Author Topic: Rigol MSO5000 Artifacts  (Read 4829 times)

0 Members and 1 Guest are viewing this topic.

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 6600
  • Country: hr
Re: Rigol MSO5000 Artifacts
« Reply #25 on: January 17, 2022, 10:49:29 pm »
The only thing that seems a bit baffling is why it goes away when you select CH2 or CH4.  I wonder if the 'noise' might be an instability in the ADC timing, as others previously referred to when the asked you to autocal the scope.
Or the path between the two ADCs is picking up noise. I can imagine that there is a path (PCB trace) that brings the signal from channel 1 to the ADC of channel 2. If that is routed badly, then it could pick up noise.

There are no two or whatever ADC chips inside scope. It has single ADC chip (Ankaa SP ASIC) that has  1 2 and 4 paths. Inside ADC are several (at least 4) interleaved ADCs. They went Keysight way of inside that chip can actually having 64 slower interleaved ADC.. We don't know for sure. There is a patent, they are using parallel ADC inside chip.
That Ankaa chip is also said to have oscilloscope specific DSP functions. That could be some corrections or what not.

And with massively parallel ADC, adjusting them all together to achieve low noise and monotonic performance is hard..
 
The following users thanked this post: JeremyC

Offline phs

  • Contributor
  • Posts: 26
Re: Rigol MSO5000 Artifacts
« Reply #26 on: January 18, 2022, 12:02:22 am »
Just for another reference:  SDG2042X, 2Vpp Sin @ 120 MHz from channel 1 via a bnc “T” on the SDG output, set to 50 ohms, in to channels 1 & 3 of each of the scopes (MSO5000 and RTB2000), which are each set to 1:1 with a 50 ohm terminator on each channel input, via a 50 ohm impedance cable (RG316), bandwidth limiting off.

Guess we need to send our RTBs back too, haha.  Actually not very worried about this - there are quite a few variables that could contribute, I would guess.  Though it's an interesting phenomenon, likely related to the effects of the sampling mechanisms/relatively low bit-depths/sig gen jitter/trigger mechanisms, I don't see it as a major issue, though I certainly could be wrong, and am not afraid to admit it. 

This kind of reminds me of the (over?) obsession about input noise levels, though I can certainly understand the OP wanting to understand exactly what’s going on. 

In my experience I actually don't mind the slightly higher noise floor on the Rigol, compared to the RTB.  Its waveform update rate is many times more than that of the RTB, esp. at low signal levels/small time scales, and it seems like it's actually easier to see some low-level signals on the Rigol, maybe because it's not filtering so aggressively, and likely other factors, too.  I'd rather see a bit more noise, rather than filter out more of the signal.

Here, the Rigols get used more frequently than the RTBs simply because they have numerous features the RTB scopes lack (Nth-edge triggering, at least 10 Mbps UART decoding, zone triggering, etc.) and the generous array of options on the Rigol scopes allow a wide variety of tasks to just get done, every day.  In my opinion, for the money especially, they are bad-ass powerful tools that can bring in the cash -- as long as you're aware of the limitations -- just like _any_ tool. 

I do personally like much about the RTB scopes, though – mainly the web interface, which is the best I’ve seen for a scope so far.  And, they have been adding sorely missed features … very slowly.  The FFT is also very nicely done and looks beautiful, too.  But still, the Rigols are the go-to tools for many jobs where the added features really come in handy – crude though they may sometimes be, they get the job done. 

Don’t really like touch screens on scopes at all, and am very thankful these instruments work so well with a wireless mouse!  This is another area where the Rigol wins over the RTB – no need to use the touch screen, ever.

In the end, as I'm sure you all know, it's what you accomplish using the tools, rather then the obsession over the instruments themselves -- that's what's important (to me, at least...)

Would love to try Siglent scopes sometime -- if they ever decide to give the user full control over zoom functions.  As they are now, they're a bit like a manual mode on a DSLR that decides it will set certain parameters for you -- whether you want it to or not.  No thanks!

Good luck on any continued investigation of this phenomenon, and hope the RTB image will provide some added perspective.  Will watch from afar.  All the best!




 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28328
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Rigol MSO5000 Artifacts
« Reply #27 on: January 18, 2022, 12:10:23 am »
Would love to try Siglent scopes sometime -- if they ever decide to give the user full control over zoom functions.  As they are now, they're a bit like a manual mode on a DSLR that decides it will set certain parameters for you -- whether you want it to or not.  No thanks!
Seems you won't have to wait much longer as alternative memory management options are to be implemented in the new 12 bit HD model.
https://www.eevblog.com/forum/testgear/siglent-sds2000x-hd-12bit-(published-for-chinese-domestic-market-only)/
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 
The following users thanked this post: phs

Offline phs

  • Contributor
  • Posts: 26
Re: Rigol MSO5000 Artifacts
« Reply #28 on: January 18, 2022, 12:18:38 am »
Quote
Seems you won't have to wait much longer as alternative memory management options are to be implemented in the new 12 bit HD model.
https://www.eevblog.com/forum/testgear/siglent-sds2000x-hd-12bit-(published-for-chinese-domestic-market-only)/

Been wondering whether the new model might have addressed that -- excellent news!  Would be great if it could filter down the line with a f/w update at some point.

