The Flukes of this type use the zebra strip, like the 87. For generally faded segments, cleaning the strips, and the PCB contacts will usually solve the problem. Be very careful - the bonded conductive pads on the glass are VERY fragile. I'd say if the zebra strip is stuck to the LCD, don't separate them. Also, don't attempt to clean the LCD contacts.
The common display problem with these is black blobs. They start small, and get larger. Sometimes the small blobs don't get larger very fast, sometimes they do.