Products => Test Equipment => Topic started by: icemanh on July 05, 2020, 02:04:48 pm
-
Hello
My DG1022U is bricked while upgrading firmware. My hardware version is 3, bootloader version is 0. So I upgraded the bootloader and it looked fine.
However, I found it halted during DSP. Only the Mod, Store/Recall, Burst buttons lighted up forever. No choice that I had to switch it off. Retried the upgrade and everything is the same..
I am using a 2G usb stick.
Any idea how I can fix it?
Thank you...
-
I solved the problem by using different FAT32 formatted USB stick.
I noticed that the DG1022 kept trying to read USB when it stuck at Burst button lighted on. It may indicate there is some wrong in USB reading.
So I tried different USB sticks, 1G, 2G, 4G etc. Finally a Lexar 4G worked good and now DG1022U resumed to normal...
-
Glad you fixed it. Was it a Rigol (and not Rigor)? Rigol gear is crazy-fussy about the USB drives, has been a source of problems for me in several occasions now. I recently had one where the drive that loaded the new bootloader worked but then that same drive would not work for the GE file upload.
When you find a drive that works for a specific piece of test gear, label it and put it on one side, that's what I have done.