Author Topic: SDM3055 reading speed and storing  (Read 1492 times)
Offline chrisb741Topic starter
« on: December 13, 2023, 12:02:12 am »
The SDM3055 has a reported reading speed of 150/sec
When I set it to store data I only get about 60 recorded entries per second
Is it possible to store the 150 readings per second?
« Last Edit: December 13, 2023, 12:26:16 am by chrisb741 »
Online Kean
    • Kean Electronics
Re: SDM3055 reading speed and storing
« Reply #1 on: December 13, 2023, 12:56:05 am »
Did you select the Fast measurement mode?
From the manual:
The instrument provides three types of measurement rate: 5 reading/s, 50 reading/s and 150 reading/s.
5 reading/s belongs to “Slow” rate; 50 reading/s belongs to “Middle” rate; 150 reading/s belongs to
“Fast” rate.
Measurement speed can be controlled by softkey menu. Press [Speed] and then press [Slow], [Middle]
or [Fast] to choose measurement speed.
« Reply #2 on: December 13, 2023, 01:10:31 am »
yep rate set to fast i get about 60 readings, slow is about 5
Could it be another option?
I'll check again
Online Kean
    • Kean Electronics
« Reply #3 on: December 13, 2023, 01:20:45 am »
What measurement type?  Some don't support Fast readings, and some only Slow.
I don't have one of these, but I almost bought one so I have read the manual previously.
« Reply #4 on: December 13, 2023, 01:36:40 am »
I was measuring DCV
Filtering is off
Online Kean
    • Kean Electronics
« Reply #5 on: December 13, 2023, 07:27:41 am »
OK, I believe DCV should support "up to 160rdgs/s".  As I don't have one here, I cannot help further.
Maybe @tautech has some experience, or can follow up for you.
Online Tjuurko
« Reply #6 on: December 13, 2023, 08:30:24 am »
You can try to write down on an external disk. For SDM3065X (for SDM3055 it should be similarly), you need to do:
Acquire
Save Reading
U-disk Record
.../external
Record -> On
Done
Record -> Off
I received data 1000 and 10000 Readings/s.
« Last Edit: December 13, 2023, 08:33:43 am by Tjuurko »
« Reply #7 on: December 13, 2023, 08:55:57 am »
Yep that's what I was doing
Then counting t how many rows each second has
« Last Edit: December 13, 2023, 08:59:51 am by chrisb741 »