Hi all,
Very long time lurker here, but never had a real need to post as most of my questions had already been asked and answered. Search button = time saver.
But I now have an issue I need advice/feedback on. I have one of the original Rigol DS2072 scopes. As I am not in the business it has sufficed for the occasions I needed a scope. Work and family commitments never left me much time to play anyway. Now I am retired and getting back into Ham Radio and currently updating the shack workbench.
To the issue at hand. I hope I can make this coherent.
When I turn on a measurement such as Vmax, Vpp there is a value displayed that is much higher than I would expect, even accounting for the intrinsic noise of the scope. It is also there when a signal is supplied.
On CH1 At 10V/Div it shows a Vpp value of 800mV (sometime flickering to 1.20V), doesn't matter if coupling is DC or AC, or if the inputs are open or shorted. Even tried a 50Ohm terminator and no difference.
If I input 10V DC from my Agilent 8001 PS (Confirmed the 10V accuracy with my Agilent 1241 and R&S HMC-8012, I don't have anything more accurate), the scope will show 10.8V. But as near as I can tell I do not see this extra 800mV signal on the trace. If I change coupling to AC the DC input goes as expected and I have just the 800mV. I thought it was noise but the only values showing if I display all measurements are Max, Vpp, Vrms, Avg, Area, and Vmin (though this tends to be always 0. None of the others like Vtop, Rise, Fall, have values.
This 800mV changes when changing V/Div.
I ran through all the settings from 10V/Div to 500uV/Div with both channels on all measurements displayed. No input signals.
CH1
5V/Div - 400mV
2V/Div - 160mV
1V/Div - 80mV
500mV/Div - 40mV
..........
5mV/Div - 800uV
This pattern continues until we reach 2mV/Div where its starts fluctuating around 500uV, At 1mV/Div most if not all the other measurements start to show, duty, period, freq. At 500uV/Div all show and we are in the noise floor. I thought I read somewhere that these scopes go to 2mV/Div then use digital magnification or something to get down to 500uV.
This pattern is similar on channel 2 also, although sometimes it will show 600mV instead of 800mV, but it divides down same ratio as CH1. Values a little bit different but consistent as V/Div changed. I am beginning to think it is firmware but I updated it to the latest available with no change. Also did a number of calibrations. I did note on one that CH2 came back showing half the values of CH1, ran it again and it then matched CH1. Once I turned off CH2 and CH1 value halved from 800mV to 400mv, re-enable Ch2 back to 800mV and it wouldn't do it again.
I noted that with a signal applied to CH1 turning off CH2 changes the "added value".
It was a bit different inputting a sinewave. I input a 1Khz sine at 5Vpp. It skewed the values of each half of the waveform. + went up by 40mV and - down by 20mV. Too small to see on the trace with V/Div at 1V/Div. Oddly changing to 500mV/Div and using vernier adjustment to bring the trace back onto the screen fully, the values were much closer to the input signal (if the Sig-gen is correct of course). I gather a lot of these values are done based on whats on screen or it seems that way.
I even removed it from the mains and ran it off a battery and inverter. I get a lot of QRM around here and wanted to check if anything was getting in via the mains. The inverter is not a good one but I don't think it was introducing anything as there was no change in the values.
I have even gone old school and counted the division/sub-divisions and the input voltage looks correct. The traces look correct. Just the damn measurements are off. A signal from the DG4000 Sig-gen "looks" ok and frequency is measured correctly. I don't have another scope to compare to as my Tek 2215A died and this is where I need the DSO to fix it.
I am at the point where I can view signals but don't trust the measurements. It's painful to check the offset for that setting and allow for it.
Does this sounds like a hardware or software issue?. I don't have the experience to decide either way. I hope the above doesn't sound like the drivel of an idiot.
I am going to buy a new scope anyway, and no I am not going to start one of those threads.
Regards
Rob