I am trying to find out if and how far SDM3045X can convert an input like 4-20mA (pressure transmitter) to display pressure values.
So far I can read from the DM3058E manual this can be done by the function "Any sensor Measurement"
The 4-20mA sensor example is very common but conversion of other types could be nice.

I dont own any of these 2 instruments but it could be the reason for chosing either of them.

Knud
 

Online knudchTopic starter

  • Regular Contributor
  • *
  • Posts: 64
  • Country: dk
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #1 on: October 19, 2021, 08:16:22 am »
I can see that recent firmware release for SDM3045X (date 24 sept 2021) says added "customer sensor support"
But no documentation how ?
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #2 on: October 19, 2021, 09:43:56 am »
I can see that recent firmware release for SDM3045X (date 24 sept 2021) says added "customer sensor support"
But no documentation how ?
Yes the documentation is poor.  :palm:
Just got a 3065X out to examine use of custom sensors which is in the Temp sensor menu and it seems simple enough to use were you can load sensor definitions and even make custom adjustments to tune accuracy.

Have to dig deeper and get a 3045X out to see if it has the same functionality and if so will put up some pics.
Later I'll ask Siglent for updated manuals that include instructions for custom sensor use.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online knudchTopic starter

  • Regular Contributor
  • *
  • Posts: 64
  • Country: dk
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #3 on: October 19, 2021, 10:23:08 am »
If the SDM3045X have custom sensor function like the Rigol I would prefer the siglent (LAN included, better display functions).

Important for 4-20mA function is that you can insert at least 2 conversion points and then it will interpolate between these.

Difference in accuracy is not so big as long we are not going to very low DC values.

When I not can rely on documentation (manuals) are there any functional differences between SDM3045x and 3055 ?
 

Offline voltsandjolts

  • Supporter
  • ****
  • Posts: 2300
  • Country: gb
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #4 on: October 19, 2021, 10:42:29 am »
Important for 4-20mA function is that you can insert at least 2 conversion points and then it will interpolate between these.

I'm not aware of a bench meter that does that.
More common is to enter scale and offset (mx+b).
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #5 on: October 19, 2021, 11:07:13 am »
SDM3045X Custom sensor menus = SDM3065X and presumably SDM3055 is identical too. (no stock to check)

Custom sensors are accessed via the Temp menu where you can load a definition, Modify it and set sensor values.
See screenshots.

When I not can rely on documentation (manuals) are there any functional differences between SDM3045x and 3055 ?
Only the ranges of measurement as outlined in datasheets however the 3045X has the advantage of being silent.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online knudchTopic starter

  • Regular Contributor
  • *
  • Posts: 64
  • Country: dk
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #6 on: October 19, 2021, 11:18:17 am »
Important for 4-20mA function is that you can insert at least 2 conversion points and then it will interpolate between these.

I'm not aware of a bench meter that does that.
More common is to enter scale and offset (mx+b).

Look in user manual for DM3058 page 2-29 onwards
 

Online knudchTopic starter

  • Regular Contributor
  • *
  • Posts: 64
  • Country: dk
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #7 on: October 19, 2021, 11:22:57 am »
SDM3045X Custom sensor menus = SDM3065X and presumably SDM3055 is identical too. (no stock to check)

Custom sensors are accessed via the Temp menu where you can load a definition, Modify it and set sensor values.
See screenshots.

When I not can rely on documentation (manuals) are there any functional differences between SDM3045x and 3055 ?
Only the ranges of measurement as outlined in datasheets however the 3045X has the advantage of being silent.

thanks for your effort

As I don't has access to a SDM3045X (or is familiar with it) it is not quite clear for me how it works
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #8 on: October 19, 2021, 11:36:31 am »
SDM3045X Custom sensor menus = SDM3065X and presumably SDM3055 is identical too. (no stock to check)

Custom sensors are accessed via the Temp menu where you can load a definition, Modify it and set sensor values.
See screenshots.

