Author Topic: R&S to launch MXO5  (Read 23058 times)
Offline Martin72
  • Testfield Technician
Re: R&S to launch MXO5
« Reply #125 on: December 13, 2023, 09:48:44 pm »
When I buy a scope for work, I also buy the options I need - it doesn't hurt me to pay for them.
For at home...Well.... :-X
Offline nctnico
« Reply #126 on: December 13, 2023, 09:51:37 pm »
You just get a lower bonus at the end of the year  :-DD
But likely your employer likes to see an ROI on the expense.
Offline Martin72
« Reply #127 on: December 13, 2023, 10:00:41 pm »
Well, to be honest, I only needed the options for development, so I only needed to equip 2 scopes with them.
When I submitted the investment request for 2 Lecroys, which said 42% discount, I was still the hero - even though they still cost 26000... Everything is relative. ;)
Offline Swirve
« Reply #128 on: December 14, 2023, 11:56:39 am »
Thanks for the input. As I guessed no one really have hands on experience in comparing the three scopes. Actually I do have the MXO5 on my bench and have started evaluating it, looking very promising so far! We do have a couple of HDO6140A, and if the only difference to 6000B is the bigger screen this doesn't give much.. But me being new in the company I haven't had time to play with the lecroy scopes. Also borrowed the power-rail probe, which together with the excellent spectrum is a great tool to evaluate noise on power rails.
The following users thanked this post: 2N3055, pdenisowski, Martin72
Offline egonotto
« Reply #129 on: December 15, 2023, 09:53:42 am »
In the Specifications of MXO 5 under
Memory segmentation states, among other things, that:
number of segments: 2113
record length: 1 Gpoints
total memory: 2113.000 Gpoints
Is that really correct? Then there should be 4 TB of storage.
Thanks to you and the other posters for mentioning this.  We just updated the spec sheet and will be posting it to the website soon.
Hello,
If I understand the new spec sheet correctly, 4 GB is available for segmented memory. So exactly the same as the RTA4000.
Or does that apply to each channel?
Best regards
egonotto
Offline pdenisowski
  • Product Management Engineer, Rohde & Schwarz
    • Test and Measurement Fundamentals Playlist on the R&S YouTube channel
« Reply #130 on: May 15, 2024, 10:21:15 am »
Just an FYI: the MXO5C (compact) was launched today
https://www.rohde-schwarz.com/products/test-and-measurement/oscilloscopes/rs-mxo-5c-oscilloscope_334265.html
Same scope, just smaller form factor.  Also available in 4 or 8 channel models
Test and Measurement Fundamentals video series on the Rohde & Schwarz YouTube channel:  https://www.youtube.com/playlist?list=PLKxVoO5jUTlvsVtDcqrVn0ybqBVlLj2z8
The following users thanked this post: ddrl46, shabaz, egonotto, trophosphere, tv84, jjoonathan, 2N3055, Martin72
Offline shabaz
« Reply #131 on: May 15, 2024, 01:48:45 pm »
Nice status display on that front panel!
This instrument would be great for permanent testbeds too, monitoring signals from one or more racks, to an attached monitor. At one workplace we ended up using a dedicated spectrum analyzer per rack, just for monitoring purposes. Nice that there's the potential with such types of test beds, to use the multiple spectrum capability of the MXO 5 to share it across several racks (if the signals are within the bandwidth of course).
The following users thanked this post: egonotto, pdenisowski
Offline moerm
  • Country: aq
  • pragmatic realist
« Reply #132 on: May 16, 2024, 01:53:12 am »
Front-up: I'll never buy that scope, not because it's not good (it probably is), but because it's way out of the hobby league - and I know that and while I understand that humans, once triggered, just love to discuss about "things in another galaxy" I also know that all the matters, facts, and "facts" discussed here are of no, none, nada practical concern to (certainly not only) me.
If this was a discussion in real life I'd probably jokingly complain to the R&S guy "what? Only 500 Mpts/ch? Come back again once you have a scope with >= 50 Gpts/ch!"
What I see here basically is just a MXO4.2. Some more of this, some more of that, a faster processor, yada,yada, but at the end of the day I just see a scope that makes better use of R&S' ASIC. That's nothing to sneeze at, but it also is not like we finally could fly to the next star, like "a new era".
And I'm amazed yet again how happily (almost) everyone accepts the usual BS. Example: If an ADC does 12 bits that doesn't mean that you actually get 12 bits because at the end of the day measuring, as opposed to guesstimating, means ENOB, and that is more like 9 bits - if you are very lucky and using a low sampling rate. Yes, averaging helps if some conditions are met, e.g. the thing you measure is repetitive. But still 12 bits is the limits.
Beyond that one finds oneself in BS territory, or in polite marketing lingo in math/statistics territory. Allow me to introduce you the 5 cats in my neighborhood. I know their precise size: 34.567 cm. OK, granted, not even a single one of them actually is 34.567 cm long, one, a kitten, easily and cozily fits in my hand while it's (likely) father is well over 50 cm long. Of bloody course we all like precision, accuracy, and high resolution, but let us not forget what it's all about: measuring that is, obtaining quantitative information about something physical!