Will be keeping an eye out -- thanks for the info!
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 28328
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Rigol MSO5000 Artifacts
« Reply #29 on: January 18, 2022, 12:25:41 am »
Quote
Seems you won't have to wait much longer as alternative memory management options are to be implemented in the new 12 bit HD model.
https://www.eevblog.com/forum/testgear/siglent-sds2000x-hd-12bit-(published-for-chinese-domestic-market-only)/

Been wondering whether the new model might have addressed that -- excellent news!  Would be great if it could filter down the line with a f/w update at some point.

Will be keeping an eye out -- thanks for the info!
I have a hunch this is in the new SDS6000A too but until I get mine in a month or so....no comment.  ;)

There was talk of implementing it in existing models but don't know where that's gone as if they do they sure don't want to break any existing functionality.  :scared:
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 
The following users thanked this post: phs

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4091
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Rigol MSO5000 Artifacts
« Reply #30 on: January 18, 2022, 04:28:51 am »
Quote
Seems you won't have to wait much longer as alternative memory management options are to be implemented in the new 12 bit HD model.
https://www.eevblog.com/forum/testgear/siglent-sds2000x-hd-12bit-(published-for-chinese-domestic-market-only)/

Been wondering whether the new model might have addressed that -- excellent news!  Would be great if it could filter down the line with a f/w update at some point.

Will be keeping an eye out -- thanks for the info!
I have a hunch this is in the new SDS6000A too but until I get mine in a month or so....no comment.  ;)

There was talk of implementing it in existing models but don't know where that's gone as if they do they sure don't want to break any existing functionality.  :scared:

"I have a hunch this is in..."

No need to hunch neither guess.
 
Simply: It is there in SDS6kA: Memory management. Normal auto and then if user really need for some rare or weird purposes there is also available fixed sample rate and fixed memory.
SDS6000A Series Digital Oscilloscope User Manual version EN01A, page 90: Memory Management
SDS6000A Quick Start version EN01A, page 25: (D) Select the Memory Management mode (Auto, Fixed Sample Rate, and Fixed Memory)
« Last Edit: January 18, 2022, 04:38:38 am by rf-loop »
I drive a LEC (low el. consumption) BEV car. Smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the wises gone?
 

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4091
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: Rigol MSO5000 Artifacts
« Reply #31 on: January 18, 2022, 05:02:22 am »
Don’t really like touch screens on scopes at all, and am very thankful these instruments work so well with a wireless mouse!  This is another area where the Rigol wins over the RTB – no need to use the touch screen, ever.

Example Siglent SDS6000A UI is made so that if user want he do not need use touch panel or physical control panel at all. It can well use alone with wired or wireless mouse. (and naturally also with all control methods mixed)
I drive a LEC (low el. consumption) BEV car. Smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the wises gone?
 
The following users thanked this post: phs

Offline bdunham7

  • Super Contributor
  • ***
  • Posts: 7823
  • Country: us
Re: Rigol MSO5000 Artifacts
« Reply #32 on: January 18, 2022, 05:15:18 am »
Guess we need to send our RTBs back too, haha.  Actually not very worried about this - there are quite a few variables that could contribute, I would guess.  Though it's an interesting phenomenon, likely related to the effects of the sampling mechanisms/relatively low bit-depths/sig gen jitter/trigger mechanisms, I don't see it as a major issue, though I certainly could be wrong, and am not afraid to admit it. 

This kind of reminds me of the (over?) obsession about input noise levels, though I can certainly understand the OP wanting to understand exactly what’s going on.

Neither of your photos looks unusual and certainly doesn't show any issues like the ones the OP is reporting.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 
The following users thanked this post: nctnico

Offline phs

  • Contributor
  • Posts: 26
Re: Rigol MSO5000 Artifacts
« Reply #33 on: January 18, 2022, 05:44:23 am »
Quote
Example Siglent SDS6000A UI is made so that if user want he do not need use touch panel or physical control panel at all. It can well use alone with wired or wireless mouse. (and naturally also with all control methods mixed)

Great to hear!  Have appreciated your posts over the years -- this is another good one.  Haha!

Siglent continues to impress.  It's exciting to watch...
 

