Sorry for the delay, but with holidays here in Germany it ended that I finally had a follow up call with Rigol Germany yesterday.
The situation is the following:
The DHO800/900 series has been designed by Rigol to mainly target educational purposes (aka they will not works seriously with the device

).
For some reason the chinese design team has willingly designed the Logic Analyser to be disabled for low timebase speeds.

I have no idea how many cents they save by doing so, but it is a hardware thing.

This might be interesting for Dave for a teardown/reverse engineering...
Therefore Logic Analyser [LA] in the RIGOL DHO 900 series DHO914 / DHO924 will NEVER work for timebases slower than 100ms/div and 20ms/div in roll mode. With the screen being 10 div wide the max time that you can sample for an event using the LA (either with or without analog channels) is only 1sec!!!

If your project is spinning up a motor or some other longer time thing you are limited to the remaining 4 analogue channels for debugging.

What a waste of chances considering that almost all resources (memory) are available...

I am in the happy position that I can always steal another scope from my colleagues bench, but
- I have to get used to another scope
- I cannot use it on the road (DHO8xx/9xx) will power from power bank
- My colleague will get mad at me if his scope is missing
And this shortcoming is
not documented in the datasheet, it is hidden in some subordinate clause in the 253page manual!
I am definately not happy with this and my personal advice for someone looking for a new scope would be to either stick with the DHO800 series or buy from some other brand.
Regarding the shortcomings in the software Rigol might provide a firmware update
if they consider some issue to be a bug.
I pointed out that you can configure only some channels of the LA to be active (16 channels fill the complete screen). But if you press the LA button twice (to hide and then to display again) ALL channels will be enabled again.
For me this is a bug, since it is annoying and also other Rigol scopes remember the LA configuration.
Rigol Germany has reported this issue to the dev team in China, but for them it "works as described in the requirement spec".

So yeah, I have filed some more things to them and it will be interesting to see what new FW-releases will come up with.
Rigol told me that they would be happy to get more feedback from customers. So find your nearest office here:
https://de.rigol.com/de/contact/infoIf there are many customers complaining maybe the dev team can be convinced to provide a bugfix.
Currently I have FW 00.01.04 installed and, yep there is still room for improvement...