PA4TIM:
The IET makes a very good job out of the 5 wire technic. Only strange the bananas are added. You can easy exand the 5 wire conection by altering the crocs adapter. That s what I probably will do. For a DC uOhm meter the adapers like they are now, would not be the way to go. But this is a LCR meter.
The resistance added by the crocks themselve or the last cm of the tweezers is neglectable if you consider the accuraty and resolution. I did buy this for LC measurements. ESR is a non accurate parameter ( it is instable, every measurement will differ more as the R residu from the adapters)
And for mOhm I use my GR1608, Boonton or 7,5 digit multimeters. For high Resistance measurement
Guarding has an advantage too, but then shielding is often enough. I have a POhm meter that uses a guard driven sense input,

I made a 4 wire adapter yesterday that uses 4 bnc and optionl guard to connect to lcr meters, bridges and multimeters. Also made rg179 wires to my kelvin clips and rebuild two hirshman probes. The difference between the two is 4-8 mOhm. Easy to zero out and in most modes in the error zone ( seebeck, EMC, accuracy error, ect)

More important is the fact they use 4 wires to go to the crocs and tweezers and use a guard signal for the shield. The shield is called the 5th wire. This is important for capacitance meaurement because it minimises capacitance between the wires. By driving the shield with a guard signal there is no voltage difference between the sense line nd shield, hence no capacitance. Ths technique was common on lab grade bridges. They most times use 3 wire because reistance is not as important there. ( ESR measuring pF or aF is in the order of kOhms, so a few mOhm in the wire is noproblem, but upto 100 pF is. The guard driveh coax minimises this stray to under 1 pF after 1 meter of coax. On my GR1620 in a guarded farraday cup it s in the fF region.

GR has a few experimenters written about guarding ad multiwire Measurements.
BravoV:

--- Quote from: Wytnucls on March 11, 2013, 08:28:58 am ---At least, with this meter, you have the option to improve on the standard IET gear, by making your own Kelvin clips or tweezers, using the 4-wire blade connections.

--- End quote ---

Ahh.... finally, actually I've been waiting and expecting to hear these words straight from you.  ;D

Btw, thanks again for those extra measurement results with pics, well done and really like the presentation !  :-+
robrenz:

--- Quote from: PA4TIM on March 11, 2013, 09:53:33 am ---More important is the fact they use 4 wires to go to the crocs and tweezers and use a guard signal for the shield. The shield is called the 5th wire. This is important for capacitance meaurement because it minimises capacitance between the wires. By driving the shield with a guard signal there is no voltage difference between the sense line nd shield, hence no capacitance. Ths technique was common on lab grade bridges. They most times use 3 wire because reistance is not as important there. ( ESR measuring pF or aF is in the order of kOhms, so a few mOhm in the wire is noproblem, but upto 100 pF is. The guard driveh coax minimises this stray to under 1 pF after 1 meter of coax. On my GR1620 in a guarded farraday cup it s in the fF region.

--- End quote ---

I hesitate to differ with you on any topic but aren't you confusing a guard connection and a driven shield? The guard is the circuit common of a simple auto ballancing bridge.  (Covered very nicely in the Agilent impedance handbook starting in section 2-14)  I think the DE-5000 is only a simple guard.
PA4TIM:
There is much confusion about guarding because the term is used not very uniform. Not only between brands but also between models from brands themselve.

The configuration is also not obvious. The main function is to shield a sensitive input from "something"
In case of the PetaOhm meter it is a driven guard to make sure the current does not leak over the shield. If there is no potential difference there will flow no current. My Tek 130 LC meter also as a driven guard. But here to minimise stray capacitance.

Some meters have active guards and separate ground connections, but with an optional shorting bar between them. Like the Fluke 845 , 8500 and my HP sourcemeter.

It can be a passive form, i have an electrometer that has a sort of grounded guard grid between inputs and the rest. The HP 4260 also uses this. This adds capacitance,(no problem for DC) between traces to decouple, but also sometimes to make sure no current leaks from the source or higher voltage traces to the input.

I have a Fluke DC standard with a guard connector ( printed on the cabinet)  and that is just connected to ground, seperated from the low terminal by a capacitor, so this is a passive ground. An other Fluke calibrator has a guard, sense, and gnd ( 6 terminal connection) the ground is the cabinet, so fluke uses guard for two different things and both instruments are from the same line/area. I also have seen a resistor between low and guard, or low and ground.

In a capacitance meter you see passive and active guards ( guarding is in fact protecting, shielding is protecting too, so that makes it difficult to see what they mean by it.). Sometimes it is not a closed loop, so the shield is connected at one end only, sometimes it is closed ( for instance when shielding for EMC or RF ( to keep signal from entering or the otherway around, preventing radiation) .  It depends on the construction. You want an as less possible stray capacitance between souce and sense. If the sence impedance and signal is very low, a passive shield can sometimes be enough. This is very nice explained in several GR experimenters ( downloadable at IET website) but sometimes it needs to be driven. It is not allways very clear what they do. For instance in the GR1620 there are 3 ways to measure and two of those are 3 terminal. One uses two shielded GR connectors, this gives allmost no strays, the other is using three banana terminals, the third is using 2 of those bananas. You need to switch to those modes. The GR1608 and 1650 use two or three wire mode but you do not have to siwitch between modes.

I made some guarded circuits, sometimes passive by using a high Ohm resistor to couple the measured signal to the shield. No current flow, just an electric field to make potentials equial. But I used driven guard to by using an opamp to drive the shield. And things like traces, sometimes grounded, somtimes driven by a resistor or opamp.

The IET can be passive, or active, i have mot measured, but it does makes a difference in result.

Most important thing is that a manufacturer uses the way that fits the design and optimizes performance and I think it is save to say that if IET choses a way they do it for the right reason because they do know how to build good LCR meters. I think IET(GR), ESI and Wayne Kerr are the brands with the most experience in builing LCR meters, bridges and standards.

Just some things i collected about guarding, upto now i have not find a uniform explanation. Every brand seems to have its own definition about guarding. The Keithly book also talks about guarding.

Wytnucls:
The guard connection appears to be the same configuration on all four meters (DE-5000, Mastech MS5308, CEM DT-9935 and UNI-T UT612): A simple direct connection to the PCB ground plane.