Author Topic: Question regarding electronic loads  (Read 462 times)
Offline Pack34
« on: March 28, 2019, 02:30:20 am »
I have an application where I need to ramp a resistive load from 10-ohm to 550-ohm and back down as part of a calibration routine for some ADCs. I was thinking of hooking up a couple DMMs and an electronic load to measure the voltage and current of the system as the load ramps up and down.
Would this work?
Offline colorado.rob
Re: Question regarding electronic loads
« Reply #1 on: March 28, 2019, 02:40:35 am »
The Rigol DL3021A shows a resistive load range from 2.0 - 15k, so in theory it should work.  You probably don't need a DMM as it should show you current and voltage.  It is certainly worth trying.
Offline Pack34
« Reply #2 on: March 28, 2019, 03:39:26 am »
I'm looking at taking advantage of the 5ksps of the Keysight DMM I have, then wrap everything in an automated Python script.
Offline TurboTom
« Reply #3 on: March 28, 2019, 03:51:39 am »
An electronic load needs considerable time to settle in CR-Mode and also requires some "macroscopic" current to flow (depending on brand/model), otherwise it cannot control the resistance. Also please consider that on all loads (that I'm familiar with at least -- which are Rigol, Maynuo and similar ones, Applent, Chroma), there's a considerable capacitance present across their input terminals.
If you need the resistance for small signal calibration, you may consider a different option like those digital trimpots (i.e. AD5270 or the like), maybe equipped with a floating supply / digital isolators. Or maybe, you can include these devices directly in your design to facilitate calibration.
If you need a set of fixed resistances, you may be better off (and possibly more accurate) if you use a set of fixed resistors / preset trimmers and a bunch of reed relays, or maybe an R-2R cascade to form a primitive resistive D/A converter.
Cheers,
Thomas
« Last Edit: March 28, 2019, 03:55:27 am by TurboTom »
Offline Keysight DanielBogdanoff
  • ALL THE SCOPES!
    • Keysight Scopes YouTube channel
« Reply #4 on: March 28, 2019, 06:33:27 am »
This is a pretty common way to do things like this.
Here's an app note we just recently put together on eloads that might be useful. Direct PDF download:
http://bit.ly/eLoadAppNotePDF
It's mostly focused around our new load, but has some useful info in it.
The following users thanked this post: colorado.rob
Offline mairo
« Reply #5 on: March 28, 2019, 08:56:24 am »
Hi Daniel, do these new N6790 Series loads compatible with older N6705A/B series mainframes?
Pack34, sorry for jumping in your topic with a different question.
Offline LapTop006
« Reply #6 on: March 31, 2019, 12:56:57 am »
Given how recently the 6705B was discontinued you'd hope so, although given the final firmware for them still says Agilent I suspect it's unlikely.
Which is a shame. I have a 6705A, which could be excellent with just a few UI tweaks, until then it's useful for specialist things (to be fair, that was the goal), and the new load modules would go well in mine.