alm:
Looks like a nice unit! Thanks for sharing those pictures! The downside of (alkaline) AAA cells is leakage, so be sure to check and consider if the unit could take primary lithium cells in the same form factor (mind the slightly higher open terminal voltage).

Re the calibration kit the ideal scenario wow be to buy it and have someone measure the Q. Because even parasitic inductance and capacitance in the measurement setup could affect it. So Ideally you'd mount the resistor in an aluminum (eg. Pomona) box with BNC connectors and then characterize the assembled unit with another calibrated unit that you have access to, for example through friends, work or uni. I'm guessing this is the original cal kit came with BNC connectors? Or was it just leaded resistors?

coromonadalix:
not being rude

you are going into a loop,    the problem is:  you need to build and characterize the builded jig(s) .....  BUT you need a calibrated equipment to do so 

i think the best would be to: once calibrate your lcr,  build the test jigs and caracterize them,  and  keep them in safety  ...


my 2 cents

Vgkid:
Thanks for tearing it down. That is a lot more complex  than I would have thought, certainly more than my gr1689. Interestingly enough , the reference resistors are the same, except your 24.9 is a custom EBG, while mine sticks with the S102.

RAQU_PL:

--- Quote from: alm on May 23, 2023, 10:53:17 am ---Looks like a nice unit! Thanks for sharing those pictures! The downside of (alkaline) AAA cells is leakage, so be sure to check and consider if the unit could take primary lithium cells in the same form factor (mind the slightly higher open terminal voltage).

Re the calibration kit the ideal scenario wow be to buy it and have someone measure the Q. Because even parasitic inductance and capacitance in the measurement setup could affect it. So Ideally you'd mount the resistor in an aluminum (eg. Pomona) box with BNC connectors and then characterize the assembled unit with another calibrated unit that you have access to, for example through friends, work or uni. I'm guessing this is the original cal kit came with BNC connectors? Or was it just leaded resistors?

--- End quote ---

Thank You for information about Li-ion AAA cells. To be honest i only know about  Ni-MH existence in these form factor.

Yes, i want to replicate form factor. It comes with bnc in metal? box and Pelican case.

Unfortunately, I don't have anyone in souch places so commercial labs are only option.



--- Quote from: coromonadalix on May 23, 2023, 12:38:03 pm ---not being rude

you are going into a loop, the problem is:  you need to build and characterize the builded jig(s) .....  BUT you need a calibrated equipment to do so 

i think the best would be to: once calibrate your lcr,  build the test jigs and caracterize them,  and  keep them in safety  ...


my 2 cents

--- End quote ---

Yes, I mostly agree but there is a catch.

Few months ago, wanted to calibrate and adjust my Solartron 7081 dmm and I've called two major acredited labs.
In first one I got answer that they do not adjust meters others than Gw-instek and the other only cal as well!

So even if I will send the meter i will have to adjust it myself..


After scrolling the IET labs cal kit i found information about calibration: "Calibration Accuracy: These resistors are compared with a precision of better than 40 ppm at DC .  Measurement uncertainty of the nominal Q  values are within 100 ppm at 1 kHz.   All other R and Q values at other frequencies are mathematically calculated.  All calibrations are made at 23°C ± 1°C."

On the top of pdf is stated that this is Z-foil bulk metal resistor.

I have few of vishay VFR resistors, selected 1k and did a sweep from 10Hz to 2 MHz. I hoped it is linear and DC only measurement will be enough to achieve datasheet accuracy.

Unfortunately, this is not the case or after measurement (kelvin clips – neither frequency response nor accuracy friendly setup) I don't know whether it is setup’s fault or resistors. I do not have currently anything better on hand.

Measurement was taken with 10x average, compensated leads. Sweep takes over 2 hours.

In attachment You can find sweep graph and original measurements.
Today I will order bnc, boxes and one z-foil resistor, I will post the results.



--- Quote from: Vgkid on May 24, 2023, 01:32:45 am ---Thanks for tearing it down. That is a lot more complex  than I would have thought, certainly more than my gr1689. Interestingly enough , the reference resistors are the same, except your 24.9 is a custom EBG, while mine sticks with the S102.

--- End quote ---

This is strange, according to datasheet, gr1689 is more accurate 0,02% instead of 0,05% for 7600.

Cyclotron:
The thread is a bit old, but I just acquired this meter and was wondering if the thread creator was successful in gathering components to calibrate the unit.