Title: Rigol DS1104Z , self calibration menu not availible
Post by: newbadboy on December 19, 2019, 09:42:07 am
I updated my scope with latest FW and the last thing the PDF instruction manual said was to restart the scope and run the self calibration found under the utility menu.

But under utility self cal is not present, i cannot remember that i have done self cal before so i cannot say if it has ever been present.

What to do?

Title: Re: Rigol DS1104Z , self calibration menu not availible
Post by: tv84 on December 19, 2019, 01:48:38 pm
Attached.
Title: Re: Rigol DS1104Z , self calibration menu not availible
Post by: JohnPen on December 19, 2019, 02:14:20 pm
It should be present in normal circumstances.

Select Utility
Press down arrow button on right side of screen
Top right menu item should show Self Cal
Select Self Cal
Select Start

Wait a few minutes to finish.
Title: Re: Rigol DS1104Z , self calibration menu not availible
Post by: MarkF on December 19, 2019, 02:29:25 pm
Press Utility menu
[attachimg=1]

Press Page Down button
[attachimg=2]

Start Self-cal.  (Insure scope has been powered 'ON' for 20 minutes or so.)
[attachimg=3]
Title: Re: Rigol DS1104Z , self calibration menu not availible
Post by: newbadboy on December 20, 2019, 07:29:39 am
Thx. ill check tonight.
Title: Re: Rigol DS1104Z , self calibration menu not availible
Post by: newbadboy on December 20, 2019, 08:04:15 pm
Yup found it :)

Running the cal as we speak!!

THX