's particularly unfortunate with the 6632B since it is a two-quadrant supply and can sink current up to the maximum rated output current.
But I don't mean to hijack this thread as it is about a different topic.
« Last Edit: October 03, 2024, 09:18:45 pm by rf-messkopf »
The following users thanked this post: egonotto, thm_w, 2N3055
Offline blackdog
  • Please stop pushing bullshit...
« Reply #59 on: October 03, 2024, 09:11:19 pm »
As promised some measurements I made on LAB power supplies I own that have an enable key that are here on the test bench.
Almost all measurements were made under the same conditions.
Of the Agilent and the Siglent LAB power supply I have made additional measurements with in the case of the Siglent a higher current of 35ma and in the case of the Agilent I have increased the voltage and the current setting, this to investigate to what extent their behavior depends on the current setting.
For convenience I have placed a picture of the LAB power supply under test in each picture.
Basic settings
Voltage setting 8V
Current 10mA
The power supply is switched on with the enable button.
Two measurements per LAB power supply channel were taken using the single shot method on the scoop.
This to get a better understanding of the power-on behavior of the LAB power supply.
A Pk-Pk mA measurement is visible at the bottom left in yellow.
Let's start with the Rigol DP832.
First channel-1, but... that is R&S behavior  :-DD
Let's zoom in on the horror.
And now to channel-2 which, by the way, has the same characteristics and virtually the same peak current of about 25mA.
And zoomed in again, that looks distinctly different from channel-1.
Channel three of the Rigol can only deliver a maximum of 5.3V and the current is again set at 10mA.
Hir is clearly visible that this channel behaves well, the peak current remains below 10mA.
The following pictures are of a single channel Siglent LAB power supply and that is the SPD1305.
This Siglent is not so kind to its load at this small current, 36mA peak current.
Lets ZOOM in.
It takes 30mSec for the current to drop to double the value of 20mA, you can do better Siglent!
I wanted to see if this Siglent Power Supply shows a different behavior at larger currents, below is a measurement at 35mA.
At this 35mA, the behavior is much better.
Only a small overshoot is visible and the abberations just after power-up are not significant.
Now shall we go watch some more horror?  ;)
WTF 130ms more than 160mA!
One positive is that the pulse is completely clean.
Now let's see if at 200mA the behavior is below, I increased the voltage to 12V.
There is comparatively less overshoot now, but, but.... 500msec long!
Now to the last one LAB power supply that behaves better and that is the GW Instek GPP4323.
This Power Supply has four channels and channel one and two are the same so I am only showing channel one.
Look R&S, Agilens and Rigol, this is how it should be.
Immediately after turning on, a small abberation is visible but that is not important, the peak current remains below the set 10mA.
This is a measurement performed on channel-4.
Looks good.
There are some abberations in the beginning, but the peak current remains below the set value.
And now for the final measurements, still the GW Instek Power Supply and now channel-3.
This channel can only deliver a maximum of 5.5V and that is what the measurements below were made with.
A little hickup at the beginning, but well below the set peak current, no problemo, are you watching R&S, this is how it should be *grin*.
From these measurements, it is clear to see that the “ current loop” of the GW Instek is a lot faster than the other Power Supply's I measured.
Many of the Power Supply manufacturers make the current loop comparatively slow, then the power supply as a whole is easier to keep “stable”, but everything has its price, which is also visible in my measurements an from others in this topic.
I hope to see more measurements from you!
Greetings,
Bram
Necessity is not an established fact, but an interpretation.
Offline thm_w
  • Non-expert
« Reply #60 on: October 03, 2024, 09:26:39 pm »
Its good that you measured current, probably should standardize on that.
Peak of the first one was 25mA, R&S was 120mA, so not nearly as bad, IMO, but still not good.
edit: Measured on DP832 and initial peak was only ever 18mA maximum with 50 ohm, regardless of current setting. So its only an issue if your current limit is sub ~15mA.
« Last Edit: October 04, 2024, 10:16:29 pm by thm_w »
Offline blackdog
« Reply #61 on: October 03, 2024, 09:38:52 pm »
Hi thm_w ,  :D
I measured the voltage across a 50 Ohm resistor and set the scope channel to indicate the current value.