Online nctnico

  • Super Contributor
  • ***
  • Posts: 26896
  • Country: nl
    • NCT Developments
Re: Rigol MSO5000 Artifacts
« Reply #34 on: January 18, 2022, 11:30:34 am »

Simply: It is there in SDS6kA: Memory management. Normal auto and then if user really need for some rare or weird purposes there is also available fixed sample rate and fixed memory.
SDS6000A Series Digital Oscilloscope User Manual version EN01A, page 90: Memory Management
SDS6000A Quick Start version EN01A, page 25: (D) Select the Memory Management mode (Auto, Fixed Sample Rate, and Fixed Memory)
But decoding, math, measurements, etc are still limited to what is on screen. So it is still useless  :palm:

But this is getting wildly offtopic here.
« Last Edit: January 18, 2022, 02:03:00 pm by nctnico »
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline core

  • Regular Contributor
  • *
  • Posts: 131
  • Country: ro
Re: Rigol MSO5000 Artifacts
« Reply #35 on: January 18, 2022, 03:15:58 pm »
Here Rigol MSO5074 CH1 and CH3, , 2 x 50 ohm cables with 50 ohm terminators.
Siglent SDG2042X, 120Mhz sinus, 2Vpp, 50 ohm output.

It seems there are no problems. I don't see any peaks like yours.

Just in case, check the installed firmware version.

v00.01.03.00.03 2021/10/18
      - Optimized waveform display in XY mode.
      - Optimized the DC gain calibration algorithm.
      - The La channel is decoded in parallel, which solved the problem of decoding error in negative polarity.

Any firmware optimisation regarding ADC calibration can help.
 

Online nctnico

  • Super Contributor
  • ***
  • Posts: 26896
  • Country: nl
    • NCT Developments
Re: Rigol MSO5000 Artifacts
« Reply #36 on: January 18, 2022, 03:22:37 pm »
Please use the same time scale the OP uses; likely this is key to reproducing the issue. Also, the issue may not show always so it could be missed. Also try 130MHz instead of 120MHz.
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline core

  • Regular Contributor
  • *
  • Posts: 131
  • Country: ro
Re: Rigol MSO5000 Artifacts
« Reply #37 on: January 18, 2022, 04:21:24 pm »
Please use the same time scale the OP uses; likely this is key to reproducing the issue. Also, the issue may not show always so it could be missed. Also try 130MHz instead of 120MHz.

I have used the same time scale and 500mV/div, 120MHz. I don't have a 130MHz signal.
Almost no problem. The maximum peak I can find it's like in the attached picture. But it can be just noise, my environment is not clean. Near the oscilloscope it's a powerfull desktop and in the same room a wi-fi router.

I have attached another 100mV/div capture, with the maximum peak I can find.
« Last Edit: January 18, 2022, 05:12:03 pm by core »
 

Offline jacekowskiTopic starter

  • Contributor
  • Posts: 14
  • Country: gb
Re: Rigol MSO5000 Artifacts
« Reply #38 on: January 18, 2022, 05:19:56 pm »
Can you change trigger slope and set trigger voltage as low as it goes and check what you see there?
 

Online nctnico

  • Super Contributor
  • ***
  • Posts: 26896
  • Country: nl
    • NCT Developments
Re: Rigol MSO5000 Artifacts
« Reply #39 on: January 18, 2022, 05:39:23 pm »
Please use the same time scale the OP uses; likely this is key to reproducing the issue. Also, the issue may not show always so it could be missed. Also try 130MHz instead of 120MHz.

I have used the same time scale and 500mV/div, 120MHz. I don't have a 130MHz signal.
Almost no problem. The maximum peak I can find it's like in the attached picture. But it can be just noise, my environment is not clean. Near the oscilloscope it's a powerfull desktop and in the same room a wi-fi router.

I have attached another 100mV/div capture, with the maximum peak I can find.
If it is external noise, then I would expect both traces to show the exact same signal disturbance. However, that doesn't seem to be happening.
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline masterx81

  • Frequent Contributor
  • **
  • Posts: 481
  • Country: it
Re: Rigol MSO5000 Artifacts
« Reply #40 on: January 18, 2022, 05:46:53 pm »
If i've well understood, as it's specified in the first post, the problem on one scope was at 120mhz, and on a replacement unit "moved" at 130mhz, so maybe it's not on all the scopes at the same frequency (so in yours at 120mhz can be fine, but show it at 130mhz).
Maybe the artifact/noise can move a bit.
 

Offline core

  • Regular Contributor
  • *
  • Posts: 131
  • Country: ro
Re: Rigol MSO5000 Artifacts
« Reply #41 on: January 18, 2022, 06:26:29 pm »
Can you change trigger slope and set trigger voltage as low as it goes and check what you see there?

