While commenting on another thread about DSO probe effects at the DUT thought it might be useful to start a new thread dedicated to such.
https://www.eevblog.com/forum/testgear/siglent-sds800x-hd-12-bit-dsos-coming/msg5759401/#msg5759401This is about the effects of the DSO probe as "Seen" by the DUT, rather than the usual probes effects as displayed on the DSO.
The setup is crude (not technically correct and calibrated) but good enough to show the effects, see thread above for details. An AWG signal generator (SDG6022X) is set to 0dBm and swept from 1MHz to 300MHz. The test DSO is the SDS824X HD (enhanced to 200MHz) in this case and thus the selected 1MHz to 300MHz Frequency Range. For the AWG 50Ω termination we utilized the input of a Spectrum Analyzer (SSA 3121X +) so we could "see" the DSO Probe effects on the SA since it's now the DUT for the DSO Probe. A BNC "T" was used to allow the DSO Probe BNC Probe Adapter to be inserted and the various test probes inserted into the adapter, see mentioned thread for setup images.
The SA was set to 1MHz to 300MHz range, peak detect, and +1dBm Reference level with 1/2dB/div. A reference was captured to show the results without the DSO probe and should show ~0dBm across the 300MHz span as the Yellow Trace. This shows about ~1/2dB variation due to the AWG, cables, adapters and such, and was not calibrated out. A more diligent test would calibrate this out and utilize better cables, adapters and such, however we felt this unnecessary for this case.
Anyway, the Yellow Trace is the reference without any DSO Probe attached. The Magenta trace is with the Siglent PP510 Probe @ 10X, the Cyan trace is the Siglent 10X SP2030A and the Green is the Rigol PVP3150 @ 10X. All probes were attached to a Siglent SDS824X HD, altho we found the SA traces didn't change when the Probe was attached or not on the DSO side. The PNG116 trace is from the Siglent SDS824X HD FFT with the PP510 Probe @ 10X, compare with the SA Magenta trace, note scale differences tho.
Note that the SA is showing a result which indicates the probe represents a more complex "load" to the DUT (SA in this case) than a simple shunt capacitance. This may cause the DUT to behave differently than expected and one should also realize that the result displayed on the DSO screen is not exactly what the DUT is experiencing due to the probe loading. In addition, the frequency response behavior can't be just related to a clock rate at the DUT for example, since the clock edges contain significant energy well beyond the clock rate and this must be considered

Happy Holidays,
Best