I think it is all about not exceeding the set current and then it is more obvious that you indicate the current value on the scope image.
That an LED can handle 2 or 4 times the maximum current for time “X” is not so important.
Every LAB power supply is used for all kinds of other applications.
My moral from almost 60 years of experience.
I always test an LED with a series resistor.
If I ever want to push some power into a battery, I always do it via a series diode.
I don't like “fast”, I like to be careful and make sure that I don't break the d.u.t. I connect to a LAB power supply.
Kind regards,
Bram
The following users thanked this post: thm_w, 2N3055, Jacon
Offline thm_w
  • Non-expert
« Reply #62 on: October 03, 2024, 09:40:25 pm »
Agreed, updated my post.
Offline Martin72
  • Testfield Technician
« Reply #63 on: October 03, 2024, 10:32:45 pm »
Fast current control is a good thing for sensitive components if you want to connect them directly to the power supply.
However, there are situations where this is rather counterproductive.
Certain DC/DC converter circuits have an inrush current that can easily be many times the normal input current.
If this is not taken into account when switching on, it can, at best, lead to non-functioning without side effects, but in some cases it can lead to permanent current flow, causing the circuit to “hang”.
As always, the dose makes the poison.
However, I also prefer series resistors for LEDs, it's just in my nature... ;)
The following users thanked this post: pdenisowski
Offline rf-messkopf
  • Mario H.
    • Homepage
« Reply #64 on: October 04, 2024, 12:16:41 am »
The delayed switchover from CV to CC mode, where the supply sits there for a few milliseconds before an exponential ramp down is seen (as expected for a capacitor discharged through a resistor), might actually be implemented on purpose. This delay might also be responsible for the overshoot that is seen when the output is switched on with the enable button with a load connected.
The current control loop itself seems to be much faster. Attached is the result of the following experiment: Supply (R&S HMP4040) is set to 8 V, 150 mA, output on, and loaded with a 50 ohms resistor (i.e., it is in CC mode with 7.5 V at the output). Then an additional 4.7 ohms resistor is switched in parallel to the 50 ohms (i.e., output voltage should jump down to 0.64 V). The transition takes about 1.7 ms (from 90% to 10%) and looks exponential as expected for a cap discharged through a resistor, with no obvious delay before the exponential decrease.
Maybe R&S can comment on this behavior.
If it is intentional, it would be nice if the user had control over it, though.
Edit: Repeated the same measurement, but with the 4.7 ohms resistor switched in by a MOSFET in order to demonstrate that there is no appreciable delay between switching and supply response. Lower trace is the gate drive pulse.
« Last Edit: October 04, 2024, 01:12:16 am by rf-messkopf »
Offline KungFuJosh
  • TEAS is real.
« Reply #65 on: October 04, 2024, 12:58:58 am »
What probe were you using to do these measurements?
Thanks,
Josh
Offline moerm
  • Country: aq
  • pragmatic realist
« Reply #66 on: October 04, 2024, 02:11:00 am »
Maybe it is time we start reconsidering difference between good reputation and legends and lore..
And assign brand value according to actual results and not "reputable brand" urban legends...
I logged in specifically and just to thank you for that.
But being logged in anyway ...
I get it but let me play a bit Cassandra ...
With a *linear* *lab* PSU my main concern is potentially sensible stuff (being powered). You are right, there are exceptions and corner cases in almost everything which leads to a need for deciding for priorities (mine I just told). Besides switchers tend to care little about the quality of input power, many circuits however *do* care.
Some of my - almost all self designed and built - PSUs do have a "don't limit (DUT) inrush current for 3 or 10 ms (selectable) with a 0 ms option (~no tolerance, strictly stick to the set current limit)" feature and most of them have a separate "emergency brake" set to double the set current limit (max the highest output current of that supply). Reason: a supply should be "generous and tolerant" but not to the point of getting damaged itself.
What I've seen here (thanks guys!) would have been, had I known it (I didn't) yet another reason to  design and build my PSUs myself rather than buying "well known brands". I'm not really surprised though that "famous A-tier brands" sell what I call crap; maybe my wording is a bit harsh but "A tier brand" quality certainly is not what I see here (modulo one "B-tier" brand).
Whatever, thanks for all the data (and nice images *g).
