Another Rohde & Schwarz HMP4040 user here...
I can confirm that the mains power switch does affect the (turned-off) outputs. On switch-on there's a short 0.3-0.4 V pulse on all outputs, and on switch-off the outputs ramp up to 1 V and stay there for several seconds.
However, I am unable to reproduce output 1 climbing above 1 V, let alone 10 V... I've retried multiple times, with different set voltages (outputs off), but the traces always looked the same. I did not experiment with different loads - only the oscilloscope or a multimeter.
The screenshots below are switching on and off with the mains switch, no load (other than the oscilloscope) and all outputs off.
Switch-onIn about 10 ms all outputs ramp up to 0.3-0.4 V and then fade back to 0 V in 30 ms. A fluke 289 in peak detect (250µs response time) measures the peaks between 0.33 and 0.38 V.
Switch-offThe outputs ramp up to 1 V in less than a second. Output 1 ramps up noticeably faster than the other three outputs. It gets a 600 ms head start, but even without that the slew rate is much higher.
Just under 6 seconds after switching off, the outputs suddenly drop to 0.5 V in under half a second. From there it takes another 4 seconds before the outputs finally reach 0 V again.
A Fluke 289 in peak detect does not read higher than 1.04 V over the entire period.
For reference: My HMP4040 is running the latest firmware (2.72). I don't know the hardware revision, but it is a "pre-facelift" unit (light front and single fan).
The Element14
review by Gough Lui shows a later "facelift" unit. Note that this power supply was seriously damaged in transport.
Which version do you have, electr_peter?