Author Topic: Rigol DP832 Random Reboots  (Read 16209 times)

0 Members and 1 Guest are viewing this topic.

Offline cidcorpTopic starter

  • Supporter
  • ****
  • Posts: 105
  • Country: ca
Rigol DP832 Random Reboots
« on: February 10, 2014, 09:28:03 pm »

I sat back and watched the issues relating to the 'lack' of heat dissipation on one of the regulators used in the Rigol DP832 and
waited for it to get sorted out before I committed to purchasing a unit.  I ordered at the beginning of the new year and to the best
of my knowledge (haven't opened the unit) it's the latest board revision with the larger silver heat sink (I can see it through the grill).

My problem is in the past couple days I started using it longer periods of time and I've noticed a serious issue.  If left idle (I've only
noticed this with the channels OFF) eventually it reboots.  I saw it through the corner of my eye a couple days ago and I noticed it 10 minutes
ago when I was actually looking straight at it.  It was only on for about 20 minutes. Temperature listed on the unit (in the Utility Test/Cal area) hovers
around 34.42.

I'm glad I noticed the reboot because I was using it on a 3.3v 5" LCD display/controller board, and silly me I didn't set the 'Power On' option to LAST setup and it
was instead set to DEFAULT (CH3 5v)...things would have ended badly for the display and board. if I enabled the output without noticing the change.

I just wanted to know if anyone else has noticed this happening?  I'm on firmware 1.06, and yes I'll consider the later firmware if it seems like it's a fix.  Has anyone else
had issues with random reboots?  The dead giveaway is the sound of the FAN switching to FULL while it is rebooting...

I may have to check inside to see how the 'fix' was implemented on my unit.  For now I'm letting it idle and checking the temperature for when it reboots.

Chris
 

Offline taemun

  • Regular Contributor
  • *
  • Posts: 110
  • Country: au
Re: Rigol DP832 Random Reboots
« Reply #1 on: February 11, 2014, 02:36:30 am »
Quote
I'm glad I noticed the reboot because I was using it on a 3.3v 5" LCD display/controller board, and silly me I didn't set the 'Power On' option to LAST setup and itwas instead set to DEFAULT (CH3 5v)...things would have ended badly for the display and board. if I enabled the output without noticing the change.

Ouch. Are you able to set the power-on defaults to 0V so that this doesn't happen in future?
 

Offline cidcorpTopic starter

  • Supporter
  • ****
  • Posts: 105
  • Country: ca
Re: Rigol DP832 Random Reboots
« Reply #2 on: February 11, 2014, 04:32:02 am »

Well the good thing is that all channels are OFF by default when the power supply cycles.  I have now set it to LAST in the Power On setting, so I
at least feel a little better walking away from the bench for a while.  I've tested it and the settings are retained after it cycles.

After watching the temperature on the supply for close to an hour and half of use, I can't say that it's probably not the temperature of the device.  I do admit that it starts to
climb very very slowly over time...starting at 34 and climbing over a hour and a half to 35.25 and I suspect it will continue, but this doesn't alarm me because
I was using the supply during this time and I'm surprised that's all it climbed to.

I'm going to assume no one else has run into this problem... the few I've talked to are at the latest firmware and haven't seen this issue.  I maybe just my unit
as well.  I'll start documenting the reboots to keep a record.

Chris
 

Offline H.O

  • Frequent Contributor
  • **
  • Posts: 813
  • Country: se
Re: Rigol DP832 Random Reboots
« Reply #3 on: February 11, 2014, 07:03:18 am »
Hi,
I've got mine for ~6months, it's the old board with the small heatsink. I don't use it on a daily basis, sometimes not even weakly but I'm yet to see it reboot. I'm on 230VAC nominal, haven't measured actual.

Have contacted Rigol - twice - about getting the board replaced without even getting a response. I'm not TOO bothered though, I'll probabaly live with it, and fix it myself when the times comes.

But, your unit is with the new heatsink which is supposed to fix the issue. I can't believe it's STILL the same regulator overheating....
 

Offline uski

  • Frequent Contributor
  • **
  • Posts: 295
  • Country: us
Re: Rigol DP832 Random Reboots
« Reply #4 on: February 11, 2014, 03:02:08 pm »
Hi,

I've heard about people replacing that linear regulator with a small switch-mode power supply module out of eBay, like this one :

http://www.ebay.com/itm/DC-DC-Buck-Converter-Step-Down-Module-LM2596-Power-Supply-Output-1-23V-30V-/181227591312?pt=LH_DefaultDomain_0&hash=item2a32019690
DC-DC Buck Converter Step Down Module LM2596 Power Supply Output 1.23V-30V

$0.99 each, shipped.

