After doing some more testing I wanted to add more info to this thread, hopefully it can twig some ideas as to why I get the results I see
I have actually narrowed this down and realize that this thread probably belongs in the test and repair forum (if any admins agree and can move it feel free)..
I have attached a rough schematic/block diagram of the battery charger circuit I am testing.
The first test shows a standard x10 probe connected to TP2 and ground clip connected to GND (0V) and shows the expected +15VDC.
The second plot shows a measurement again of TP2 but this time using the TA041 differential probe (red probe connected to TP2 and Black connected to GND (0V). Again this shows the expected +15VDC no problem.
I repeat the exact same tests for TP3. Third plot attached for the standard probe which shows the expected 15VDC and fourth plot for the differential probe.
I understand that I most likely have some significant common-mode noise present on the secondary side of the charger.. this can come from a number of sources for example the Y-class capacitors C5 and C4, transformer etc.
I have measured this CM noise with a Fluke 179 RMS meter by placing the red probe on any point of the circuit and the black probe on the EARTH at the power outlet circuit (not DC ground). I get roughly around 90VACrms.
So I basically understand how there is 50Hz AC noise present what I can't understand is why the signal becomes rectified at TP3? I understand there are diodes in the circuit but I don't see an AC signal at TP2 least of all a rectified signal. Also as can been seen on plot 5 using the differential probe connected to OUT+ (red) and Out-(black) I can see the full un-rectified signal with DC offset, the RMS value is lower though.
Finally I understand that by connecting the standard probe clip to the circuit I am grounding out the CM Noise so therefore I don't see the AC signal for the TP3CH2 plot.. So therefore which probe is actually giving the correct plot? (note this power supply is a 2 core L+N insulated class product i.e. has no connection to EARTH on the mains socket)
Appreciate any help.. I can provide plenty of other data if it helps.