Rigol DS1000Z series buglist continued (latest: 00.04.04.04.03, 2019-05-30)
rob040:
--- Quote from: ankerwolf on March 31, 2018, 08:03:40 pm ---Simply rename the .01.zip to .z01 and open the .zip with any packer (WinRAR, 7zip, ...).
--- End quote ---
Yessss, that does the trick. Thanks!
Payne:
New bug found?
Hi all,
When I do the following, something weird happens with the vertical position of the signal line:
Set VerticalRef = Center (Utility, System)
No probes connected
Set vertical sensitivity to 5V
Center the vertical line to 0V
Center the trigger point to 0V
Set vertical to -5V
Set trigger point to -5V
Now turn the vertical sensitivity knob to 2V, 1V, 500mV
When you turn to 200mV (and even lower), the trace will appear on the screen again (but it should have been close to the floor in your room!)
I can reproduce this error also with other settings for the Y-axes sensitivity and other channels
Rigol DS1054z
SW: 00.04.04.SP3
Board: 0.1.4
frozenfrogz:
I can not reproduce that on my scope. The trace leaves the screen (yellow line glued to the bottom) in all Vscale settings.
Edit: Same board revision by the way, Rigloled DSER.
Boot: 0.0.1.4
FW: 0.2.3.11
Payne:
Try to set the probe for CH1 at 1x, then it should be there!
Fungus:
--- Quote from: Payne on April 09, 2018, 06:56:40 am ---When you turn to 200mV (and even lower), the trace will appear on the screen again (but it should have been close to the floor in your room!)
--- End quote ---
You hear that click as you go to 200mV? That's a clue that the 'scope switched ranges there.
If the line is in the wrong place in those ranges it means you need to run calibration on your scope.