Author Topic: RTA4004 Warranty Issue or Firmware Bug?  (Read 758 times)
Offline OdysseusTopic starter
« on: March 03, 2024, 02:25:23 pm »
I'm looking for anyone with an RTA4000 series R&S scope that could check if this problem I've encountered is isolated to my instrument. The summary is that channels 1 and 2 will sometimes appear on the display as if they were un-triggered, with random time alignment, even though channels 3 & 4 are stable and channel 1 (or 2) is the trigger source. The attached screenshot does a good job exemplifying the issue. Just a single 10MHz square connected to all four inputs.
It only happens when the FFT function is enabled with high resolution acquisition enabled. I've also exported and attached the setup file to make it easier to replicate.
« Last Edit: March 03, 2024, 02:30:51 pm by Odysseus »
Online egonotto
Re: RTA4004 Warranty Issue or Firmware Bug?
« Reply #1 on: March 03, 2024, 04:17:45 pm »
Hello,
i can not load you setupfile.
Best regards
egonotto
Online 2N3055
  • Country: hr
« Reply #2 on: March 03, 2024, 04:55:55 pm »
Make sure you are running latest FW version in a first place.
Also if  it is a 10 k€ scope in warranty, first step is to contact point of purchase and R&S support.
The following users thanked this post: pdenisowski
« Reply #3 on: March 04, 2024, 12:59:28 am »
Hello,
Best regards
egonotto
Thank you for trying.
I am able to load it back onto my scope, but it does seem like I uploaded the wrong version.
I've attached another, which I've confirmed loads correctly after a Preset and recreates the issue.
Note that I've uploaded a .zip file, which needs to be extracted first.
Yes, the very first thing I checked, the firmware is the most recent for the RTA4000 series, 01.700.
I did eventually figure out how to submit a ticket to R&S, but it was more convoluted than I expected. I'm not confident I'll even get a response.
I would also be frustrated if I sent it in for warranty and I got it back with the same problem, because it's a firmware issue. Hence my post.
« Last Edit: March 04, 2024, 04:08:45 am by Odysseus »
Online egonotto
« Reply #4 on: March 04, 2024, 02:46:23 am »
Hello,
i can again not load you setupfile.
I have Firmware Version:    01.700 but not US
I only changed the voltage to 50 mV because I use a BNC connection.
If I activate FFT, I can no longer change the deflection time. It is now 34.95 us/div.
You can explain to me how I can make the setting by hand, or you can contact pdenisowski from R&S who recently wrote here: https://www.eevblog.com/forum/testgear/scope-recomendations-in-$15-20k-range/
Enclosed is my setup file.
Best regards
egonotto
« Reply #5 on: March 04, 2024, 04:06:17 am »
Your setup file will load on my scope but throws a similar server error. Oh well.
The minimum steps to recreate is:
  • Preset
  • Set time scale to 80ns/div
  • Set Acquisition to hi-res
  • Turn on channel 3
  • Apply identical signal to channels 1 and 3
  • Enable FFT
  • Wait 5 to 30 seconds
Thanks for taking the time to help, and thanks for the contact. I'll see if my ticket gets a response, and ping pdenisowski otherwise.
Online pdenisowski
  • Product Management Engineer, Rohde & Schwarz
    • Test and Measurement Fundamentals Playlist on the R&S YouTube channel
« Reply #6 on: March 04, 2024, 01:52:01 pm »
Please do :)
Test and Measurement Fundamentals video series on the Rohde & Schwarz YouTube channel:  https://www.youtube.com/playlist?list=PLKxVoO5jUTlvsVtDcqrVn0ybqBVlLj2z8
Free online test and measurement fundamentals courses from Rohde & Schwarz:  https://tinyurl.com/mv7a4vb6
The following users thanked this post: egonotto, Odysseus
Online egonotto
« Reply #7 on: March 04, 2024, 01:59:24 pm »
Hello,
I am also experiencing this problem. After a few seconds, the triggered channel runs and the untriggered channel remains stable.
It even gets worse later.
Best regards
egonotto
« Last Edit: March 04, 2024, 02:17:51 pm by egonotto »
The following users thanked this post: Odysseus
Offline nctnico
« Reply #8 on: March 04, 2024, 03:54:25 pm »
As I expected, I can reproduce the effect on my RTM3004 as well. Let's classify this as a firmware bug. Cycling through trigger or FFT settings makes the signal stable for a while but then it goes off again.
« Last Edit: March 04, 2024, 03:59:30 pm by nctnico »
« Reply #9 on: March 05, 2024, 10:07:58 pm »
R&S support has also been able to replicate the issue and filed a bug report, but advised me there is no guarantee this will receive a firmware update, as engineering is currently focused on the MXO series.
It appears that using only channels 1 & 2 or only channels 3 & 4 at a time will avoid triggering the bug.
Unfortunately, that puts the scope into 2.5GS/s instead of 5GS/s, so aliasing becomes an issue at higher frequencies.
Thanks again for your help.
« Last Edit: March 05, 2024, 10:16:17 pm by Odysseus »
The following users thanked this post: nctnico, egonotto
Offline nctnico
« Reply #10 on: March 05, 2024, 10:11:27 pm »
That would be very dissapointing...
« Last Edit: March 05, 2024, 10:48:22 pm by nctnico »
