quarks:
Unfortunately I totally missed this very interesting information until today.
Great post, many thanks for sharing.

dr.diesel:

--- Quote from: quarks on February 27, 2016, 06:02:13 am ---Unfortunately I totally missed this very interesting information until today.

--- End quote ---

Yeah, I can't keep up anymore either.  Sure wish @mentions worked!   :scared:

dl1640:
Very informative investigations. I set up a 3458A with OCOMP DELAY5 NPLC100 4W to check a 100 ohm class .005 resistor, I'm going to see how good it can get. The resistor is obviously out of spec and I need to use it by this calibrated value. :)

Dr. Frank:

--- Quote from: dl1640 on December 01, 2017, 01:45:07 pm ---Very informative investigations. I set up a 3458A with OCOMP DELAY5 NPLC100 4W to check a 100 ohm class .005 resistor, I'm going to see how good it can get. The resistor is obviously out of spec and I need to use it by this calibrated value. :)

--- End quote ---

This delay is necessary only for R >= 10k.
For 100Ohm, you can have a delay of a few ten msec, otherwise you waste time.

Frank

dl1640:
Thanks Frank,
I will try different delay time.