Products > Test Equipment

Problems with CH 4 in Lecroy LC574AM Error CL_match_gains_or_offsets: 4,8.

(1/1)

hafrse:
Hello,

I have newly acquired a used LC574AM, but it seems there is a problem in channel 4. When I connect to the cal signal and press Auto Setup, a message appears in red  CL_match_gains_or_offsets: 4,8.

The displayed trace shows noise only when Volts/Div is lower  or equal  0.2V.

Any clue where to look? the instrument had manual Reset a number of times but still the problem shown. The other 3 channels are ok.

Please see the attached images, last image is when zoomed
Thanks in advance!

George

nctnico:
With a bit of luck you can swap the EXT input module with the module of channel 4. But likely this is a contact issue to re-seating the input module itself might do the trick.

hafrse:

--- Quote from: nctnico on August 16, 2024, 05:30:51 pm ---With a bit of luck you can swap the EXT input module with the module of channel 4. But likely this is a contact issue to re-seating the input module itself might do the trick.

--- End quote ---
Thanks, do I need to open the instrument or just release the 2 screws and slide back?

nctnico:

--- Quote from: hafrse on August 16, 2024, 05:44:04 pm ---
--- Quote from: nctnico on August 16, 2024, 05:30:51 pm ---With a bit of luck you can swap the EXT input module with the module of channel 4. But likely this is a contact issue to re-seating the input module itself might do the trick.

--- End quote ---
Thanks, do I need to open the instrument or just release the 2 screws and slide back?

--- End quote ---
You'll need to open & disassemble the oscilloscope quite a bit. On my Lecroy Wavepro 7200A it was a difficult job especially due to Lecroy using threaded standoffs which may unthread at the wrong end. As in turning the entire threaded standoff instead of releasing the screw. You'll see once you open it up. Be really carefull not to bump / scrape the casing onto the boards. If you can flip burgers at McDonalds, you can also have a career as a senior mechanical engineer at Lecroy  :rant:

The screws at the front just hold a bit of plastic in place to make the connectors stick out nicely.

Navigation

[0] Message Index

There was an error while thanking
Thanking...
Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod