Hello,

as the 8845A does not have that feature build in the software I simply used the math with offset and multiplicator to get the temperature shown
it is in kOhms range, but the value represents 1:1 to °C

Used a cheap 2x3mm 100 Ohms 38.5K PTC

build myself a kelvin cable with detachable sensor cutting an USB cable in half
- cutting off the male USB plug
- female side has on the cut end to 4 banana plugs
- male side has the PTC soldered with 4 silicone leads as Kelvin probe
- so the actual cable resistance is eliminated

math can be also adjusted to calibrate the PTC in ice and boiling water

https://youtu.be/FFzoJUZd9ak
Title: Re: PTCs on Fluke 8845A bench DMM using math functions
Post by: kado on May 10, 2019, 07:10:00 am
[emoji106] Thanks for sharing.
Karsten


Gesendet von iPhone mit Tapatalk
Title: Re: PTCs on Fluke 8845A bench DMM using math functions
Post by: voltsandjolts on May 10, 2019, 07:53:51 am
I think using a 100Ohm PRTD would be a better option.
They are supported by the 8845/46 for degc readout and are widely available for a couple of bucks (smd chip PRTDs).
Title: Re: PTCs on Fluke 8845A bench DMM using math functions
Post by: BLF Lexel on May 10, 2019, 10:31:36 pm
I think using a 100Ohm PRTD would be a better option.
They are supported by the 8845/46 for degc readout and are widely available for a couple of bucks (smd chip PRTDs).

I use a 100Ohms Platinum 3x4mm chip with 38.5Ohm/100K
the 8845A seems not to have °C option for them, or am I missing something?
Title: Re: PTCs on Fluke 8845A bench DMM using math functions
Post by: voltsandjolts on May 12, 2019, 04:15:13 pm
Apologies, I skim read your post and I misunderstood it.
I use the the terms NTC/PTC for thermistors and PRTD for platinum sensors - I thought you were using a thermistor not a PRTD.
And I was doubly wrong  :-\ ..8845 doesn't have PRTD degc readout like the 8846, hence your post.
Hah, ignore me and carry on!