(don't forget to adjust the output voltage before connecting it to the circuit)

uski
 

Offline Electro Fan

  • Super Contributor
  • ***
  • Posts: 3197
Re: Rigol DP832 Random Reboots
« Reply #5 on: February 11, 2014, 08:12:54 pm »
Hi,
I've got mine for ~6months, it's the old board with the small heatsink. I don't use it on a daily basis, sometimes not even weakly but I'm yet to see it reboot. I'm on 230VAC nominal, haven't measured actual.

Have contacted Rigol - twice - about getting the board replaced without even getting a response. I'm not TOO bothered though, I'll probabaly live with it, and fix it myself when the times comes.

But, your unit is with the new heatsink which is supposed to fix the issue. I can't believe it's STILL the same regulator overheating....

Maybe there is a possibility that the reboots other people are experiencing are caused by something other than the regulator and/or the overheating?
 

Offline cidcorpTopic starter

  • Supporter
  • ****
  • Posts: 105
  • Country: ca
Re: Rigol DP832 Random Reboots
« Reply #6 on: February 11, 2014, 08:33:30 pm »

I'm not going to tag it as a regulator problem.   I can't trust the temperature shown on the cal screen.  Now the power supply
is playing nice, and I haven't caught it rebooting since I posted. If it starts doing it again, I'll be directly checking the temperature
of that regulator with a probe.  And yes, I should upgrade the firmware to eliminate that as the issue as well.

It could be anything, and could be isolated to mine... so for now I'm happy that no one else out there has the same problem.

Chris
 

Offline skrap

  • Contributor
  • Posts: 29
Re: Rigol DP832 Random Reboots
« Reply #7 on: November 19, 2016, 09:53:26 am »
Did you figure out what the problem was? I have a brand new unit that does the same thing but even more often. Sometimes it will just be a short glitch on the screen and then it will hang until manual power cycling.
 

Offline Macbeth

  • Super Contributor
  • ***
  • Posts: 2571
  • Country: gb
Re: Rigol DP832 Random Reboots
« Reply #8 on: November 19, 2016, 10:10:19 pm »
Did you figure out what the problem was? I have a brand new unit that does the same thing but even more often. Sometimes it will just be a short glitch on the screen and then it will hang until manual power cycling.

I just had mine reboot even though it had not been used in a month and was just switched on for 10 minutes. I have never had it hang, but it certainly has rebooted a few times over the past few years for absolutely no reason.

In actual use it also has issues with powering certain boost converters like XL6009. It just goes loco.

It can not be trusted as unaided test equipment. I now use my old school HP 6632B for anything like that - 20V and 5A - it's a great boat anchor (or monitor plinth in my case)
 

Offline skrap

  • Contributor
  • Posts: 29
Re: Rigol DP832 Random Reboots
« Reply #9 on: November 20, 2016, 03:28:15 am »
Is that one with the updated PCB with a larger heatsink and mosfet or the older one with the heat problem?

For me it seems like I was just unlucky, now it reboots or hangs within 1 min of turning on so something is really wrong with this one. |O Would have been interesting to take it apart to figure out what's happening but it's probably not worth to gamble the warranty. I have an HP as well that I enjoy but I kind of like this DP832 unit even though all the quirks.
 

Offline _Wim_

  • Super Contributor
  • ***
  • Posts: 1517
  • Country: be
Re: Rigol DP832 Random Reboots
« Reply #10 on: November 20, 2016, 07:21:56 am »
I have 2 DP832 units that I use quite regularly (bought end 2014) So far I have never seen the reboot issue, but I haven’t watched for it specifically. They have never hang so far, that I am sure.
 

Online RoGeorge

  • Super Contributor
  • ***
  • Posts: 6185
  • Country: ro
Re: Rigol DP832 Random Reboots
« Reply #11 on: November 20, 2016, 10:02:35 am »
The only reset I saw on my DP832 was last month. Ch1 and Ch2 were in parallel (through a double diode, in order to charge a car battery). Both channels were set to 3A/18V each, while constantly logging the outputs values using SCPI commands over LAN (once a minute).

I was just increased the current from 0.3 to 3A (manually, on each CH1 and CH2 so the total generated power was under 85W), then after about 5 minutes at 80W, it reset under my eyes. After the reset, all the channels were in stop mode.

