need cal procedure for the HP3478A
« on: May 10, 2013, 07:42:17 pm »
hey

I have got any standard needed , just don't know the procedure , can someone instruct me ?

I've rotated the "Cal" rotary swich and the letter "C" is now on screen , then I pushed the shift and thee "cal" buttons , then the device ask " 10.0000 ? " and from there I don't know what's next  :-//

thank you in advance ! 
 

Offline PA0PBZ

  • Super Contributor
  • ***
  • Posts: 5192
  • Country: nl
Re: need cal procedure for the HP3478A
« Reply #1 on: May 10, 2013, 07:47:21 pm »
Did you bother to read the manual?
Keyboard error: Press F1 to continue.
 

Offline mikes

  • Regular Contributor
  • *
  • Posts: 127
  • Country: us
Re: need cal procedure for the HP3478A
« Reply #2 on: May 10, 2013, 07:50:05 pm »
go to agilent.com

under "Electronic Test & Measurement", click on "Manuals"

Find "3478a" using the search box

Download the Service Manual (you may have to register)

Calibration starts on page 4-11 (section 4-48)

(you really should have full cal instructions in front of you, and read them, before you start setting things into cal mode)
 

Offline eevblogfanTopic starter

  • Frequent Contributor
  • **
  • Posts: 569
  • Country: 00
Re: need cal procedure for the HP3478A
« Reply #3 on: May 10, 2013, 07:50:27 pm »
the menual is long and I understood that there are few ( I am shure I am wrong but that's my impression)
 

Offline eevblogfanTopic starter

  • Frequent Contributor
  • **
  • Posts: 569
  • Country: 00
Re: need cal procedure for the HP3478A
« Reply #4 on: May 10, 2013, 08:39:27 pm »
thank you mikes , I've tweaked the meter and now it reads down to the LSD ( compare to fluke 5700 )

 