Author Topic: Powerbanks tested with DHO800/900  (Read 31721 times)
Offline Houseman
Re: Powerbanks tested with DHO800/900
« Reply #25 on: February 13, 2024, 10:18:54 am »
Anker 737  24,000mAh (powercore 24k) FAILED !
It is a shame if your Anker737 dont work with your Rigol, as the Anker737 (Powercore24 A1289) does seem to work with the newer Rigol products, but it could be they changed something from the older Rigol products to the newer Rigol models.
I can CONFIRM that even if new generation IT DOES NOT WORK with RIGOL DHO914.
Even 140W Seires 7 Generation 2 these beefy PB does freeze the Rigol at startup since cannot go beyond 20W. EITHER plugged in C1 and in C2 jacks. See attach
DO NOT BUY IT.
I am asking before claiming for refund if Anker will release a new firmware for this item...
« Last Edit: February 13, 2024, 10:21:43 am by Houseman »
Offline DaneLaw
« Reply #26 on: February 14, 2024, 03:03:34 am »
DO NOT BUY IT.
Not sure what Rigol had in mind with their DHOxxx line, and could look like they placed themself between two chairs 12V PD & 15V PD
As mentioned newer products from Rigol, using the same form factor, works on all the high current powerbanks I got access to.
An example of the start-up cycle stays constantly in 15v PD.
You can obviously decoy trigger a locked 15V PD, then it should work according to the info on page1.
What I do with my Micsig scope that doesnt have native PD, so can use most of these PD powerbanks with that unit.
« Last Edit: February 14, 2024, 03:09:57 am by DaneLaw »
The following users thanked this post: Houseman
Offline Proxy64Topic starter
  • Country: th
« Last Edit: February 14, 2024, 03:28:14 am by Proxy64 »
  • Country: th
« Reply #28 on: February 14, 2024, 04:28:23 am »
Here a test with a PD trigger board to use an Anker powerbank to power a DHO800/900.
https://youtu.be/CBg0HLug-W4?si=Qaa56oDNEBTkxZSL
edit: Note that the FNB58 is only here to display the voltage and power.
« Last Edit: February 14, 2024, 04:47:15 am by Proxy64 »
  • Country: th
« Reply #29 on: February 15, 2024, 01:33:10 pm »
Based on to this video Shargeek Storm2 works
https://youtu.be/TXL916et9_Y?si=ufldtgFa_a2auhS4
Offline Houseman
« Reply #30 on: February 16, 2024, 04:48:13 pm »
Thank You all. Would like to know a way to measure how much noise in injected into the rigol via power supply with this setup.
And why the power supply provides 12V instead of 15V provided by the powerbank
« Last Edit: February 16, 2024, 04:50:12 pm by Houseman »
Offline ebastler
« Reply #31 on: February 16, 2024, 05:33:51 pm »
Where "the power supply" means which power supply??
Only the very early power supplies shipped with beta units (and maybe to early Chinese customers) were of the cheap fixed 12V type described in the user manual. All DHO800/900 scopes shipped to the Western world should come with proper supplies which negotiate the voltage via the USB PD protocol, and should negotiate 15V.
Rigol have just neglected to update the user manual. Or did you actually measure 12V being used with your scope's power supply?
Offline Demon Xanth
« Reply #32 on: February 16, 2024, 09:18:08 pm »
Baesus 140W powerbank DOES NOT work works on the C2 output, but not the C1 input/output. Very curious.
Edit:
The C2 and USB-A outlets cut out due to lack of current, the C1 seems to have a handshaking issue. The 65W works fine though.
« Last Edit: February 19, 2024, 08:20:25 pm by Demon Xanth »
The following users thanked this post: Proxy64
Offline shapirus
  • Country: ua
« Reply #33 on: February 17, 2024, 07:19:55 pm »
This one works too:
The PSU here is a 69W Ugreen. The battery is a 3s LiPo.
Curiously, just like the Baseus mentioned in the previous post, it doesn't work (or the scope doesn't work) on port USB-C1. The other two, USB-C2 and type A, work fine.
update: I checked the specs, and now it's obvious: the specs state 20W max on port C1, 60W max on port C2. But then again, it says max 22.5W on port USB-A, and the scope works fine powered from that port.
