I updated my scope with latest FW and the last thing the PDF instruction manual said was to restart the scope and run the self calibration found under the utility menu.
But under utility self cal is not present, i cannot remember that i have done self cal before so i cannot say if it has ever been present.
What to do?
Online tv84
  • Country: pt
Re: Rigol DS1104Z , self calibration menu not availible
« Reply #1 on: December 19, 2019, 01:48:38 pm »
Attached.
Offline JohnPen
« Reply #2 on: December 19, 2019, 02:14:20 pm »
It should be present in normal circumstances.
Select Utility
Press down arrow button on right side of screen
Top right menu item should show Self Cal
Select Self Cal
Select Start
Wait a few minutes to finish.
The following users thanked this post: I wanted a rude username
Offline MarkF
« Reply #3 on: December 19, 2019, 02:29:25 pm »
Press Utility menu
[attachimg=1]
Press Page Down button
[attachimg=2]
Start Self-cal.  (Insure scope has been powered 'ON' for 20 minutes or so.)
[attachimg=3]
Offline newbadboy
« Reply #4 on: December 20, 2019, 07:29:39 am »
Thx. ill check tonight.
Offline newbadboy
« Reply #5 on: December 20, 2019, 08:04:15 pm »
Yup found it :)
Running the cal as we speak!!