Rigol DP832 - Firmware list and bugs
<< < (144/153) > >>
skander36:
You must consult Keysight Keythley product support page for your DMM . Here you will find software and documentation about SCPI. Also you may use software like Ultra Sigma from Rigol or  Command Expert from Keysight to send SCPI commands through USB or Telnet .
If DMM freeze either on USB then you must adapt commands to Keythely SCPI sintax.
Trident900fi:
Thanks for the hints ! I check and I let you know :-)
Trident900fi:
Just installed Keysight Connection Expert and command like *IDN?, *RST or *OPC? work fine.
Command like CONF:VOLT:DC AUTO make a error on the DMM

After a *RST command, DMM show only -.------ instead of numbers and stay like this.
With INIT it freeze !
bson:
As soon as you issue an SCPI command the instrument will switch into remote mode and the local panel is disabled.  There's usually a way to regain local control.

*RST disables continuous auto triggering in some instruments, so unless you start a measurement (MEASure), set a trigger, or auto trigger-and-wait (READ) the instrument won't take measurements or display a readout (readout being a calibration data corrected measurement).  My Keithley 2001 works the same; if I recall it also permits choosing what settings to load on reset, with the default being a complete reset with everything disabled (which is more practical for remote operation).
Trident900fi:

--- Quote from: bson on March 22, 2020, 10:06:35 pm ---As soon as you issue an SCPI command the instrument will switch into remote mode and the local panel is disabled.  There's usually a way to regain local control.

*RST disables continuous auto triggering in some instruments, so unless you start a measurement (MEASure), set a trigger, or auto trigger-and-wait (READ) the instrument won't take measurements or display a readout (readout being a calibration data corrected measurement).  My Keithley 2001 works the same; if I recall it also permits choosing what settings to load on reset, with the default being a complete reset with everything disabled (which is more practical for remote operation).

--- End quote ---

Many thanks for the informations ! I will try again  :)