I couldn't reproduce the reset, so I let it charging the car battery over night at 2x3A and SCPI logging over LAN, and it didn't reset any more. Also, no other reset that I know of since I bought the DP832 about 2 years or so, but it was used only rarely, and usually for very small loads (1-10W for all 3 channels added).

I suspect the reset has something to do with the fact that SCPI commands were sent (for data logging) while manually changing the current from 0.3 to 3A. I am saying this because the last DP832 logged (in the PC) was at 0.3A, but the reset happened a few minutes later after the current was increased from 0.3 to 3A, so I should have seen a few records in the log at 3A, but there are no records logged at 3A. These are the last 3 records in the PC log:

Code: [Select]
YYYY-MM-DD,HH:MM:SS,CH1(V),CH1(A),CH1(W),CH2(V),CH2(A),CH2(W),CH3(V),CH3(A),CH3(W)
...
2016-10-30,01:19:25,12.5654,0.3993,5.017,12.5716,0.3947,4.962,0.0000,0.0000,0.000
2016-10-30,01:20:25,12.5648,0.3993,5.017,12.5711,0.3947,4.962,0.0000,0.0000,0.000
2016-10-30,01:21:25,12.5644,0.3993,5.017,12.5706,0.3947,4.962,0.0000,0.0000,0.000

During all this time, a DMM was also logging the battery voltage and current. The multimeter logged a few minutes between the manual current increase and the reset, but these minutes are missing from the DP832 log, so I guess it was like this:
- DP832 generating under 10W while logging the data into PC
- manually change the current (from the DP832 front panel) for CH1 and CH2 from 0.3A to 3A
- the software hang, because I have no 3A records
- after a couple of minutes, the DP832 reset. During these minutes, the DP832 was generating less then 100W, and the current was stable, and at the expected (as shown in the DMM logs).

I don't know where to look for the firmware version details. The 'Utility' -> 'System' shows 'Digital version: 00.01.11'. The hardware is the version with a big radiator (the radiator can be seen through the air holes).

This is what the DMM was logging. The blue cross is marking the DP832 reset. After the reset, all the DP832 were off. The charts shows the same values before and after reset because I manually turned on the DP832 channels after the reset. The various steps in the charts are because the current or voltage were manually adjusted in the DP832 a few times during the whole battery charging.

Offline Pinkus

  • Frequent Contributor
  • **
  • Posts: 773
Re: Rigol DP832 Random Reboots
« Reply #12 on: November 20, 2016, 10:49:38 am »
Hi,

I've heard about people replacing that linear regulator with a small switch-mode power supply module out of eBay, like this one :

http://www.ebay.com/itm/DC-DC-Buck-Converter-Step-Down-Module-LM2596-Power-Supply-Output-1-23V-30V-/181227591312?pt=LH_DefaultDomain_0&hash=item2a32019690
DC-DC Buck Converter Step Down Module LM2596 Power Supply Output 1.23V-30V

$0.99 each, shipped.

(don't forget to adjust the output voltage before connecting it to the circuit)

uski
Seriously? Using a crappy Chinese $1 (including shipping) stepdown module (=made of 100% fake/dubious quality products*) for an expensive lab equipment  :scared:? If this fails (and chances are high it will at some day), it will fry your DP832.

* try to calculate. subtract shipping, packing, margin etc. you will find, they are producing this for a BOM of maybe $0.3. Now try to find LM2596, 2x 36V caps, 3A inductivity, 3A schottky, precision trimmer for $0.3 .... no way, even if you would go an buy 100K or 1M of them.


 

Offline sdouble

  • Frequent Contributor
  • **
  • Posts: 275
  • Country: fr
Re: Rigol DP832 Random Reboots
« Reply #13 on: November 20, 2016, 08:08:54 pm »
I suffer from that issue too.
I reported that here :
https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/msg1071713/#msg1071713
I use them with the 230V 50 Hz main. Are you also using it in europe ?
 

Offline skrap

  • Contributor
  • Posts: 29
Re: Rigol DP832 Random Reboots
« Reply #14 on: November 21, 2016, 04:50:07 am »
Europe here, 230 VAC / 50Hz with what seems to be the latest hardware revision. At least the large heatsink but I can't confirm the new mosfets or not, firmware is the latest version and calibration was done in June. The supply works really well and is within spec. when it doesn't hang or reboot but this typically happens within a minute now no matter if the outputs are on or off. Since it's very repeatable I guess something on the digital side is overheating for some reason.
 

Offline skrap

  • Contributor
  • Posts: 29
Re: Rigol DP832 Random Reboots
« Reply #15 on: November 21, 2016, 11:41:03 pm »
Got a new one and this one is flawless, at least so far ;-) All the design quirks are still there of course but I can't really say I mind them much at all actually. Latest hw and software revisions really makes this a nice powersupply indeed.

At least until the dreaded reboot rears its ugly head once again... but hopefully no more.
 

Offline sdouble

  • Frequent Contributor
  • **
  • Posts: 275
  • Country: fr
Re: Rigol DP832 Random Reboots
« Reply #16 on: November 22, 2016, 07:34:16 pm »
Europe here, 230 VAC / 50Hz with what seems to be the latest hardware revision. At least the large heatsink but I can't confirm the new mosfets or not, firmware is the latest version and calibration was done in June. The supply works really well and is within spec. when it doesn't hang or reboot but this typically happens within a minute now no matter if the outputs are on or off. Since it's very repeatable I guess something on the digital side is overheating for some reason.
what is the latest revision ?
 

Offline skrap

  • Contributor
  • Posts: 29
Re: Rigol DP832 Random Reboots
« Reply #17 on: November 23, 2016, 03:01:20 am »
It's DP8C 2.20 I beleive, I'm not sure about the number but it's the one with new heatsink and larger mosfet.
« Last Edit: November 23, 2016, 03:11:31 am by skrap »
 

Offline envisionelec

  • Frequent Contributor
  • **
  • Posts: 286
  • Country: us
Re: Rigol DP832 Random Reboots
« Reply #18 on: June 01, 2017, 03:10:09 pm »
I just got a message from a friend who bought my 2016 (post regulator update) DP832 a couple months ago. He says it get stuck in a reboot loop especially when cold. It warms up then stops resetting and begins to work. Anyone see this?
 

Online RoGeorge

  • Super Contributor
  • ***
  • Posts: 6185
  • Country: ro
Re: Rigol DP832 Random Reboots
« Reply #19 on: June 02, 2017, 08:23:07 am »
If any electronics are doing strange things only when cold, then check for bad capacitors with high ESR.

Offline taydin

  • Frequent Contributor
  • **
  • Posts: 520
  • Country: tr
Re: Rigol DP832 Random Reboots
« Reply #20 on: April 09, 2018, 02:02:09 pm »
Just saw the reboot on my DP832A. I have this unit for several years, and I think I saw this spontaneous reboot only two times, including this latest reboot. Firmware version is 1.13.00.1

When this happened, only CH1 was on, set at 12V and supplying around 150 mA current. Temperature is 30 degrees celsius right now, after reboot.

I don't know what the status is with the heatsink, I will check this.
Real programmers use machine code!

My hobby projects http://mekatronik.org/forum
 

Offline Jeroen3

  • Super Contributor
  • ***
  • Posts: 4078
  • Country: nl
  • Embedded Engineer
    • jeroen3.nl
Re: Rigol DP832 Random Reboots
« Reply #21 on: April 22, 2018, 06:57:55 pm »
I was using my DP832 to charge a battery, came back a few hours later and it turned off channel 1.
OVP or OCP were disabled... So I guess it did a reboot.

Do we know the cause, or is there a fix? A power supply that reboots isn't very reliable imho.
 

Offline Smokey

  • Super Contributor
  • ***
  • Posts: 2572
  • Country: us
  • Not An Expert
Re: Rigol DP832 Random Reboots
« Reply #22 on: November 29, 2018, 05:28:07 am »
Bummer.  My DP832 just starting this whole reboot thing.  Happens intermittently now about every 20-45 seconds.   I've had it for a couple years with no problems until now.  That sucks.

Just as another data point, I outfitted all the technician benches at my last place with these supplies.  Probably around 10 in total, doing daily work in a production environment.  Out of the original order, one unit had this reboot issue out of the box.   We sent it back for a replacement and had zero problems with any of the supplies since.

:EDIT:  Its not one of original batch with questionable linear regulator.
 

Offline Smokey

  • Super Contributor
  • ***
  • Posts: 2572
  • Country: us
  • Not An Expert
Re: Rigol DP832 Random Reboots
« Reply #23 on: November 29, 2018, 06:12:36 am »
Well shit.  That went downhill fast.  I was just about to post that I have the 1.14 firmware and it got worse.

Within the day it went from rebooting once every hour, to every minute, and now it's mostly in an infinite reboot loop where as soon as it comes up it reboots.  I can hear the fan spin up sometimes, but not every time.

Of course this happens when I'm in the middle of an actual project.  Fun. 
 

