PatrickB:
Hello,
With my DS1102Z-E scope, the brightness of the waveform decreases sharply when the time base is 5ns. At 10 ns and above or at 2 ns, the brightness is normal, strange! >:(
Do you have this problem with this scope or have you heard of this problem?
The firmware version is 00.06.3 SP2 and board version 0.0.1
Do you think it's a hardware problem with this board version?
Bye
lexvts:
Hello! A similar situation, even saves a picture with the same dim beam, in the PC program the beam for 5Ns is already normal. More likely, it's the firmware. The support service is not responding.
Fungus:
This isn't a CRT. Trace brightness is done in software.
RogerG:
Hello,
I own a brand new DS1102Z-E since a few weeks and tried to replace this horrible, non-indexed multipurpose encoder. I found the Bourns PEC11R-4215F-S0024 (24 positions) recommended at a german youtube channel, which is nicely indexed. But with this encoder the Scope always triggers the menu twice, so it jumps over one menu item :palm:.
Any other suggestions for the correct encoder? The Alps that is often discussed seems to be discontinued. It should be for SMD mount, but thats not mandatory.
Thanks in advance!
Regards from germany
Roger
Piton:
Hello. I also tried cheap Chinese encoders of the EC11 series, but they do not work well, skip and jump. Bought at local store EC11J1524413. Works as it should, but is a bit stiff when shifting and needs some mechanical tweaking.