bitseeker:
Sounds plausible to me.
BravoV:
Helped friend upgrade his DS1054Z from 04.03.02.03 to the latest 04.04.03.02, so far no problem in the upgrading process and also the self cal after that. Also all DSER options are still active in this new firmware.

Spotted minor screen's change, not as many as my previous DS1104Z-S as it was a big jump, but this time is just the right panel buttons, watch closely the minor shift at right side at the attached GIF animation.
BravoV:

--- Quote from: ProBang2 on April 21, 2017, 02:01:25 pm ---How to get in the "expanded mode"?
Fast pressing the hard buttons [Menu] [Menu] [Force] [Menu]
Then pressing the hard button [Utility]

After that:
Press the soft buttons <System> and <System Info> gives the expanded system info.
Or, press the soft button <Self-Cal> and the scope shows the usual two soft buttons "Start" and "Exit"
and additional two soft buttons: "LFCal" and "Output".

With a DS1054Z both functions are failing.                             :-//
Perhaps the first is associated to an inbuild logic analyzer, the second to the function generator of a 1054S?

--- End quote ---

The new extra "Output" button does some extra magic.  ;)

Here are the steps :

* While in the "Expanded" mode, insert an USB flash drive, and of course assuming its trouble free like you can save files without problem.
* Go to the Self-Cal menu, and then press the "Output" button
* It will save a new file at root directory in the USB flash named "CalData.txt".  (PS : It will not notify anything on screen when it saves that file, but it should be fast, say in few seconds)
Now, the fun of decoding that CalData.txt file is started ...  :-DD


For those who still confused, attached below pictures of two versions of Self-Cal screens, the normal one and in the "Expanded" mode.

ted572:
The following procedure can be used to Sanitize Memory in the DS1000Z Series Oscilloscopes. This is similar to the procedure released by Rigol (Titled: Sanitization of memory with the DS1000Z series of oscilloscopes) with some additional details.

Edit Added -> Re-Calibration Requirement: Sanitization of the instrument's memory will of course erase the stored Self-Calibration values.  Therefore it is important to perform Self-Calibration following Sanitization.
Gary.M:
Just did the update. I was a little cautious as my hardware and boot version seemed to be the troubled ones. However all went well. I did the full memory sanitise first.