« Last Edit: February 17, 2024, 07:57:18 pm by shapirus »
Offline Houseman
« Reply #34 on: February 19, 2024, 11:13:21 am »
Nice!.
Isn't it a little bit undervoltage powered?
I see 11,1V instead of minimum 12V
How long does it last?
Regards
Offline Houseman
« Reply #35 on: February 19, 2024, 11:14:44 am »
Ah, ok got it now.
It is not just a cigar plug.
Offline shapirus
  • Country: ua
« Reply #36 on: February 19, 2024, 11:30:03 am »
Ah, ok got it now.
But it is. It's a plain automotive 12V USB power supply designed for the cigarette lighter socket, only of the fancier ones -- higher power and with support of PD, QC and all the type C paraphernalia.
Haven't measured. Will be equal to: battery energy capacity * DC/DC converter efficiency * wires and connectors loss factor / scope power. Very rougly, that's a 50 Wh battery, let's assume 80% conversion efficiency, then at 35 W consumption it's going to last 50 Wh * 0.8 / 35 W ~= 1.14 h.
Speaking of min voltage, it seems to work fine at (currently) ~11V, and there's really no reason for it not to: there's a plain switching buck-boost converter inside, and they typically have a wide range of acceptable input voltages. I will test it later with a lab PSU to see how low a voltage it can accept on the input.
Offline jsobell
« Reply #37 on: February 22, 2024, 10:54:42 am »
I can confirm that the INIU 25000mAh unit works fine - https://www.amazon.com.au/dp/B08VDJP7WN
I timed it, and it gives almost exactly two hours out of a 100% charge.
That's not overly impressive, as at 3A draw it mathematically should be over 7 hours, and even allowing for losses it would be nice to have more than 25% of the power delivered.
But, if two hours is sufficient, then this model works fine and is nice and cheap.
I've also ordered a 40,000mAh unit to test, and I'll post if it works or not.  It's a bit of a minefield choosing the packs, as some are only 5V, and many split their current over several ports, rather than a single one with 15V@3A as we require.
Offline shapirus
  • Country: ua
« Reply #38 on: February 22, 2024, 11:11:50 am »
No that's actually good enough. It's 25000 mAh at 3.7 V -- the usual disgusting trick of the powerbank manufacturers who don't specify capacity at the actual battery voltage, or, better, energy capacity. Always use energy -- measured in Wh -- to estimate the battery run time. This makes the battery 25 Ah * 3.7 V = ~92 Wh, and at 35 W power draw it theoretically means slightly over 2.6 hours. Apply the conversion losses and the chinese Ah derating, and 2 hours run time doesn't look bad at all.
Offline jsobell
« Reply #39 on: February 23, 2024, 03:47:10 am »
I confirm that powerbank INIU BI-B63 clamed as PD 100W 25000mAh does not complete Rigol's boot.
Fan goes on but after a while DHO shuts down.
Asked for refund.
Interesting, I have the same one and have no issues running the scope.
I can use it in either output socket (45W or 100W) and I get about 110mins out of a full charge.
The pack is rated to 3A@15V, which meets the requirements.
Offline jsobell
« Reply #40 on: February 23, 2024, 07:41:33 am »
That is interesting, and I see what you mean. I guess it's a tricky one to specify accurately, as you can only specify the battery capacity as it supports so many voltages. It is a bit disappointing though!
The good thing about this particular pack is that it also works in pass-through, so you can leave it connected and charge it (quite slowly) while running the scope.
I'm just running a test on a 40000mAh pack to see how that fares, but it doesn't support pass-through, so may not be worth the extra hassle. It reports 2.3-2.4A of current draw at 15V, so on that basis (40Ah * 3.7V = 148W) it's got over 4 hours of running.
I'll post up the actual running time once the scope dies :)
J.
Offline shapirus
  • Country: ua
« Reply #41 on: February 23, 2024, 10:11:19 am »
This is exactly the reason why specifying not charge, but energy capacity, in Watt-hours, and conversion efficiency at different output voltages, is the right and universal way of doing it. Having these two, the customer can adequately compare different powerbanks and estimate how long they are going to last in a particular scenario.
