I recently bought an Agilent 54701A for cheap which was said to be in good working order ("voll funktionstüchtig").
Since I don't have the official supply 1143A which is hard to get over here and very expensive anyway, I wanted to figure out how to build my own power supply.
From what I can tell from the service manuals of the probe and power supply, the pinning on the Lemo connector should be as follows:
1 red +17V
2 orange -17V
3 green GND
4 black (Offset +/-5mA)
The colors are the wires of the opened Lemo connector. The pinning order is actually marked on the soldering side of the Lemo connector (circle around pin 1, then a line connecting the following pins in their order).
I created a dual 17V supply from two (isolated) outputs of my bench supply, both set to 200mA (since the manual defined a maximum current draw of 110mA).
However, when I turn on the outputs, there are are 200mA drawn on the +17V line and nothing at the -200mA line. When I measure the resistance between these pin 1 (red) and 3 (green), it's only 500Ohm.
Actually, when measuring the resistance from the shielding to any of the pins, it looks like black pin (4 according to the Lemo pinout) is ground.
So it looks like the pinning is mirrored for pin 1..6 (the center pin 7 is not connected anyway).
If so, the following pinning would be correct:
6 purple +17V
5 blue -17V
4 black GND
5 green Offset +/-5mA
I'm a bit hesitant to try this as it seems to contradict the service manuals. I don't know how I could misinterpret the pinning order on the connector and I'm unsure if it's dared to even consider that Agilent did.
Side note: according to the service manual, the pins 5 and 6 should be iD0 and ID1. In Figure 16, they seem to be both connected to ground. However, even with the mirrored pinning it seems like only of of them is pulled to ground (500 Ohm) but the other one is not connected to GND at all.