Offline InductorbackEMF

  • Regular Contributor
  • *
  • Posts: 73
  • Country: gb
  • Repair/Rebuilding.
Re: Rigol DP832 Random Reboots
« Reply #24 on: December 02, 2018, 10:35:45 am »
You know i had it reboot 2 times as well,i almost thought it was a power cut!
good thing i was just testing some resistors or that would of ended badly as well hehe.

Alex.
Want a cup of TEA? (just one more meter please..)
Alex.
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23018
  • Country: gb
Re: Rigol DP832 Random Reboots
« Reply #25 on: December 02, 2018, 11:08:51 am »
Slightly glad I didn’t buy one now. Historically speaking I had an issue with some Farnell supplies at university years ago. They were weak so if there was a power glitch the output cap would discharge across the pass transistor and blow it up. On powerup full unregulated DC would hit the DUT. To work around this we used a fat 5W Zener across and a fuse in series with the power supply.

Shouldn’t have to worry about this in 2018  :palm:
 

Offline TurboTom

  • Super Contributor
  • ***
  • Posts: 1389
  • Country: de
Re: Rigol DP832 Random Reboots
« Reply #26 on: December 02, 2018, 02:51:28 pm »
I've got some issues with a DP811 (20V/10A or 40V/5A single output unit) also rebooting all by itself once in a while. It's working stand-alone, no LAN or USB connection so the behaviour shouldn't be induced externally. I recently updated it to the 01.16 beta firmware but it didn't make any difference regarding the reboot issue. Yet, it affected the 10s digit of the preset current display as not being able to properly display "0" -- the leftmost two vertical segments are clipped (btw, stupid idea to imitate a seven segment display on a graphic screen anyway...). So I'll probably return back to F/W 01.14 on that instrument. I'm curious if anyone will be able to find out what's actually causing the reboots on the DP800 series. A machine that's doing this all the time should be a good condidate for a fix. My one so far is only rebooting once every odd day or so...
« Last Edit: December 03, 2018, 06:52:03 am by TurboTom »
 

Offline Smokey

  • Super Contributor
  • ***
  • Posts: 2572
  • Country: us
  • Not An Expert
Re: Rigol DP832 Random Reboots
« Reply #27 on: December 03, 2018, 05:18:56 am »
Since I bought the supply from the rigol clearance, it only had a 90 day warranty.  I would still be under the 3 year mfg warranty though.  They want $250usd eval to fix it.  I pushed back a bit since this reboot thing appears to be some sort of design problem. 
 

Offline Commodore8888

  • Contributor
  • Posts: 32
  • Country: us
Re: Rigol DP832 Random Reboots
« Reply #28 on: October 20, 2019, 08:26:54 pm »
Bit of a necro here, but I had the 832 out for some "upgrades" (namely changing to an 832A via the SCPI command), and saw this thread while digging around.

Figured I should add the data point that I too had an 832 from circa mid 2016 (post regulator "fix") that encountered random reboots. It had been sitting idle for about 5mo on a shelf. I was lucky to get even 20s out of it, sometimes it would just sit in a boot loop.

FWIW, it did seem to be somewhat temperature dependent. I sent the unit in for warranty repair
in late 2017 and it came back fine (knock on wood). All the service sheet said was "replaced control board".
Mike D
 

Offline Smokey

  • Super Contributor
  • ***
  • Posts: 2572
  • Country: us
  • Not An Expert
Re: Rigol DP832 Random Reboots
« Reply #29 on: October 21, 2019, 12:23:04 am »
Ya, I should have followed up.  They ended up giving me a little break on the repair cost.

Before I sent my unit in it was pretty much in a constant reboot loop on my bench.  They claimed that they could not reproduce the problem, which is very odd considering it was 100% nonfunctional when I sent it.  Only thing I could figure was shipping knocked something back into place maybe?
Anyway I asked if they new exactly what the problem was with these "reboot" failures and he didn't tell me anything other than when they get one they just replace the whole control board, like Commodore8888 said.  All of my options were reset, so that does make sense.

It's been running fine since then.
 

Offline T2

  • Newbie
  • Posts: 7
  • Country: ch
Re: Rigol DP832 Random Reboots
« Reply #30 on: March 03, 2020, 03:12:49 pm »
Hi there,

I know this thread is quite old.

Myself, I'm looking at a dp832 that has "developed" this reboot issue over time.
My Rigol has the v2.10 (2013-03-26) top board (with the large heat sink).