Make sure you make no mistake in units calculation to avoid misunderstanding the concepts of power, current, charge, and energy. Multiplying Ah and Volts won't yield Watts:
1 A = 1 C / 1 s [current, aka charge over time]
1 W = 1 J/s [power, aka energy over time]
1 A * 1 V = 1W
1 A*h = 1 A * 1 h = 1 A * 3600 s = 3600 C [charge]
1 W*h = 1 A*h * 1 V = 1 W * 1 h = 1 W * 3600 s = 3600 W*s = 3600 J [energy]
Offline LonnieU
« Reply #42 on: February 25, 2024, 02:16:11 am »
I purchased a Baseus Model BS-30KP365 for a USB Soldering Iron and for other general purpose usage awhile back.
https://www.amazon.com/s?k=B08JV4W4NY
Ran my DHO924 for an hour and 45 minutes today and it took the pack fully charged down to 56 percent. Negotiated for 15vdc and ran 2.1-2.2 amps.
Offline jsobell
« Reply #43 on: February 28, 2024, 10:45:07 am »
J.
Yes, I get almost exactly 4 hours of running from the ROMOSS 65W 40,000 mAh pack.
The only thing I'm not keen on is that it only has a single USB-C ports, so it doesn't support pass-through and you can't leave it connected to recharge when the scope is powered off.
Offline jim_griff
  • Personal Text goes here.
    • Principal Audio
« Reply #44 on: May 09, 2024, 02:32:31 am »
For those who are curious, you CAN indeed hook up your Rigol DHO800 directly to a battery. I've tested between 11V and 15V on the bench PSU connected to a USB-C cable and the scope powers up fine. It draws roughly between 3.1A and 2.8A depending on input voltage.
I haven't tested in-depth to see if there are any long-term issues when running at lower voltages, but the fact it powers up shows there is no USB PD negotiation going on, so no trigger board is necessary to "trick" it into working.
I'm going to build a 3S LiPo pack with in-line fuse and BMS protection to run mine when out and about. Don't see the point of inefficiencies in power banks, unless there is a solid requirement for it to constantly have 15V on the scope's USB-C power input. Maybe I'll figure out some issues later down the line! However, the scope has many Vregs inside it anyway which are serving all the necessary voltages, so I don't see any point regulating the input to 'precisely' 15V.
The following users thanked this post: shapirus
Offline aXit
« Reply #45 on: May 09, 2024, 03:22:27 am »
What's the UVLO behaviour of the scope for voltages under 11V? 3.66V per cell could be less than 50% capacity used, depending on the exact chemistry.
Offline ebastler
« Reply #46 on: May 09, 2024, 06:24:55 am »
It's known that the scope works without PD negotiation, since the original power supply (shipped with the very first units only) was a cheap fixed 12V type. The user manual, in section 4.3.2, still describes that supply and specifies a 12V/4A power requirement for the scope.
The following users thanked this post: jim_griff
Offline shapirus
  • Country: ua
« Reply #47 on: May 09, 2024, 08:00:59 am »
I would guess that it should work fine down to 10V or maybe even less. Everything gets converted to lower voltages anyway, the highest probably being the fan power rail at its 8V.
If that is so, then the only remaining concern should be the somewhat increased current (compared to 15V and 12V) that the tracks going from the USB socket to the switching converters will have to carry. But unless they are super thin, that shouldn't be an issue.
Offline csuhi17
  • Country: hu
« Reply #48 on: May 09, 2024, 08:14:14 am »
I would mainly be interested in the maximum voltage it can withstand. :-BROKE
Fnirsi oscilloscope = waste&regret
Offline jim_griff
    • Principal Audio
« Reply #49 on: May 10, 2024, 01:26:25 am »
I've had it down to 9.3V here (3.1V per 18650 LiIon cell). Was still functioning properly, but I wanted to stop the test there as I didn't want to damage the cells I was using.
With 2500mAh cells (a suspect figure for cheapo cells), I managed to run it for approx. 15 mins, give or take a few mins. Didn't set a timer. When recharging from 3.1V to to 4.15V, the cells each took in an average of 1350mAh.
In terms of the maximum voltage the scope can withstand, I haven't tested haha! I was thinking about testing a 4S LiIon battery (16.8V max), but the scope is a lot of money to put on the line for a simple battery test. Since it works at 12V and quite a bit below, I'm not so happy to test it up to 16.8V. :)