TRN:
Thanks for the feedback, and I will try to dig into this a little further, and will let you know if there is anything to report

Diesus:
@myhobo
I hope that you can help me.
Im saw a oscilloscope RTO1024 in ebay, so I'm thinking to buy it or not and it depends of your response so this oscilloscope has the same issues that your reported with self alignment failing, and even didn't detect the channels. I attached the pictures with the issue.
Can you tell me the last status of your investigation? Was you able to fix it?
This to understand if the effort to repair the oscilloscope is worth it.

TRN:
Hi Diesus,

As far as I know, myhobo was unable to fix the issues with his scope.

The mask-test results from the partial self-test log indicate that there is no communication between the ADC's & the Gigabit receivers on any channel.

This might mean that;

1)either, all ADC's on the frontend board are defect, which is very unlikely
2 there is something wrong with the two HSEC8 Boards which connect the front-end with the main board, which is also very unlikely.
3)there is a problem with the two RTC ASIC's, which is more likely.

These ASIC's are manufactured by R&S, and they do not sell these. So if the problem is related to them you are out of luck.

So in my opinion it is very unlikely that you will be able to fix this scope.

leonardo:
Hello,

Does anybody manage to hack the RTO software options? R&S still sell those options?

nix239:
Hi.
I have couple of questions.
1. I have old hdd with FW 2.45.1.1 but some files are damaged and now options not work. maybe sombody have image or same fw file.
2. Update to 3.70 solve problem with options?Сan i revert to 2.45.1.1 with old hdd if something goes wrong?