- the digital load seems quite "dynamic", the current comsumption is not stable but "jumps around"
- using an external 5V power source for the digital supply just intensivies the problem (Minutes instead of hours/days)
  -> longer cables, higher impedance

- Measuring the voltage with an oscilloscope (AC coupled) shows sporadic, rather high spikes/bursts
- >1.2Vpp (measured with the built-in 5V supply) !!

This leads me to believe that instabilities/noise on the 5V rail may be culpable for the spontaneous reset/reboots fo the DP832.
My next palnned steps are to dismantle the logic board (front board) and look there closer.
- Caps going bad?
- Insufficient decoupling close to the MCU or reset logic?


/Thomas
942946-0
« Last Edit: March 03, 2020, 03:15:13 pm by T2 »
 

Offline T2

  • Newbie
  • Posts: 7
  • Country: ch
Re: Rigol DP832 Random Reboots
« Reply #31 on: March 03, 2020, 09:46:00 pm »
Dug further down. Some of my previsouly observed 5V spikes may have been coupled in by my measurement setup.

The digital front board seems to be really a finicky thingie. I came across an 1117 linear regulator (5V in, 3.8V out, labeled "Battery") that basically 1:1 passed the input ripple from the 5V line to the output. With a maximum drop-out voltage of (according to datasheet) 1.1 to 1.3V, this seems a bit on the close side (5V supply has around 5.2V, so deltaV is 1.4V). There are 3 ceramic output caps present (see schematic fragment) but I don't trust them.

The other supply voltages (3.3V, 1.8V, etc) look more or less OK but have quite some digital noise on them.

I added an input (22u, yeah I rounded up) and an output cap (100u) to the 3.8V regulator as the datasheet sugegsts. The output voltage then became rather stable, not passing on any ripple from input. Check the pictures for my hack.

I also noted that the CPU gets rather hot (you wouldn't let your finger there for long time), so I put a small heatsink on.

Now starts the waiting part...


943052-0

943056-1

EDIT: Still rebooting. Need to look further.
« Last Edit: March 04, 2020, 08:32:09 am by T2 »
 
The following users thanked this post: Smokey, thm_w, baldengineer, MegaVolt

Offline T2

  • Newbie
  • Posts: 7
  • Country: ch
Re: Rigol DP832 Random Reboots (FIXED!)
« Reply #32 on: March 05, 2020, 06:26:56 pm »
Hi there,

I think I finally fixed it. My DP832 is now running without reboot for more than 24h.

If you have a Rigol DP832 with a DigitalBoard V02.00 2012-12-07 (front-side board with MCU)
and your Rigol is not doing sporadic reboots you may just be lucky.

TopBoard: v2.10 (large heat-sink for 5V regulator)
DigitalBoard V02.00

My DP832 started to develop the sporadic reboots over time, from once a month to about once a
day to multiple times a day by now. Let's face it - unusable as a bench supply. And yes, my DP832
is running 24x7 - but that's something you'd expect from a bench-supply.
Some days ago I dismantled my Rigol and started poking around on the digital board.

To summarize, aside from the linear 5V regulator/toaster on the top board that we all know about,
I identified 2 issues while checking the voltage rails on the digital front board.


The "Battery" rail:
The AC noise from the 5V rail was passed straight through to the output of the regulator (1117 ADJ).
I added input and output caps (although there are some ceramic caps on the output side, but I
don't trust them). Also the deltaV (Vin-Vout) was very close to the dropout voltage. My guess is that
this is the main reason for the ripple being passed straight through (the regulator was not regulating
anymore).

Studying the iMX28 (Application processor / MCU) datasheet a bit, shows, that this beast
is designed to be powered from two power sources. A 5V source as well as a Li-ion battery
source (3.1 ... 4.2V). The MCU automatically switches between those sources.


The "VDD4P2" rail:
This is an intermediate power rail (4.2V) from the MCUs integrated switching supply. I did
measure a *sh%@tload* of switching ripple (>2Vpp) on this rail. According to the reference
schematic for the iMX28, there *must* be a handful of decoupling caps installed on this rail.
However, by just quickly probing around I could not find any serious caps (just one tiny 0402 cap)
on the Rigol board (I don't say there are no caps, just say that I did not find any - hard to tell
for sure  without schematic). Luckily this rail is accessible from the back side of the board, no
need to disassemble the front.
I added some caps (ceramic and electrolyte) to this rail and the switching noise got *way* better.

