I updated my scope with latest FW and the last thing the PDF instruction manual said was to restart the scope and run the self calibration found under the utility menu.
But under utility self cal is not present, i cannot remember that i have done self cal before so i cannot say if it has ever been present.
What to do?
-
Attached.
-
It should be present in normal circumstances.
Select Utility
Press down arrow button on right side of screen
Top right menu item should show Self Cal
Select Self Cal
Select Start
Wait a few minutes to finish.
-
Press Utility menu
[attachimg=1]
Press Page Down button
[attachimg=2]
Start Self-cal. (Insure scope has been powered 'ON' for 20 minutes or so.)
[attachimg=3]
-
Thx. ill check tonight.
-
Yup found it :)
Running the cal as we speak!!
THX