See also: the miracle of 6.5, 7.5, and 8.5 multimeters with single or even 2 digit ppM measurement certainty (and, of bloody course, hardly any of them actually is beyond x.3 digits. In fact, some of the "very best of the best" actually are very close to x.0 digits ...).
I'll end with (surprise!) kudos to R&S because AFAIK they stopped at 5 3/4 digits. I may be wrong but my guess is that some of their engineers back then said "Nope. We don't take part in the BS-your-customers game!".
Btw. does R&S also sell the MXO5 in six-packs? I'm asking because buying a scope for 60 k€ + VAT isn't worth to leave the house *g
VxWorks - Yes! Linux - meh. Windows - Thanks no, definitely.
Offline shabaz
« Reply #133 on: May 16, 2024, 03:18:03 am »
...And I'm amazed yet again how happily (almost) everyone accepts the usual BS. Example: If an ADC does 12 bits that doesn't mean that you actually get 12 bits...
What makes you think people don't know that? No-one normal cares about a number '12', they care about actually using 'scopes in real-life scenarios, to see the benefit (or lack of) for real applications. There's a whole thread concerning that with a different (lower-cost) 12-bit 'scope.
Those who want to wait around for 12 to mean what you want it to mean, can knock themselves out, meanwhile, the normal world will still try out the instruments, and if there is no improvement, or if performance is worse, or is sluggish in response, then people won't buy it - that's what forums are good at, sharing the feedback.
You actually have zero useful feedback in that long comment; everyone knows the 12 doesn't indicate ENOB. You also didn't get the architecture right; it's not just a faster processor; it's doubled-up silicon, there are two multi-core processors.
The following users thanked this post: nctnico, egonotto, jjoonathan, jusaca, pdenisowski, KungFuJosh
Offline jjoonathan
« Reply #134 on: May 17, 2024, 02:54:30 am »
The real scandal is how many scopes they keep releasing that aren't the MXO6, which is the one I want. At this rate, they are going to release the MXO7 and MXO8, then loop back to do a prequel trilogy with MXO1, 2, and 3, and then they will start into the low profile version of each of those, and then an 8 channel version of each of those, and then a dedicated logic analyzer RF hybrid thing, and by now it's 2050 and we all have flying cars, and the thread is called "Which entry level 16 bit oscilloscope should I buy," and then the MXO6 comes out.
The following users thanked this post: egonotto, tv84, pdenisowski
Offline pdenisowski
« Reply #135 on: May 17, 2024, 09:53:33 am »
(laughs)  Point taken. Our current "6" series scope is the RTO6, which is no slouch:
https://www.rohde-schwarz.com/us/products/test-and-measurement/oscilloscopes/rs-rto6-oscilloscope_63493-1079745.html
I hope you don't mind me using the "prequel trilogy" comment internally - most of our scope team would absolutely love that reference :)
The following users thanked this post: egonotto, tv84, jjoonathan
Offline jjoonathan
« Reply #136 on: May 17, 2024, 02:25:02 pm »
Ha, go for it! The RTO6 is a beauty, but I believe it uses the "Ooma" ADCs. These are some of the best 8b converters to ever be put in a scope -- spurious is excellent, they can beat certain 10b converters at high frequency ENOB, and the R&S "HD Mode" does an atypically good job of trading speed for bit depth -- but I already have a RTO scope on my desk, so if I am spending this much on an upgrade I am looking for a new analog front end as well. That's why I am eagerly awaiting the MXO6. Of course, at this price tier my timing and choices are partly driven by external requirements, so who knows how it will actually come down, but I can at least lust over the brochures  ;D
Offline ddrl46
« Reply #137 on: May 27, 2024, 06:48:06 pm »
While reading over the specifications of the scopes in the MXO series I noticed that the trigger sensitivity specification seems to be rather optimistic in comparison with other oscilloscopes. Or has Rohde & Schwarz perhaps changed their interpretation of trigger sensitivity since the RTB / RTA series?
Offline nctnico
« Reply #138 on: May 27, 2024, 07:06:50 pm »
I'd say it is very overly optimistic because there will be more noise than 0.0001 division for sure. I think somebody took the minimum hysteresis setting and forgot about noise.
The following users thanked this post: ddrl46, egonotto
Online nidlaX
« Reply #139 on: May 30, 2024, 10:19:49 pm »
I just got the marketing email today. Can you comment publicly on pricing? This looks like the best headless scope on the market at the moment!
The following users thanked this post: pdenisowski
Offline pdenisowski
« Reply #140 on: May 31, 2024, 09:33:51 am »
Base model pricing is available on our website, and you can also "build your MXO5C" and then click a button to get a quote
https://www.rohde-schwarz.com/us/products/test-and-measurement/oscilloscopes/rs-mxo-5c-oscilloscope_334265.html
Offline egonotto
« Reply #141 on: August 24, 2024, 07:56:53 am »
Hello,
Shahriar takes a look at the Rohde & Schwarz MXO-5 Series, their first 8-Channel oscilloscope.
