Products > Test Equipment
Review of the Korad KEL103 programmable load.
<< < (11/11)
electr_peter:

--- Quote from: masterx81 on December 22, 2021, 10:42:54 am ---I want to add that the device can be used as current limiting device. In Constant Current Mode, the voltage drop across it without Current limiting acting is almost 0v.

--- End quote ---
DC loads can be used in more advanced situations, you described one of them. DC load in conjunction with PSU can test DC chargers (charger sees positive or negative current depending on voltage). Be mindful of allowable parameters in full setup.
Kaziq:

--- Quote from: sp9bsl on August 10, 2021, 10:35:15 am ---Hi,
just a silly question - what is adjustment resolution in CR mode?  I can't find it reviewed in any video. It is very useful in testing limit of SMPS.

--- End quote ---
Hi. It's been some long time, but I saw your question only now.
I have KEL102, and the adjustment resolution in CR depends on the range:

* 0 - 9.9999 R the resolution is 0.0001 R
* 10 - 99.999R it is 0.001 R
* 1000 - 9999.9R it is 0.1 RMaximum is 7500 R. Minimum is 0.05 R.

0.0001 R is crazy resolution. I tried to confirm it really works, using Keysight 34461A and Li-Ion battery (the most stable voltage supply I have right now), but the lowest I saw the difference was 0.01 R. Everything below 1 mA was noise. Perhaps with better setup it could be measured.

killingtime:
Digging up and old thread here, but it's the correct place for this question.

Anybody know what the ripple / noise is on the KEL103 while under load?

An important parameter for testing PSU's. It's quite large on some of the cheaper loads like the KP184.

Thanks.
Hydron:
This is with 100mA CC load (from a linear 5V supply). The ripple is from magnetic coupling from the (shielded) transformer to the shunt and/or the amplifier circuitry.
Ripple gets worse when either fan starts up (at 2A+, or when the temperature of the main heatsink rises too high).

Edit: has anyone else had issues with getting DHCP to stay on, or to connect with TestController? I think there might be a new revision and I'm not having any luck with those things, which ontop of the transformer/fan noise is a bit annoying.
Gediminas:

--- Quote from: masterx81 on December 22, 2021, 10:42:54 am ---I want to add that the device can be used as current limiting device. In Constant Current Mode, the voltage drop across it without Current limiting acting is almost 0v. I've had the need to test a device using it's Power supply and I've had the need to Current limit it. The response to sudden load changes is quite fast. Can do something that isn't designer for. Quite Happy with it. Excluding the user interface😅

--- End quote ---

That is good idea!
I have one more good use case, it can do short circuit (shift + short), in this mode load can be used as 30 Amp ammeter.
Navigation
Message Index
Previous page
There was an error while thanking
Thanking...

Go to full version