I bought a Sencore lc102 meter at a hamfest for an insane good price. I calibrated it and it worked flawlessly for month.
Now, it says "error 7" when one attempts to zero it, or measure capacitance below 2uF (but not above). Inductance and ESR measurements are fine. The onset of the problem was sudden as if something folded. I have been working on this thing for about a month now, trying to diagnose the problem and it is absolutely kicking my ass. I have replaced the relays, the lm319 comparator, and for awhile suspected IC12, a flip-flop that addresses the current sources for capacitance measurements since only measurements below 2uf were affected (therefore, I suspect, screwing the ability to zero compensate). One theory was a power supply issue, but all the rails are nominal, except for the 12 volt unregulated rail which is at 13.8 volts, but I think that is normal. My most recent theory has been that there is a problem with one line in the control buss, but swapping the IC and checking all the components on that buss (F5) yielded nothing.
In fact, though, I am still guessing. I still don't even know which section of the circuit is problematic.
If someone who has some familiarity with this instrument could throw me a bone, it would be very grateful. I am running out of ideas.
This is such a nice instrument. I refuse to give up.