I have a Quadtech 1880 milliohm meter that I would like to calibrate but I can find nothing out online as to the exact procedure.
So far, it seems to be:
1)  Turn unit on
2)  Move calibration switch to enable position
3)  Unit shows "0", short probes and press the "Man/Ext" key - unit does call.
4)  Press "Man/Ext" again, unit shows required reference value.  Connect to probes, press "Man/Ext" - unit does cal
5)  Continue steps 3 & 4 through 1.9M range; unit resets.
6)  Move calibration switch to disabled.
7)  Cycle power, done.
What is happening is that after this, meter only reads somewhat correctly if the measured value (1st digit) is in the lower part of the current range.  A value of 1.5 milliohms, 1.5 ohms, 1.5 K or 1.5M measures correctly; values of 4 milliohms, 4 K, 6 milliohms cause erratic readings.
Question is, it this due to a bad cal or are there hardware issues with the meter.  If I knew for sure the exact calibration procedure, it would at least help me to know where to start troubleshooting.
TIA,
Hal
--- If it isn't broken... Fix it until it is ---
Online Kleinstein
Re: Quadtech 1880 Milliohm eter calibration procedure
« Reply #1 on: August 12, 2019, 05:13:14 pm »
This sound a bit like a hardware problem, like a wrong current source causing the source to drive into saturation.
With a militohm meter, I would suggest using 4 wire technique. So the short may be not just connecting 2 wires. Or closing Kelvin probes together. There are special 4 way shorts. Other wise the sens wires should be connected, the drive wire should be connected and some wire / resistance between the 2 pairs.
Offline Shock
« Reply #2 on: August 12, 2019, 05:35:02 pm »
Contact distributors or manufacturers for the schematic, service manual and calibration info whatever they can provide. Manufacturers often include test points or short points to confirm the test equipment agrees with itself and is operating correctly.
You mentioned erratic measurements, does this mean reproducible fluctuating measurements or reproducible inaccurate measurements? There is a big difference between the two.
Are you using the original kelvin leads sold with the device as it may be similar to my micro ohm meter which has two 100k resistors between each drive and sense.
First you must rule out the components under test to ensure they are not complicit in the inaccuracies, then if it's not try and include offsets in your calibration to see which way the inaccuracies swing. You might be able to characterize where it is occurring. Obviously you should be able to produce reproducible measurements regardless of accuracy or drift in ranges.
Probably about that time you might want to scope the calibration procedure and during normal measurements to see what is different. Also you can verify the power supply noise/ripple and general component quality. Normal warnings about connecting the scope ground clip to a circuit that is not isolated/floating or at ground potential apply.
Offline BobBeBob
« Reply #3 on: January 08, 2021, 08:13:53 am »
HI there, I was wondering what happened with the issue you were having with your quadtech 1880?
I just got one and it seems it is not reading anything. I zero it out. but it keeps reading the same values (-.5522 to -.5421) no mater what is connected to it.
where you able to find any schematics or a service manual for it?
« Reply #4 on: January 08, 2021, 08:18:29 am »
This is what I have for the 1880. Mine works great btw  :)
The following users thanked this post: BobBeBob
Offline BobBeBob
« Reply #5 on: January 08, 2021, 09:47:22 pm »
thank you for your reply,
I have those manuals . they are basically the user manual and the brochure. I need the service manual if it exists with schematics and or pictures of the big flat board inside.
Offline BobBeBob
« Reply #6 on: January 08, 2021, 10:26:28 pm »
Sorry again, I am just figuring out this blog. I was replying to your massage in a different way. so you may have several reply.
I have the manual and brochure.
I need either the schematics or a few pictures of the board setting flat in the unit.
on mine it looks like someone had modified it for their use.
there are 2 snipped wiring of 5 wiring that extent from the rear of the test lead connector to the board. snipped off wiring connect to the right of the connector on a couple of diodes. and a few resisters around the board looks like done after the manufacturing. I could send you pictures if you want.
