| Rigol DP811 remote sensing |
| (1/1) |
| terrance:
I got my DP811 calibrated, and the voltage output and measurement is within spec now. but, when I toggle "sense", voltage output (externally measured) jumps between 2 values, is 1.8% lower when "sense off", but internal readings are the same. There's 0 load, and sense+ sense- is short to output +/- at the terminal. - set: 1.000 V, meas 0.982 V - set: 10.000 V, meas 9.819 V - set 40V. meas 39.284V |
| terrance:
>:( this drives me nuts, I did a "sense off" calibration, and now, it is 1.8% lower when "sense on" than "sense off" :wtf: And the "ADC cal" (page 2) fails, and "sense cal" has only 1 calibration point. I'll try a "sense" calibration in script. |