I'm still measuring up to 400mVpp switching noise, but stability has improved massively.
Before this modification, with the voltage reduced to 4.9V on the 5V rail, I got multiple reboots
per minute. After adding the caps it just kept running. No reboots observed. :-)

So my guess is: fix #2 is actually the one that cures the spontaneous reboots. Problem #1
probably has no direct impact on that (depends maybe when/if the MCU actually takes
the power feed from the "Battery" rail - if at all).


A lower voltage an the 5V rail causes the reboots to happen faster / more frequent.
My guess is: lower voltage -> higher currents (switching supply in the iMX28) -> more switching ripple on 4P2 rail
-> at some point the MCU resets (likely the built-in reset circuit or brown-out detection triggers)

This could even explain why it happens more frequent on devices with the small heat-sink (old top board).
If the LM317 starts to cut back due to thermal conditions -> lower voltage -> higher probability for reboot

Note: This VDD4P2 rail issue might me systematic! A cross check with a properly working (no reboots)
DP832 of a friend (same DigitalBoard rev) did show massive ripple of close to 2Vpp on this rail as well.



To summarize my modifications:
1) I modified the 1117 "Battery" voltage regulator circuit to deliver 3.5V
(instead of the 3.8V) to get some voltage margin (dropout voltage 1.25V!)
and avoid issues there. I also added input (10uF) and output (100uF) caps.
944160-0]

To adjust the voltage, I put a 4k7 resistor in parallel to the existing 680E
(going to ground). Yeah, ugly...
944156-1

2) Add a handful of decoupling caps to the VDD4P2 rail
10nF // 100nF // 1uF // 47uF
The VDD4P2 is on the cathode side of the fat shottky diode on the back side of the board.
944176-2
I scraped away some solder-mask on a ground-plane and botched in a handful of caps.
944164-3
944168-4

Fragment of the reference schematics:
944172-5

I'm quite sure that modification 2 alone will fix this issue, but I didn't bother to undo my other modification.

I hope this will help some Rigol DP832 owners to get their device working properly again!

/Thomas

Update: Running for one week now, without issues.
« Last Edit: March 11, 2020, 06:03:24 pm by T2 »
 

Offline _Wim_

  • Super Contributor
  • ***
  • Posts: 1517
  • Country: be
Re: Rigol DP832 Random Reboots (FIXED!)
« Reply #33 on: March 05, 2020, 07:45:49 pm »
Hi there,

I think I finally fixed it. My DP832 is now running without reboot for more than 24h.

Great work! I do not experience the reboots (yet?) of both of my DP832s, but it is good to know there is a solution when it happens. How old is your DP832? I wonder if this problem is related to a certain batch of instruments...
 
The following users thanked this post: Dazed+Confused

Offline Smokey

  • Super Contributor
  • ***
  • Posts: 2572
  • Country: us
  • Not An Expert
Re: Rigol DP832 Random Reboots
« Reply #34 on: March 06, 2020, 06:35:03 am »
Great job! 

Who is going to be the brave soul that tries to simulate the conditions on a perfectly functional supply to see if it starts rebooting?
 

Offline T2

  • Newbie
  • Posts: 7
  • Country: ch
Re: Rigol DP832 Random Reboots (FIXED!)
« Reply #35 on: March 06, 2020, 04:53:35 pm »
Hi there,

I think I finally fixed it. My DP832 is now running without reboot for more than 24h.

Great work! I do not experience the reboots (yet?) of both of my DP832s, but it is good to know there is a solution when it happens. How old is your DP832? I wonder if this problem is related to a certain batch of instruments...

I bought my DP832 in Sept 2014. It already had the Rev 2.1 TobBoard with the large heatsink. FrontBoard is V2.0 (with a 2012-12 revision date).

By the way, still running without issue (48h now).
« Last Edit: March 06, 2020, 04:55:29 pm by T2 »
 

Offline T2

  • Newbie
  • Posts: 7
  • Country: ch
Re: Rigol DP832 Random Reboots
« Reply #36 on: May 03, 2020, 09:00:19 am »
Just a heads up: My DP832 is not running for 2 months with this fix and has not experienced any issue (= reboot) since.
This modification definitively fixes the reboot issue.

/Thomas
 
The following users thanked this post: Smokey, Macbeth, sequoia

Offline cgz2001

  • Contributor
  • Posts: 15
  • Country: us
Re: Rigol DP832 Random Reboots
« Reply #37 on: August 15, 2020, 08:20:57 pm »
All,

I came into possession of one of these units this week which has the constant reboot symptom described above.  The unit never manages to stay up longer than a few seconds, if that.  I intend on performing a modification to 4p2 V rail on the unit as T2 has described.  For reference this unit has the silver heatsink.  Screenshots below are documenting the performance on the LM317T regulator input and output as well as 4.2 V rail.

