hey guys,
Does anybody have a suggestion on what kind of high voltage probe I need to get that will work with my Fluke 87-V to measure 5KV-10KV?
I know nothing about these probes. I assume they do voltage division of 10:1 or 100:1. Do I have that right?
I was hoping to get something on on eBay for $50-$60, if not, what do I need--or do I have the wrong idea entirely?
Thanks,
I have the Fluke 80K-6 and it works fine business. I used it to check voltages on an analog scope with no issues. I paid $65 for a Fluke 27FM multimeter, a set of probes and a hard case along with the HV probe. It was military surplus and the probe looked like it hadn't been used.
Since you stated that you need to measure voltages between 5 and 10 kV, an 80K-6 would not be adequate.
The Fluke 80K-15 or 80K-40 probes would do the job, and can be readily found on eBay:
This 80K-40 is currently on offer at $89.95.
hey guys,
Does anybody have a suggestion on what kind of high voltage probe I need to get that will work with my Fluke 87-V to measure 5KV-10KV?
I know nothing about these probes. I assume they do voltage division of 10:1 or 100:1. Do I have that right?
I was hoping to get something on on eBay for $50-$60, if not, what do I need--or do I have the wrong idea entirely?
Thanks,
I did a video on this topic... ...be careful that the input impedance of the meter will change depending on the voltage range!
I did a video on this topic... ...be careful that the input impedance of the meter will change depending on the voltage range!
w2aew,
What a nice Video you made, thanks for that.
I had no idea, that the RCA probe was available at so many different resistor values, always thought that they are all 990 MOhm.
Thanks guys, that clears it all up...and excellent video too!
Regards,
Tom
I did a video on this topic... ...be careful that the input impedance of the meter will change depending on the voltage range!
I learned this lesson with a Fluke 80K-6 which I picked up for measuring the cathode voltage on a Tektronix 7834 that I was refurbishing. I wanted to check the calibration of the probe which was used but in excellent condition and the multimeter I was using but when I broke out my HP3478A, it read high on the lower voltage ranges (30mV, 300mV, 3V) while all of my other multimeters, I have quite a collection, were consistent. I had already suspected this would happen and a quick calculation showed that the high reading was consistent with the others when the high input resistance was considered. I ultimately satisfied myself that accuracy would be well within the specification for the probe although it would be nice to have a high voltage calibrated source for such checks.
This would be an even larger problem with cheap multimeters that do not have a calibrated 10 MOhm input resistance.
I still use the 7834. It is faster than any DSO that I can afford.