Author Topic: Schematic for Agilent E5070B  (Read 571 times)
Offline NeuroTopic starter
« on: May 03, 2022, 04:36:48 pm »
Hi everyone!
I've got a couple of dead Agilent E5070B that I'd like to repair and I was hoping to get a schematic somewhere. Agilent's website does not have one.
Any idea where I can find one?
Thank you very much!
« Last Edit: May 03, 2022, 05:26:31 pm by Neuro »
Offline fourierpwn
Re: Schematic for Agilent E5070B
« Reply #1 on: May 04, 2022, 12:02:08 am »
The power supplies in these units are commercial off the shelf (COTS) parts - not Keysight branded. Rather, they are Nipron PCSA-300P-X2S units
Typically, the CLIP packets for Keysight/Agilent gear don't include schematics for COTS parts, so if you suspect a dead power supply then I wouldn't bother chasing down a schematic. Rather, "thou shall check voltages"  :-DMM
The following users thanked this post: Neuro
« Reply #2 on: May 04, 2022, 06:32:33 pm »
The E5070B has a regular PC ATX power supply, so it's not a problem at all.  ;D
Internal tests are showing errors, coming from signal paths.
Online nctnico
« Reply #3 on: May 04, 2022, 06:45:00 pm »
If you have an idea which module / signal path is defective then it should be possible to get it working by doing some reverse engineering -assuming that the defective chips aren't custom ones made by Agilent for internal use-. Judging from the service manual, there is an extensive test suite that should bring you very close to the location of the problem. With an RF generator and oscilloscope / spectrum analyser it is possible to follow the signal path and see where it goes wrong.