« Last Edit: October 04, 2024, 02:14:23 am by moerm »
VxWorks - Yes! Linux - meh. Windows - Thanks no, definitely.
The following users thanked this post: mhsprang
Online ArdWar
  • Country: sc
« Reply #67 on: October 04, 2024, 03:31:26 am »
One (potentially ignorant) question:  has the behavior of the NGE (or other manufacturers' supplies) been outside of its published specifications?
Personally I'm not exactly sure since it is tricky to figure out which specifications override what, and whether those interrelated specifications should be interpreted strictly ("any" exceed) or loosely ("all" exceed).
For example, if we take my NGP800 load current regulation spec strictly (< 0.01% + 5mA) then my peak result of 160mA inrush at startup is well beyond the allowable spec range of 10mA * 0.01% + 5mA, which basically allows for 15.001mA max inrush.
The inrush duration is also well beyond "Load recovery time" spec of 400µs, so can't use that either to handwave.
NGP800 does not specify CV-to-CC transition time. Is this the "get out of jail free" card? :o
Fuse delay time *may* allow this, but that's protection function while the discussed performance should be regulation function. And sure enough if I set the fuse to protect close to the setpoint (20mA) it did indeed trip on itself during startup. Job well done there.
The following users thanked this post: thm_w, pdenisowski, mhsprang
Offline mhsprang
« Reply #68 on: October 04, 2024, 07:53:54 am »
Agreed on inrush currents, but those will be obvious if you switch on the power.
But that is not the point of all of this. If we set a voltage on our PSU, we expect that PSU to output that voltage, no more. Similarly, if we set a current, we expect that PSU to output that current, no more. If a manufacturer specifies that a set current may be +/-5% + 5 mA, I do not expect the output current to be 14 times higher than that. And if the design shows transients at enabling the output, I want to see that beforehand in the datasheet and not after the purchase on my bench.
The following users thanked this post: thm_w, Jacon, nomead
Offline mhsprang
« Reply #69 on: October 04, 2024, 08:02:15 am »
Thanks to everyone for all the analysis and comparison.  As I mentioned above, I've already communicated all of this to the power supply team in Munich and we are investigating.  It would also be very helpful if the OP could contact our technical support about this and/or contact their R&S sales representative as well.
I've seen lots of test results for lots of supplies (and again, many thanks for those), but I haven't seen any comparisons of experimental results to published specifications.  That would have been one of the first things I checked - is a given supply "in spec"?
My apologies if that's a dumb question.
Yes, I sent an email to R&S support yesterday about this.
And yes, I feel this is outside the specifications for exactly the reasons ArdWar just stated in his reply #73, just above.
Offline blackdog
« Reply #70 on: October 04, 2024, 01:26:34 pm »
Hi, :-)
KungFuJosh asked for some explanation of what scope probe I was using, which can be seen below in the picture.
Cost about 1850$ *grin*
The comment above is a hint to the major measuring instrument manufacturers who regularly charge outrageous prices for all kinds of probes.
Probing setup
The tests I did with current limiting around 10mA required a 50 Ohm resistor and about 1-Watt.
I used 2x a 100 Ohm resistor in parallel with a 1M Banana Crock cable from GW Instek and to the scope I used one 1M BNC to Crock cable.
One and all can be seen in the photo below.
I have been thinking about running the measurement setup “HF”.
But on reflection this was not really necessary, this because the LAB power supplies tested have between 100uF and 680uF across the terminals.
This does not allow fast rise times in the measurements.
I use two already older scope probes on my test benches from Micsig for commonmode measurements and current measurements, which are the DP10007 and the CD2100B.
For my work these probes are satisfactory and they are very affordable.
But Micsig also has faster probes available and still affordable prices.
In my measurements, however, these Micsig probes have not been necessary, K.I.S. was a good mindset here, so i used basic cables.
First the pictures
The used probe
Scoop setup
In this photo, I left the settings for channel-1 on, so it is easy to see what I worked with.
Coupling
The coupling is DC
Bandwidth
Bandwidth is 20MHz, otherwise too much noise is visible.
These measurements are about waveform and level and not because I want to see everything that happens while measuring.
I do always start with the “window wide open” when making measurements.
Then I have an impression if there are things that also come into play that I need to take into account.