Best regards
egonotto
The following users thanked this post: DaneLaw, pdenisowski, Martin72
Offline tooki
« Reply #142 on: August 24, 2024, 12:15:53 pm »
FYI, a key tidbit from that video, in case you decide not to watch it: they finally released the Zone Trigger function in the latest firmware update (for both MXO5 and MXO4)!
Edit: from watching the rest of the video, it looks like the Zone Trigger is really powerful, more so than the one in e.g. the Keysight 3000X/4000X series. (And that was already a useful feature!)
« Last Edit: August 24, 2024, 12:44:29 pm by tooki »
The following users thanked this post: egonotto, 2N3055, pdenisowski, Martin72
Offline 2N3055
  • Country: hr
« Reply #143 on: August 24, 2024, 01:59:37 pm »
From practical standpoint, biggest advantage is number of zones (most scopes have 2). Non rectangular zones are edge case, choosing logical operation is useful sometimes but you must pay attention what you want to accomplish. All in all, nice development. It took almost 2 years though... But at least now looks good.
Dr. Richard W. Hamming
Offline tooki
« Reply #144 on: August 25, 2024, 12:03:58 pm »
Non-rectangular is exotic, sure. But the thing that is really interesting is that you can do a zone trigger on the spectrum analyzer. I don’t think the Keysight 3000X/4000X can do that. That opens up some really interesting frequency-domain triggers.
Offline 2N3055
  • Country: hr
« Reply #145 on: August 25, 2024, 01:10:13 pm »
That is interesting. That is called Frequency Mask Triggering in SA parlance.
And no M.Z. IV scopes can do, in fact not many at all can do that. That is full blown mixed domain function.
What would be interesting to see is what is POI (probability of intercept) for that function..
The following users thanked this post: egonotto, tooki
Offline tooki
« Reply #146 on: August 25, 2024, 01:17:33 pm »
I haven’t a clue. (I have zero experience with “real” SA’s, I’ve only ever used the FFT or SA feature in oscilloscopes.) If there’s a simple way to test it, lemme know how and I’ll try it out on the MXO4 at work.
I suppose I’m not surprised that it’s R&S that added this to a DSO, given their STRONG background in RF test gear like SAs.
Offline jjoonathan
« Reply #147 on: August 25, 2024, 03:04:38 pm »
The ability to do a "zone trigger" on a FFT goes way back on the RTO/RTE instruments: you could add frequency zones to the mask test and set the action to stop/record. It would only apply to waveforms that the regular trigger had already decided to capture so the POI wasn't in the same league as an RTSA, but at the same time the R&S scopes can do so many hardware-accelerated waveform captures and FFTs that in practice it usually worked so long as you weren't hunting the rarest of events. The MXO4/5 are multiples stronger on the waveform/FFT front so I expect them to be even stronger in this regard, but still not in a position to offer RTSA-competitive POIs. Again, in practice this is often not an issue -- as you saw in Shahriar's video, you can pretty aggressively filter with the zones and still get decent update rates if the event is decently repetitive
https://youtu.be/EHEXDHxzSB0?t=2278
I've been lobbying for true RTSA zone triggers for years, because as we know from Deep Voice Guy, R&S know how to do them (1) but it turns out there might have been a reason for R&S's reticence to port over the RTSA features to the scopes due to ongoing patent litigation. Apparently Tektronix bought another patent on recording I/Q samples on an oscilloscope  :-- :-- :--
https://patents.google.com/patent/US8675719
https://patents.google.com/patent/US8521460
R&S challenged this with some success
https://developer.uspto.gov/ptab-web/#/search/documents?proceedingNumber=IPR2018-00643
but the ruling left some of the claims standing because Tek was able to make the argument fly that this was totally different from their previous patent on account of having an ever so slightly different signal path. This probably leaves open the route to RTSA functionality so long as it shares the same signal path as time-domain waveforms, but A) I'm not sure if R&S wants to touch this, B) I am not sure the details of their private agreement with Tek, and C) even if they decided to go ahead with it they probably wouldn't have started until the litigation settled in 2020. Best case is that we see RTSA-like POI in the next few scopes. Worst case is they wait for the patents to expire and we don't see it for a decade, and while the POI isn't as good as it could have been it is good enough for many purposes. These R&S scopes are fast.
(1)
« Last Edit: August 25, 2024, 05:52:41 pm by jjoonathan »
The following users thanked this post: egonotto, tooki, 2N3055, pdenisowski
Offline jjoonathan
« Reply #148 on: August 25, 2024, 05:50:02 pm »
Hey, it looks like the Tektronix patent on a Modulation Domain Trigger expires next year  :-+ :
https://patents.google.com/patent/EP1669764A1/en
The following users thanked this post: tooki, 2N3055, KungFuJosh
Offline mkissin
« Reply #149 on: October 01, 2024, 01:05:17 am »
Interestingly, there's a bunch of new bundle offers for both the MXO4 and now also the MXO5 scopes, valid until March 31 2025.
https://www.rohde-schwarz.com/us/products/test-and-measurement/oscilloscopes/promotion/oscilloscopes-promotion_253927.html
The MXO5 ones aren't on the website for some reason, but they are in the PDF flyer.