Products > Test Equipment
Searching for some precision mains power meter (0.1mA)
Swainster:
I have a RS branded GWInstek bench power meter with 100nA resolution on the 5mA range. Some numbers just to get a feel for things:
With mains 115VAC/60Hz the current is about 0.06mA even with nothing plugged into the outlet. With a long IEC lead (still no load) then the current goes up by 0.012mA. With mains set to 240VAC/50Hz then this leakage current is about 0.1mA. Plugging in a few random bits of old test equipment (power supplies, electronic load) with hard on/off switches in the off position typically adds another 0.1mA (except for one which added 14mA... this I know has crispy RIFA caps inside so I guess it is trying to telling me something). A random cheap chinese phone charger uses about 2mA without any load, though to be fair this was dropping as the internal caps reformed.
I bought this power meter to measure standby/idle currents according to IEC62301, which is actually something typical power meters cannot do. One of the requirements is 1mW resolution - thats under 9uA at 115VAC. My suggestion is that if you are serious about this type of measurement then you will need a proper bench type instrument. If you are only interested in current then a DMM could be good enough, but if you actually want to know power/energy used then a power meter is the most practical solution (2 DMMs will not work without a lot of extra work to extract and sync the data, even assuming they have enough bandwidth).
TLDR What you want to do is non-trivial and probably the preserve of application specific bench instruments. Building a custom instrument to do it is always possible, but only practical for hobby/enthusiasts - In fact I think this kind of project would generate quite some interest on this forum. For commercial purposes it makes far more sense to just buy the proper tool for the job.
Navigation
[0] Message Index
[*] Previous page
Go to full version