Pressentation
Most measurements I show on forums are “without” noise, so that the attention goes as much as possible to the measured signal of interest.
Scale factor
So the next setting is the scale of the channel used, and this is set by using the “User1” setting of the Siglent scope.
Here the setting is 50V per ampere.
Input impedance
This should be set to 1M Ohm so as not to blow up your scope.  :-DD
Unit
This should be set to Ampere unit, this because we are measuring the peak current behavior of a LAB power supply.
And now for some comments on LAB Power Supply's.
I've tested a lot of different circuits and I've developed a fondness for the “Harrison Consept.”
That is that the reference section has its own floating power supply and the control current for the power section usually comes from a current source.
Then there are two control loops one for current and the other for voltage.
Which then control the current through two diodes to the Power section.
Nowadays it is a whole lot easier to make a fast responding LAB Power Supply.
The advantage is that it is then possible to keep the capacitor across the output terminals low and still have a fast responding loops with good dynamic behavior.
This has not yet caught on with many manufacturers,
a thick capacitor across the output terminals and slow opamp/power transistors is much cheaper.
Better loop contro en component choices
Furthermore, a current source that always draws current from the output(when the channel is turned on) helps well for excellent dynamic behavior.
There are also opamps available these days that have a large phase and gain margin, look at the ADA4625 models from Analog Devices.
This opamp series has low noise, little bias current, fast and little offset, but above all a phase margin of 88 Degrees at 100pF capacitive load, which is excellent!
Everything I have written here is mine no AI, but translated by deepl.com
And I'm a dyslexic Monkey, which can lead to pretty crooked sentences.  :-DD
Kind regards,
Bram
Edit picture
« Last Edit: October 05, 2024, 02:36:36 pm by blackdog »
The following users thanked this post: nctnico, KungFuJosh
Offline rf-messkopf
  • Mario H.
    • Homepage
« Reply #71 on: October 04, 2024, 03:07:59 pm »
That spec exclusively refers to the voltage regulator. A 'normal' bench PSU with a single series pass regulator is unable to sink current. Thus, when in CC mode and the load resistance drops so that the supply has to lower the output voltage in order to keep the current constant, the excess charge in the fixed output capacitor corresponding to the voltage difference has to be dumped into the load. The supply cannot take up this charge. Therefore, we see the exponential decrease of the output voltage shown a number of times in this thread, and this decrease will depend on the load. There is no way around that, unless the PSU is a two-quadrant one like the Agilent 6632B which can down-program the current, or a SMU.
Well, transients exceeding set values are inevitable when the load resistor suddenly drops in CC mode, as the normal bench PSU cannot sink current, and the transition time will be load-dependent.
The delay time when transitioning between CV and CC mode (I'm not aware of a PSU that actually specifies that delay) is another story. I can see why they seem to be doing that. Otherwise you can get rapid oscillations between CV and CC mode. I've run into that with homebrew PSU designs and complex nonlinear loads long time ago. The question is which delay is appropriate. I think the approx. 150 ms of the Agilent are a bit too long. Would be nice to have control over it, and to have it specified at least.
That overshoot when the supply is activated with the output enable button and with a load connected which draws more current than set is yet another story. I agree that I'd prefer the supply not actually doing that. But one may also argue that it should behave just the same way at startup as when transitioning to CC mode in order to deal with complicated loads. Again, I'd prefer user control over this behavior.
So no idea whether this is in spec or not.  :-//
Offline KungFuJosh
  • TEAS is real.
« Reply #72 on: October 04, 2024, 03:19:37 pm »
Thank you for all the details!
Also... Damn it! I knew it was going to be an expensive probe. 😉
Offline jayk
« Reply #73 on: October 04, 2024, 04:14:55 pm »
All this is making me feel a lot better about my GPP-4323.  If only it had a decent web interface and were a bit quieter under load it would be perfect.  Still a great value for the money.
Offline mhsprang
« Reply #74 on: October 04, 2024, 04:21:57 pm »
If you go back and take a look at reply #37, second image, you see the current spike first, then drop from 20 mA to 10 mA (setpoint) in several steps over a period of 12 seconds. That has nothing to do with any capacitor discharge but everything with something that goes terrible wrong in the firmware.