Here's a result of an ongoing (3hr+) where you can see the effects of small temperature changes
at the DUT. These changes are due to the AC turning OFF/ON and a reduction in outside ambient temp. As you can see our "home office lab" is far from a proper temp controlled lab. With the night approaching the AC doesn't cycle as much (note the somewhat sinusoidal temp changes with a cycle time of ~15 minutes) because of the lower outside ambient. Without a local AC cooling effect as often the DUT temp rises above the earlier level before returning to a higher level. Later the room AC will be dropped 1 degree F (77) for more comfortable sleeping.
Normally we would use the DMM6500 or KS34465A trending plots and visually correlate these plots with the DUT cap measurements, or as planned later create a custom Python routine to bring these together. With TC from HKJ and Josh's LCR interfaces, we won't need to develop the Python routines!!!
These measurements were accomplished with an old Mac Pro, LAN connected to a Hioki IM3536 and DMM6500, with custom DUT fixture shown above. This illustrates the superb performance of the Hioki IM3536, which should be expected for an LCR meter of this class!!
Anyway, from the plot chart you can see the rough temp coefficient of the 20nF Mica is 25~30ppm which agrees with some earlier temperature/cap measurements. This is pretty good for a Mica cap, or any cap!! Some of these Micas have already demonstrated excellent repeatability and very high "Q", so looking forward to continuing this with the help of TC