nctnico:
Just to add another data point: The Keysight E36313A behaves OK. Set to 20V, 8mA with a 47 Ohm resistor attached.
I had to use 500mV/div other wise the scope wouldn't trigger.
rf-messkopf:
--- Quote from: mawyatt on October 03, 2024, 08:00:51 pm ---That's what we've shown, and agree on acceptable overshoot!!
--- End quote ---
Sure, but still the settling time when transiting from voltage to current regulation varies among different supplies and can be rather long. The Agilent 6632B takes over 150 ms with a 50 ohms load, for example, as I show in the post above. I find this rather disappointing. That will fry a LED instantly when connected to the supply when in CV mode with the output on. I should test that supply with the rear panel switch set to "fast".
2N3055:
--- Quote from: rf-messkopf on October 03, 2024, 08:12:26 pm ---
--- Quote from: mawyatt on October 03, 2024, 08:00:51 pm ---That's what we've shown, and agree on acceptable overshoot!!
--- End quote ---
Sure, but still the settling time when transiting from voltage to current regulation varies among different supplies and can be rather long. The Agilent 6632B takes over 150 ms with a 50 ohms load, for example, as I show in the post above. I find this rather disappointing. That will fry a LED instantly when connected to the supply when in CV mode with the output on. I should test that supply with the rear panel switch set to "fast".
--- End quote ---
Well, that is mostly because of large capacitors on the output. To make it worse, internal impedance of capacitors is in milliohms...
Current sources for testing sensitive devices usually have shorted outputs when you are connecting DUT, and you remove short to ramp up voltage/current in controlled manner. Current sources also can have voltage limiters, where you can set max voltage, and if current source is unloaded, limiters work like parallel regulators. Basically adjustable zener in paralel to DUT.
Actually, that kind of device could be easily constructed and added as additional device in parallel to PSU output to "clamp" any misbehaviour....
rf-messkopf:
--- Quote from: 2N3055 on October 03, 2024, 08:51:24 pm ---Well, that is mostly because of large capacitors on the output. To make it worse, internal impedance of capacitors is in milliohms...
--- End quote ---
The output voltage curve doesn't look like a discharging capacitor, though. Rather, the supply sits there for milliseconds before the voltage ramps down, same with the R&S, but much shorter. That's particularly unfortunate with the 6632B since it is a two-quadrant supply and can sink current up to the maximum rated output current.
But I don't mean to hijack this thread as it is about a different topic.
blackdog:
Hi,
As promised some measurements I made on LAB power supplies I own that have an enable key that are here on the test bench.
Almost all measurements were made under the same conditions.
Of the Agilent and the Siglent LAB power supply I have made additional measurements with in the case of the Siglent a higher current of 35ma and in the case of the Agilent I have increased the voltage and the current setting, this to investigate to what extent their behavior depends on the current setting.
For convenience I have placed a picture of the LAB power supply under test in each picture.
Basic settings
Voltage setting 8V
Current 10mA
The power supply is switched on with the enable button.
Two measurements per LAB power supply channel were taken using the single shot method on the scoop.
This to get a better understanding of the power-on behavior of the LAB power supply.
A Pk-Pk mA measurement is visible at the bottom left in yellow.
Let's start with the Rigol DP832.
First channel-1, but... that is R&S behavior :-DD
.
Let's zoom in on the horror.
.
And now to channel-2 which, by the way, has the same characteristics and virtually the same peak current of about 25mA.
.
And zoomed in again, that looks distinctly different from channel-1.
.
Channel three of the Rigol can only deliver a maximum of 5.3V and the current is again set at 10mA.
.
Hir is clearly visible that this channel behaves well, the peak current remains below 10mA.
.
The following pictures are of a single channel Siglent LAB power supply and that is the SPD1305.
This Siglent is not so kind to its load at this small current, 36mA peak current.
.
Lets ZOOM in.
It takes 30mSec for the current to drop to double the value of 20mA, you can do better Siglent!
.
I wanted to see if this Siglent Power Supply shows a different behavior at larger currents, below is a measurement at 35mA.
At this 35mA, the behavior is much better.
.
Only a small overshoot is visible and the abberations just after power-up are not significant.
.
Now shall we go watch some more horror? ;)
WTF 130ms more than 160mA!
.
One positive is that the pulse is completely clean.
.
Now let's see if at 200mA the behavior is below, I increased the voltage to 12V.
There is comparatively less overshoot now, but, but.... 500msec long!
.
Now to the last one LAB power supply that behaves better and that is the GW Instek GPP4323.
This Power Supply has four channels and channel one and two are the same so I am only showing channel one.
Look R&S, Agilens and Rigol, this is how it should be.
.
Immediately after turning on, a small abberation is visible but that is not important, the peak current remains below the set 10mA.
.
This is a measurement performed on channel-4.
Looks good.
.
There are some abberations in the beginning, but the peak current remains below the set value.
.
And now for the final measurements, still the GW Instek Power Supply and now channel-3.
This channel can only deliver a maximum of 5.5V and that is what the measurements below were made with.
.
A little hickup at the beginning, but well below the set peak current, no problemo, are you watching R&S, this is how it should be *grin*.
.
From these measurements, it is clear to see that the “ current loop” of the GW Instek is a lot faster than the other Power Supply's I measured.
Many of the Power Supply manufacturers make the current loop comparatively slow, then the power supply as a whole is easier to keep “stable”, but everything has its price, which is also visible in my measurements an from others in this topic.
I hope to see more measurements from you!
Greetings,