Author Topic: Rigol DS1104 Oscilloscope and DG1022U Function Waveform Generator  (Read 1079 times)
Offline Dosupi
  • Country: cn
Hello Everybody,
I use Rigol DG1022U Function Waveform Generator to generate a sin waveform , V_pp = 1V, f = 1KHz.
Then, I use Rigol DS1104 Oscilloscope to watch the sin waveform , and do FFT on DS1104.
In picture 1, FFT: 20dBV   5KHz/Div   100K Sa/S
Amplitude at 1KHz is -8.8dBV, 3KHz is -50dBV, the difference between them is -41.6dBV.
In picture 2, FFT: 20dBV   1KHz/Div   20K Sa/S
Amplitude at 1KHz is -8.8dBV, 3KHz is -46.4dBV, the difference between them is -37.6dBV.
In picture 3, FFT: 20dBV   500Hz/Div   10K Sa/S
Amplitude at 1KHz is -9.6dBV, 3KHz is -36.0dBV, the difference between them is -26.4dBV.
Amplitude at 1KHz is -9.6dBV, 2KHz is -36.0dBV, the difference between them is -31.2dBV.
I don't know What's wrong with my test? and which data is correct  :'(
Offline Neganur
Re: Rigol DS1104 Oscilloscope and DG1022U Function Waveform Generator
« Reply #1 on: February 01, 2018, 06:00:02 pm »
What is it that you’re trying to measure?
-8.8dBV is around 1Vpp so the first picture looks correct.
The yellow sine wave in picture 2 and beyond looks distorted, try to increase the time/div without decreasing the sample rate (10ms/div is down to 50MSa/s ? Does the scope have more than 7.5Mpts of memory?)
EDIT: I saw the spec sheet mention that the sample rate is shared by all channels. Disable all except for the one you’re using to maximize the sample rate (1GSa/s * 12 div * 1ms = 12Mpts memory)
« Last Edit: February 01, 2018, 06:08:32 pm by Neganur »