Author Topic: Rigol DS1054Z calibration & adjustment  (Read 9266 times)
« on: May 22, 2016, 07:39:40 am »
Out of curiosity, I was doing some searching here and on the web at large regarding calibration/adjustment for the 1054 (i.e., not referring to the built-in auto-cal), but found nothing fruitful thus far. It may or may not need an adjustment in the coming years, but if it does, then what?
I don't think the usual calibration services handle these. I suppose Rigol would be happy to take your money. However, I only need it to be accurate relative to other instruments in my lab (i.e., adjusted to an in-house reference), not NIST. In other words, DIY calibration. :-/O
Given its price class, it might be considered relatively disposable. :o Is that its ultimate fate? I suppose it could go on living for qualitative see-if-something's-there use, should it actually drift so far from where it should be.
Mine is still under warranty, so I haven't taken it apart, yet. Just thinking aloud.  :blah:
Re: Rigol DS1054Z calibration & adjustment
« Reply #1 on: May 22, 2016, 07:46:06 am »
It should be outlined in a Service manual, have you tried to find one?
TDS Teks in the past had a button press sequence to enter service mode in which each of the adjustable parameters could then be checked and adjusted against a known reference.
These days many DSO's a calibrated against the like of a Fluke scope calibrator, the model should be stated on your scopes Cal cert.
Avid Rabid Hobbyist
« Reply #2 on: May 22, 2016, 07:57:21 am »
Yes, I've tried to find a service manual. Perhaps others have better Google-Fu than I. However, I vaguely recall a post or two about the lack of much technical information about these scopes. Hence, my ponderings.
Online Fungus
« Reply #3 on: May 22, 2016, 08:31:03 am »
Is it out of spec now? If it isn't then there's no reason to think it ever will be.  :-//
That's actually its main purpose.
Oscilloscopes aren't multimeters. You shouldn't expect highly accurate voltage measurements from them.
Most DSOs use 8-bit ADCs attached to circuitry that's designed to respond to high frequencies. This means you have about 1% accuracy under ideal conditions (and that's only when your vertical gain is set so that the signal occupies the whole screen).
The specs for the DS1054Z are:
DC Gain Accuracy: <10 mV: ±4% full scale, ?10 mV: ±3% full scale
DC Offset Accuracy: ±0.1 div ±2 mV ±1% offset value
1% offset error + 3% gain error could add up to 4%. Add in the ADC error, improper vertical scale setting, some RF noise from the lights, imperfect probing, etc. and you can easily have 5-10% error in the on-screen numbers.
Bottom line: Oscilloscopes aren't multimeters.
(Yes, I'm sure there's some old green-screen "Tek" that's sub-1% accurate, but... )
I doubt they have anything more than their self-calibration function.
« Last Edit: May 22, 2016, 08:36:26 am by Fungus »
Offline EI6JA
  • Country: ie
« Reply #4 on: May 22, 2016, 05:02:56 pm »
The Performance Verification Guide is available at http://beyondmeasure.rigoltech.com/acton/attachment/1579/f-0564/1/-/-/-/-/DS1000Z%20MSO1000Z%20Performance%20Verification%20Guide.pdf
It does not cover adjustment but should let you know if your oscilloscope is within specification.
« Last Edit: May 22, 2016, 06:39:39 pm by EI6JA »
Offline uncle_bob
« Reply #5 on: May 23, 2016, 12:14:19 am »
Hi
I would not be surprised if there isn't much calibration on this scope. All of the timebase stuff is run off of a fixed crystal oscillator. If it's off ... replace it. All of the bandwidth stuff appears to be select at test parts. About all that leaves are the gains. At the precision of the scope, that may all be dependent on fixed ratio resistors. Again, if they are broke ... replace them.
Bob
« Reply #6 on: May 23, 2016, 01:53:15 am »
Thanks for the insights. My prior scope was a CRO, so it was mostly a case of curiosity as to what the long run is like for a DSO. If it does end up just being component replacements, no big deal. Alas, with the way electronics are these days, it'll probably be replaced by some bigger fancier scope before it goes out of whack.
« Reply #7 on: May 23, 2016, 02:03:30 am »
To give you some idea of what's involved with a modern DSO check the SM for a basic Siglent DSO:
http://www.siglentamerica.com/Uploadfile/file/20140925/SDS1000CML_ServiceManual_en.pdf
Pages 26 & 27 list the gear needed and describe the steps involved in the process.
Avid Rabid Hobbyist
« Reply #8 on: May 23, 2016, 04:33:26 am »
Thanks, tautech. Not much to it. Just click "test". :-DD Oh, wait. I need a Fluke 9500B.  :-BROKE
Even though there isn't a similar Rigol document handy, I'm sure it's something similarly automated.
« Reply #9 on: May 23, 2016, 05:05:48 am »
There will be but maybe not in the public domain.....yet.
One would expect the Self Cal process to flag a failure when it doesn't meet spec and warn the user that a full Calibration or repair is needed.
Depending on how involved and smart the Self Cal is it might even point to the failure, something like the warning in this post:
https://www.eevblog.com/forum/testgear/oh-no!-i'm-never-buying-any-agilent-keysight-equiment-ever-again!/msg945126/#msg945126
Until you see warning flags......continue normal transmission.  ;)
Link fixed
« Last Edit: May 23, 2016, 07:16:19 am by tautech »
Avid Rabid Hobbyist
« Reply #10 on: May 23, 2016, 06:05:30 am »
Roger! Normal transmission transmitting...normally.
