The problem is that at that voltage and frequency, you will have to look closely at the frequency derating curve of the probe. Many 100X probes won't be sufficient at that level. I imagine the power level must be quite high, but at 80kHz, your probes actual input impedance will only be a small fraction of its nominal value, so circuit loading is still a possible issue.
As for the current, the old Tek AM503 or AM503 amplifier with the A6302 probe would be nice.