I  need pictures of a good unit please.
thank you
Offline BobBeBob
« Reply #7 on: January 08, 2021, 10:30:12 pm »
you ever got that quadtech 1880 calibrated and working?
« Reply #8 on: January 09, 2021, 03:07:57 am »
Sounds like your unit is the same as mine internally and was a 'factory bodge'  :palm: These photos should be easy to compare against yours for a start. I will leave the top off mine for a couple of days if you want any voltages measured or photos elsewhere on the board.  :)
Offline BobBeBob
« Reply #9 on: January 10, 2021, 04:47:38 am »
If you wouldn't mind can you please take pictures of all around that board please and send to me. Pictures you sent looks the same as mine so far. On mine there is a few more places that look bodged.
« Reply #10 on: January 10, 2021, 05:03:16 am »
Overall and three separate ones for a bit more detail. The tight hold on a service manual and calibration service for these now older models is complete  :bullshit: time they pulled their heads out of their ....
« Reply #11 on: January 10, 2021, 05:03:53 am »
part 2
« Reply #12 on: January 10, 2021, 09:04:16 am »
Still no luck on the correct procedure by me.  Mine has/had the same bodge (I changed the connector to a standard DIN to facilitate making new test cables) -  I had been wondering if that was factory or not.
Hal
« Reply #13 on: January 10, 2021, 09:10:38 am »
The bit of recent trawling I did indicates the use of a custom 0.019. 0.19. 1.9  \$\Omega\$ set etc for $2k+ I have also looked and searched in the past for the Original Manufacturer "Chen Hwa Tek" with no success. Chroma the now brand owner seem to want to tie the units to return to them only so they can gouge you for $ instead of helping with an older model.  :--
« Reply #14 on: January 11, 2021, 01:45:55 am »
Naked Bum shot for BobbeBob of the front panel PCB connection area.
Offline BobBeBob
« Reply #15 on: January 11, 2021, 04:51:11 am »
Thank you. I took some smaller pictures Hopeful these will make it through.
would you mind taking more pictures from the back as many as you can please.
thank you
kind Regards
Bob
Offline BobBeBob
« Reply #16 on: January 11, 2021, 04:55:26 am »
more pics of the mine, modified one.
Offline BobBeBob
« Reply #17 on: January 11, 2021, 04:57:07 am »
more pics of the modified one
« Reply #18 on: January 11, 2021, 05:06:20 am »
Interesting Bodge  :-\ Across the large Pad R26 where mine has the 1 \$\Omega\$ 10W resistor on the chassis with wires going to it yours has what looks like a 1/4W resistor? What is the value on it might give a clue as to what some of the other bodges were setting out to achieve  :-//
Offline BobBeBob
« Reply #19 on: January 11, 2021, 05:08:56 am »
you got to see the bodge work on mine. I sent a bunch of pictures. the real mess is in the rear.
I was planning to remove all the add on component and fix the cut traces to see what happens.  I will leave the ones that match Beanflying boards.
just waiting for a few more pics.
Offline BobBeBob
« Reply #20 on: January 11, 2021, 05:10:45 am »
The value seems to be 100K,
Brn, Blk, Grn, Gld
Offline BobBeBob
« Reply #21 on: January 11, 2021, 05:12:47 am »
Or is it 1m ohm?
« Reply #22 on: January 11, 2021, 05:18:22 am »
1 Meg \$\Omega\$. Without tracing the circuit out or trying to sort out your rats nest from photos at a very rough guess they were trying to reduce the test current or change the range able to be tested by the unit while maybe making use of its tolerance/sorting functionality?  :-//
Offline BobBeBob
« Reply #23 on: January 11, 2021, 05:43:23 am »
I just double checked. it is a 100K.
I figured I will just remove everything they added and reconnect the traces, Add the same resister you have on yours. if I see signs of hope, them will think of calibration.
I am open to suggestions thought.