Once I perform the modification I'll advise on the performance, hopefully with screenshots.
 

Offline frank.van.dinteren@home.n

  • Newbie
  • Posts: 1
  • Country: nl
Re: Rigol DP832 Random Reboots
« Reply #38 on: August 20, 2020, 10:42:26 am »
This fixed the reboot (resetting the output to off, not a whole machine reboot)  issue for me as wel. Have mine DP832A  since mid-2014. Started having random reboots a few years ago (once every few days). It would never stay up for a whole weekend and now it's already is running for 2 weeks.

Thanks.
 
 

Offline cgz2001

  • Contributor
  • Posts: 15
  • Country: us
Re: Rigol DP832 Random Reboots
« Reply #39 on: August 26, 2020, 11:49:42 am »
All,

Adding ceramic class 1 and class 2 capacitors as alluded to above seems to have solved the problem on my unit.  I will advise if it eventually fails.
 

Offline chrisrossx

  • Newbie
  • Posts: 1
  • Country: us
Re: Rigol DP832 Random Reboots
« Reply #40 on: September 02, 2021, 09:07:00 pm »
My DP320 (Board Rev 02.20) started randomly rebooting a few months ago, once every 10 minutes or so... Today, it was rebooting every 5 seconds or so.

I took the cover off, to poke around, and I couldn't replicate the problem.  I buttoned everything back up, and it's been on for over an hour.  The only thing I can think of is that I did stress the wires on the top board trying to take it apart.  However, I did not try and reseat any of the connectors...

Not a solution, but it did stop my issue...
 

Offline blurpy

  • Regular Contributor
  • *
  • Posts: 232
  • Country: no
Re: Rigol DP832 Random Reboots
« Reply #41 on: September 03, 2021, 01:32:41 pm »
Jerry Walker has a series of videos where he troubleshoots and fixes his rebooting DP832:



I found it very interesting, even though I don't have that device myself.
 

Offline T2

  • Newbie
  • Posts: 7
  • Country: ch
Re: Rigol DP832 Random Reboots
« Reply #42 on: October 17, 2021, 03:32:23 pm »
Part 3 of this video series points out exacly what my conclision was, just *much* better explained than I could.
 

Offline TurboTom

  • Super Contributor
  • ***
  • Posts: 1389
  • Country: de
Re: Rigol DP832 Random Reboots
« Reply #43 on: October 17, 2021, 11:49:54 pm »
I just installed a 220µF/25V Rubycon ZL low impedance electrolytic across the troublesome 4.2V rail (because I had those handy) and verified that the noise on that rail dropped by at least a factor of five (from close to 2Vpp to  ~400mVpp). I used existing pads for this and tacked the caps with a drop of glue to the PCB. No reboot problems after that anymore...
 

Offline baldengineer

  • Newbie
  • Posts: 7
  • Country: us
  • James
    • Personal Blog
Re: Rigol DP832 Random Reboots
« Reply #44 on: November 15, 2021, 10:33:17 pm »
I did both the LM1117 voltage mod and added additional decoupling caps. In my case, I used a combination of polymers and MLCCs. After 48 hours, my DP830 has stopped the reboots.

Thanks for taking the time to diagnose and document your fix.

Not my favorite soldering job, but it works. :)
Capacitor Expert by Day, Enginerd at Night  ||  Personal blog:  www.baldengineer.com  ||  Electronics Videos for Beginners:  www.addohms.com
 

Offline hcoverlambda

  • Supporter
  • ****
  • Posts: 10
  • Country: us
    • Blog
Re: Rigol DP832 Random Reboots
« Reply #45 on: August 03, 2023, 02:51:41 pm »
I only added the 47µF cap and that fixed it for me. Was restarting a few times p/m. Thx to everyone who dug into this!
 

Offline BTBlueSkies

  • Newbie
  • Posts: 2
  • Country: us
Re: Rigol DP832 Random Reboots
« Reply #46 on: August 23, 2023, 02:33:29 am »
Ha! you guys should rule the world, I would vote for you.

Thanks so much for this thread and the info, it worked for me. Here is a post showing my old man solution of hot gluing the caps in a stack and then, after gluing em down so they don't move, I could solder them into place quite easily since they didn't dance around any more.

Thanks again.
 
The following users thanked this post: PeDre, tv84


Share me