I have done for CH1 falling slope trigg, CH2 also, and CH3. There are no differences. See captures.
I have tried freq down to 100MHz, no changes.

If you want to check something else, just let me know. Limited to 120Mhz.
 

Offline core

  • Regular Contributor
  • *
  • Posts: 131
  • Country: ro
Re: Rigol MSO5000 Artifacts
« Reply #42 on: January 18, 2022, 06:33:17 pm »
Please use the same time scale the OP uses; likely this is key to reproducing the issue. Also, the issue may not show always so it could be missed. Also try 130MHz instead of 120MHz.

I have used the same time scale and 500mV/div, 120MHz. I don't have a 130MHz signal.
Almost no problem. The maximum peak I can find it's like in the attached picture. But it can be just noise, my environment is not clean. Near the oscilloscope it's a powerfull desktop and in the same room a wi-fi router.

I have attached another 100mV/div capture, with the maximum peak I can find.
If it is external noise, then I would expect both traces to show the exact same signal disturbance. However, that doesn't seem to be happening.

Indeed. So it can be the internal noise / ADC artifacts.
 

Offline thinkfat

  • Supporter
  • ****
  • Posts: 2150
  • Country: de
  • This is just a hobby I spend too much time on.
    • Matthias' Hackerstübchen
Re: Rigol MSO5000 Artifacts
« Reply #43 on: January 18, 2022, 06:40:12 pm »
The MSO5000 has a reputation of being noisy. Down on the 50mV/div region you're bound to experience some of it. Not sure if that setting is actually analog gain or just zoom in on 100mV.
Everybody likes gadgets. Until they try to make them.
 

Offline core

  • Regular Contributor
  • *
  • Posts: 131
  • Country: ro
Re: Rigol MSO5000 Artifacts
« Reply #44 on: January 18, 2022, 07:43:48 pm »
The MSO5000 has a reputation of being noisy. Down on the 50mV/div region you're bound to experience some of it. Not sure if that setting is actually analog gain or just zoom in on 100mV.

I have this MSO5074 and Siglent SDS-1104X-E. Down to 20mV/div the noise is about the same. At 10mV/div I can see clear the difference in favor of Siglent.

But the signal's refresh rate is much more faster on Rigol. On Siglent is "frozen",the equivalent of a 4x average on Rigol. I believe it can be a different way to process the signal, apart from the much faster sample rate of Rigol.

Under 5mV/div there is no dubt noise advantage in favor of Siglent. The lower you go, the bigger the noise difference.

For me both of them are very useful, I like both.

Important for me is to know their strengths and weaknesses and use them accordingly.
« Last Edit: January 19, 2022, 07:27:59 am by core »
 

Offline normi

  • Regular Contributor
  • *
  • Posts: 78
  • Country: 00
Re: Rigol MSO5000 Artifacts
« Reply #45 on: January 20, 2022, 06:05:58 pm »
Most likely you have a faulty scope from the same batch as the first scope. But you are in a small group as most persons would have hacked the scope to the full bandwidth, so it could be related to the low pass filter on ch1 which limits it to 100M.

Try the new firmware as that has a calibration update, check FFT of CH 1 vs CH 3 and see if anything is noticeable different between the two.
 

Offline jacekowskiTopic starter

  • Contributor
  • Posts: 14
  • Country: gb
Re: Rigol MSO5000 Artifacts
« Reply #46 on: January 21, 2022, 06:18:01 am »
But you are in a small group as most persons would have hacked the scope to the full bandwidth, so it could be related to the low pass filter on ch1 which limits it to 100M.

Try the new firmware as that has a calibration update, check FFT of CH 1 vs CH 3 and see if anything is noticeable different between the two.

I would have hacked it, but i wanted to wait with voiding my warranty until i'm sure the thing is 100% working.
Latest firmware did nothing (it was the first thing i did).

Rigol went back to the distributor for a refund so i can't do any more testing on it.
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 16642
  • Country: 00
Re: Rigol MSO5000 Artifacts
« Reply #47 on: January 21, 2022, 07:04:22 am »
I would have hacked it, but i wanted to wait with voiding my warranty until i'm sure the thing is 100% working.

It wouldn't void your warranty, it's just the same as installing an official upgrade.

Plus it can be reset to factory state just as easily as it can be hacked.
 

Offline jacekowskiTopic starter

  • Contributor
  • Posts: 14
  • Country: gb
Re: Rigol MSO5000 Artifacts
« Reply #48 on: January 21, 2022, 09:36:55 am »
As far as i know installing unofficial "upgrades" voids the warranty, wherever they can prove it is another thing (but then me admitting that on a public forum wouldn't be a good idea).
 