Products > Test Equipment
Rigol DP832 - Firmware list and bugs
McBryce:
If you got the "Installation Fails" message then you've probably put the wrong key in more than 3 times. As far as I can remember, the DP832 will block all license keys now for the next 24 hours, so you'll have to leave it running for a day before you can try the correct key.
McBryce.
Maverick71:
Thank You, I guess I tried this Codes for more than 20 times...
I'll report tomorrow.
Regards,
Maverick
kazam:
Hi,
I'm observing some strange behavior on my DP832, FW:1.14. This is the output given from the cal program:
Starting Voltage: connect the DMM VOLTAGE inputs to PSU channel 1
Press Enter
Running DAC-V for channel 1
DACV Ch:1 - Step:0 - Nom:0.200000V - DMM:0.018149V
DACV Ch:1 - Step:1 - Nom:0.500000V - DMM:0.318225V
DACV Ch:1 - Step:2 - Nom:1.200000V - DMM:1.016125V
DACV Ch:1 - Step:3 - Nom:2.000000V - DMM:1.811329V
DACV Ch:1 - Step:4 - Nom:3.200000V - DMM:3.001484V
DACV Ch:1 - Step:5 - Nom:4.100000V - DMM:3.891267V
DACV Ch:1 - Step:6 - Nom:5.200000V - DMM:4.980547V
DACV Ch:1 - Step:7 - Nom:6.900000V - DMM:6.671674V
DACV Ch:1 - Step:8 - Nom:7.500000V - DMM:7.270375V
DACV Ch:1 - Step:9 - Nom:8.700000V - DMM:8.460897V
DACV Ch:1 - Step:10 - Nom:10.100000V - DMM:9.856518V
DACV Ch:1 - Step:11 - Nom:11.800000V - DMM:11.544770V
DACV Ch:1 - Step:12 - Nom:12.600000V - DMM:12.339850V
DACV Ch:1 - Step:13 - Nom:13.500000V - DMM:13.231560V
DACV Ch:1 - Step:14 - Nom:15.000000V - DMM:14.717630V
DACV Ch:1 - Step:15 - Nom:15.800000V - DMM:15.510480V
DACV Ch:1 - Step:16 - Nom:16.500000V - DMM:16.206660V
DACV Ch:1 - Step:17 - Nom:17.300000V - DMM:17.002720V
DACV Ch:1 - Step:18 - Nom:18.500000V - DMM:18.194970V
DACV Ch:1 - Step:19 - Nom:19.100000V - DMM:18.794810V
DACV Ch:1 - Step:20 - Nom:19.900000V - DMM:19.586580V
DACV Ch:1 - Step:21 - Nom:20.200000V - DMM:19.884750V
DACV Ch:1 - Step:22 - Nom:20.800000V - DMM:20.479630V
DACV Ch:1 - Step:23 - Nom:21.800000V - DMM:21.468360V
DACV Ch:1 - Step:24 - Nom:22.400000V - DMM:22.062930V
DACV Ch:1 - Step:25 - Nom:22.700000V - DMM:22.363750V
DACV Ch:1 - Step:26 - Nom:23.900000V - DMM:23.561490V
DACV Ch:1 - Step:27 - Nom:24.300000V - DMM:23.956440V
DACV Ch:1 - Step:28 - Nom:25.700000V - DMM:25.352280V
DACV Ch:1 - Step:29 - Nom:26.900000V - DMM:26.542070V
DACV Ch:1 - Step:30 - Nom:27.900000V - DMM:27.532260V
DACV Ch:1 - Step:31 - Nom:28.500000V - DMM:28.126970V
DACV Ch:1 - Step:32 - Nom:28.900000V - DMM:28.525230V
DACV Ch:1 - Step:33 - Nom:29.800000V - DMM:29.426520V
DACV Ch:1 - Step:34 - Nom:30.200000V - DMM:29.827590V
DACV Ch:1 - Step:35 - Nom:32.000000V - DMM:31.614140V
Running ADC-V for channel 1
ADCV Ch:1 - Step:0 - Nom:0.000000V - DMM:0.000032V
ADCV Ch:1 - Step:1 - Nom:0.050000V - DMM:0.049575V
ADCV Ch:1 - Step:2 - Nom:0.100000V - DMM:0.099314V
ADCV Ch:1 - Step:3 - Nom:0.500000V - DMM:0.499981V
ADCV Ch:1 - Step:4 - Nom:1.000000V - DMM:0.999880V
ADCV Ch:1 - Step:5 - Nom:5.000000V - DMM:4.999513V
ADCV Ch:1 - Step:6 - Nom:10.000000V - DMM:9.999473V
ADCV Ch:1 - Step:7 - Nom:12.800000V - DMM:12.799610V
ADCV Ch:1 - Step:8 - Nom:20.000000V - DMM:20.000810V
ADCV Ch:1 - Step:9 - Nom:30.000000V - DMM:30.000650V
ADCV Ch:1 - Step:10 - Nom:32.000000V - DMM:32.002000V
The DAC values seem a little whacky and sure enough, after I calibrate the voltages are spot on but the readouts are way off. It will measure 2.002V when set to 2.000V but the DP832 will read back and display 2.961V.
I haven't written the cal data so restarting brings me back to decent values.
Insights?
/K
Macbeth:
Kazam, you don't say what cal program you are using or what your equipment is?
I had similar results when using my (modified) version of bsons python script. I put the issue down to the script not waiting long enough for the PSU to settle after changing its settings, but I didn't investigate as I gave up after encountering the DM3058E bugs (which Rigol have still not fixed).
I just chose not to commit the calibration when I had obviously faulty results.
kazam:
To clarify:
Using Matlab script posted earlier in thread. modified to use my Keithley DMM7510.
I checked the strings and values manually when debugging the script and it all checks out.
I have tried different wait times between setting the output of the DP832 and reading the voltage back. It makes very little difference and now I'm using 5 seconds. Still getting really weird data.
Yes, I'm not committing any cal data at the end so no harm done. Just annoying! :)
/K
--- Quote from: Macbeth on February 26, 2016, 11:00:18 pm ---Kazam, you don't say what cal program you are using or what your equipment is?
I had similar results when using my (modified) version of bsons python script. I put the issue down to the script not waiting long enough for the PSU to settle after changing its settings, but I didn't investigate as I gave up after encountering the DM3058E bugs (which Rigol have still not fixed).
I just chose not to commit the calibration when I had obviously faulty results.
--- End quote ---