jjoonathan:
I am away from home for the week and cannot check but it seems unlikely that the problem is with those termination resistors / dc coupling capacitors. Do you have connectivity through the two 90 degree board-to-board connectors?

TRN:
Hi jjoonathan,

Thanks for the pictures, but the
link for All teardown photos: https://photos.app.goo.gl/4gtMHPGbsLmii7yJ9 does not work, and returns an error. Can you re-upload?

Also do you have a picture of the frontend without the shielding/ heat-sink cover installed?

rgds

myhobo:
I enter the service code 16052008f, then click flash, it can't be used now, the error is as follows

jjoonathan:
It looks like the flash was incomplete or put the wrong firmware on there.

Did it let you select the file?

Did it stop half way through flashing?

If you run the app as administrator and enter the service password, does it let you try again?


@TRN: I've fixed the teardown photo link, it should work now :)

jjoonathan:
Hmm, both the front panel and the FEC are missing material numbers. It seems unlikely that they both flashed incorrectly.

Hypothesis 1: a driver is broken. Running as administrator or re-installing the RTx app might fix this.

Hypothesis 2: a hardware connection upstream of both isn't fully seated.