Offline BobBeBob
« Reply #24 on: January 12, 2021, 07:54:31 am »
Hi, I was wondering if it is not too much trouble for you, at you convenience, Please send some more picture of the back side of the PCB. I need enough pictures to be able to see the entire PCB back side. please.
thank you
regards
Bob B
Offline BobBeBob
« Reply #25 on: January 14, 2021, 05:34:09 am »
Hi, any luck getting a few more pictures from the back of the PCB please?
Offline BobBeBob
« Reply #26 on: February 21, 2021, 09:30:38 am »
Hi There,
any luck with those pictures please?
Offline BobBeBob
« Reply #27 on: March 12, 2021, 09:23:24 am »
Hi it's me again, sorry to bother you. But I was wondering if you would send me a few more pictures from the back of the board of the quadtech 1880 please. pretty please.
I have a picture here from you but I need more to the right of the board. the pictures you send are is very nice and clear but I wish I had some more images of the back.
I am attaching some thermal images of my board which indicate a short I think on U29, which on yours is a OP77EP, But on mine is a LTC1052CN8. I have a 0P77 on order and will be replacing and will look for any change.
Also below chip U31 is a resister I want to guess is R51, (the number is under the piece and not able to read it), on one side the resister traces out to U30 pin 5. well, there is what appears to be a Zener Diode next to it. I want to know if the resister and the cathode side of the zener diode are connected in the back of the board. this is one of the reasons I need more pictures of the back of the board please.
Offline BobBeBob
« Reply #28 on: March 12, 2021, 09:47:56 am »
Hi, so I seem to found a short on U29, which is a OP77EP on you board but on mine is something else. I attached thermal images for viewing. ordred the OP77EP, we will see what happens.
I could use some more pic from the back of the board.
one reason I need more pictures is because I need to see if the resister in below chip U31 which one side of it traces out to pin 5 of chip U30, is connected to the zener diode cathode side. it looks as if the trace on mine was scrapped off in the back.
Offline BobBeBob
« Reply #29 on: March 12, 2021, 09:48:54 am »
more thermal pics
Offline BobBeBob
« Reply #30 on: March 12, 2021, 09:50:13 am »
more thermal pics
Offline BobBeBob
« Reply #31 on: March 13, 2021, 03:33:13 am »
 279 F. I think the measurement is within 0.2 C accurate. but as soon as I took U1 chip out of socket the heat went away and the rest of the chips turned on with normal on.
Offline BobBeBob
« Reply #32 on: March 13, 2021, 03:46:57 am »
After undoing the birds nest and soldering several cut traces. I think I did it correctly. though not sure if the cathode of zener diode next to R53 is connected to it or not. I connected it. Ran out of patience, crossed my fingers, plugged the unit in and turned it on. No sparks but, It was still not making any measurement. it was stuck on 4.232 M ohm. it was stuck on that number in every range but decimal was moving. before undoing the birds nest the unit was stuck on 5.5xx M ohm, That's when I decided to get the short finder for dummies out, the thermal camera. I was not looking for a short but just wanted to see what is on. And yes, there was a short there. But, unfortunately, I did not take a before thermal picture. So, I don't know if removing the birds nest and connecting R53 to the cathode of the zener next to it caused this issue or was it this way before.   If anyone has picture of the back side of the board I would appreciate it if you could send some clear pictures of the backside of the board of quadtech 1880. thanks.
Bob
Offline BobBeBob
« Reply #33 on: March 13, 2021, 05:10:21 am »
sorry everyone for multiple posting the same. I just figured out there are other pages to the threads. figuring out EEVblog is like the first day driving on the opposite side.
Offline BobBeBob
« Reply #34 on: March 13, 2021, 05:20:25 am »
Ok, my apologies, I think I made a few mistake, let me just make a quick correction.
U29 I think is the chip that has the short, it is OP77EP.
R51 it appear to be the resister that I would like to see if it is connected to cathode of the the zener diode next to it. these 2 are next to U31 pin 1.
any info is helpful
thanks