I appreciate test equipment having good protection, but I think we should be really careful with spec contest on input voltage. The Siglent affordable scopes (like all the lower cost oscilloscopes), ship with 1x/10x probes that have 300V peak-peak rating in 1x mode --> that's 106V rms and none of the system is CAT rated! It is not much help if the oscilloscope survives but the probe that the unsuspecting hobbyist is using is fizzing and shorting in their bare hand!
I can't think of any scenario where it is appropriate to deliberately probe 1:1 near mains voltages. One of the first things I bought was differential probes!
The important things first: Nobody here advocates probing a mains powered circuit with an oscilloscope as long as you don't absolutely know what you're doing. This includes to use appropriate probes and avoid the ones that can be switched to x1 mode.
The cheapest Probe (the one that will ship with a Siglent SDS1104X-U) is the PP510. It is rated
300 Vrms CAT II in x10 mode and
150 Vrms CAT II in x1 mode.
Since the scope cannot show more than 80 Vpp on the screen at its lowest sensitivity of 10 V/div anyway, a x1 probe is not helpful and its accidental use should become obvious quite quickly.
BTW, there are single ended probes specifically for mains, like the PB925, a x10 probe with 250 MHz bandwidth, that is 600 V CAT III and 1000 V CAT II rated.
EDIT: of course in most of today's use cases for mains voltage probing, a differential probe will be required!