When I not can rely on documentation (manuals) are there any functional differences between SDM3045x and 3055 ?
Only the ranges of measurement as outlined in datasheets however the 3045X has the advantage of being silent.

thanks for your effort

As I don't has access to a SDM3045X (or is familiar with it) it is not quite clear for me how it works
3045X works just like SDM3055 of which there are many old videos on the internet however there have been a few firmware updates since they were released. The measurement ranges are different for 3045X and only SDM3065X has PLC settings otherwise the 3 models are very similar to use and feature sets are nearly identical except for accuracy.
Each model uses a different reference for the levels of accuracy they provide.
Thread for the 3 models here:
https://www.eevblog.com/forum/testgear/siglent-new-bench-dmm-sdm3055/
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online knudchTopic starter

  • Regular Contributor
  • *
  • Posts: 64
  • Country: dk
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #9 on: October 19, 2021, 11:45:50 am »
In the firmware update for SDM3055 there are not mentioned anything about custom sensor
But 3055 firmware are some months older than 3045X
 

Online knudchTopic starter

  • Regular Contributor
  • *
  • Posts: 64
  • Country: dk
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #10 on: October 19, 2021, 12:33:31 pm »
In the SDM3045X is custom sensor implemented like offset + scale ?

Not quite clear to me
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #11 on: October 20, 2021, 08:48:44 am »
In the SDM3045X is custom sensor implemented like offset + scale ?

Not quite clear to me
Me neither however I think so.
It appears a file of the sensors characteristics must first be loaded but it what format and what it needs to contain IDK.  :-//

Sent email to Siglent asking for instruction of how Custom sensors are to be setup/used. No word back yet however also asked them to prepare an App note for Custom sensors.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online knudchTopic starter

  • Regular Contributor
  • *
  • Posts: 64
  • Country: dk
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #12 on: October 20, 2021, 09:10:47 am »
In the SDM3045X is custom sensor implemented like offset + scale ?

Not quite clear to me
Me neither however I think so.
It appears a file of the sensors characteristics must first be loaded but it what format and what it needs to contain IDK.  :-//

Sent email to Siglent asking for instruction of how Custom sensors are to be setup/used. No word back yet however also asked them to prepare an App note for Custom sensors.

Thanks for working on this
It seems that I must wait a little before making a decision
 

Online knudchTopic starter

  • Regular Contributor
  • *
  • Posts: 64
  • Country: dk
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #13 on: October 28, 2021, 08:46:05 pm »
Got a preliminary information from SiglentEU
(They are working on updated of the manual)

Custom sesnor is implemented as a lookup table
Input value => displayed value

That means you can use it also for nonlinear sensors

Just ordered a SDM 3065X :)
 
The following users thanked this post: tautech

Online knudchTopic starter

  • Regular Contributor
  • *
  • Posts: 64
  • Country: dk
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #14 on: November 05, 2021, 12:12:01 pm »
Just received and unpacked my SDM 3065X  :)

Tested right away the custom sensor function, works perfect
example: 4-20mA sensor input
table entry 4mA = 0 Bar presure
table entry 20mA = 10 Bar presure
save under a filename
Activate under temp measurement
12mA measured => display 5 bar
You can display both values at the same time....useful

Cal date was in may 2021...is ok
firmware  V3.01.01.08R1

Just let it run for a few days before update to the latest FW
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #15 on: November 05, 2021, 12:22:23 pm »
Nice, glad it worked out ok.  :-+
But do update the FW now as #3 in the release notes is an important fix.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline kcbrown

  • Frequent Contributor
  • **
  • Posts: 880
  • Country: us
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #16 on: November 05, 2021, 01:24:22 pm »
In the firmware update for SDM3055 there are not mentioned anything about custom sensor
But 3055 firmware are some months older than 3045X

The SDM3055 doesn't have the custom sensor function.  I've got the latest firmware installed on mine and the "Custom Sensor" menu item is conspicuously absent from it.
 

Online knudchTopic starter

  • Regular Contributor
  • *
  • Posts: 64
  • Country: dk
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #17 on: November 05, 2021, 01:42:26 pm »
In the latest FW (24 sept 2021) for 3045x was it introduced

None of the FW release notes for 3065X mention it, so no idea when it was introduced.

My gues is that it will be introduced in 3055 at the next FW update.

No logical reason why it not should come in 3055
 

Offline jan28

  • Contributor
  • Posts: 38
  • Country: nl
Re: SDM3045X versus DM3058E...sensor input 4-20mA
« Reply #18 on: November 07, 2021, 12:51:24 pm »
For those interested how it looks on a 3045X (with recent firmware you can now send screenshot directly over SCPI to a host...):

In this case I measure a voltage that represent a current, not very fancy but that's what is hooked up right now. I really like it also works in trend display.
At first sight it looks pretty versatile and handy. So many time I'm looking at my DMM value and calculating to what it translate I'm measuring, this really helps.



 