Hi all,
Did some work on one of my instruments here, its an E4980AL-032 (300 kHz option). The E4980AL comes in 300, 500 kHz and 1 MHz flavors. The E4980A is the high freq. version that goes up to 2 MHz, with some nice DC bias options (up to 40V internal DC bias capability). All E4980A/AL are rated at same accuracy spec of 0.05% with same measurement ranges.
I noticed that the LCD glass was showing signs of delamination (it is an EMI shielded glass, so it contains layers with metal mesh). So this gave me an excuse to open it up and take a peek inside. I have not seen any detailed board pics from inside the E4980A/AL meters online.
Full high-res photo dump:
https://drive.google.com/drive/folders/1dR6SpquMtO1U3GbjPnZD62Gn6p-JvUbW?usp=sharingSince you only have to disassemble the front end, I did not take the entire thing apart. The back is just the processor and motherboard, so those aren't very interesting. The neat bits are the analog front end which actually does the LCR measurement. I also took some of the shielding cages off to get better pics.
I was a little surprised at some no-clean flux still being splattered all over the connectors that were soldered in by hand. A little sloppy for a "high-end" instrument but hey, the customer usually doesn't see it so...
I also had to add some extra EMI tape to shim up the LCD to the front panel/keypad assembly. I noticed that there was a gap and every time you press keys near the lower-right of the screen, the glass would move in. Tolerances aren't that great on large, thin pieces of plastic like the front panel I suppose.
Enjoy the pics.