| Products > Test Equipment |
| Sensitive LCR Meter Measurements |
| << < (13/16) > >> |
| Martin72:
The two fixtures photographed from the outside... |
| Martin72:
Finally, the clearly different inner life of the two. |
| mawyatt:
How is the bracket attached to the cover in the Left side of image 4788 and Right side of image 4789? There's no thru bolts or screws shown from the top side of the cover image for this bracket attachment!! Are these something like a welded stud to the cover bottom, then smashed/expanded to hold the bracket?? Best, |
| Martin72:
--- Quote ---How is the bracket attached to the cover in the Left side of image 4788 and Right side of image 4789? --- End quote --- Normally with screws through the housing, as these also fix the spring mechanism on top of the cover. This is not necessary on the other side, where threaded bolts are welded on and the bracket is fixed with nuts. Mechanically completely plausible. |
| mawyatt:
--- Quote from: Martin72 on August 23, 2024, 07:53:21 pm ---I have now also taken the ST fixture home with me, before that I measured 3 capacitors at 100khz at work, once with the ST, then with my nameless one. The results are congruent, which suggests that the “denser” structure of the ST does not yet play a role in the frequency. Tomorrow I will do the same with the ET3503, but then of course also at 300kHz. When I got home, I photographed both fixtures again, the pictures will be split accordingly due to the maximum number. --- End quote --- With the 10pF Caps this is showing ~20ff difference, or 0.2%. Since you used the same 10pF Cap (C0G), same quality LCR Meter, same basic setup, this leaves this Fixture as the main source of the difference. The difference effects can now be broken down into variations in Cal and main differences in physical fixture (lack of shield). A simple experiment can verify IF the shield is worthwhile. Setup the measurement after lengthly warmup with a very large average (~100), do a spot Open Cal at highest frequency the LCR supports (300KHz) with a small piece of plastic about same size as 10pF Cap DUT for proper plunger spacing with the ST Fixture. Now record the result, then replace the Fixture with the No Name and repeat, record the result. Note the small plastic spacer will allow the 2 fixtures to have ~ the same spacing and remove this uncertainty, however they will increase the fringe capacitive effect, so don't rely on this for accurate measurements, only relative differences. The ~100 averages should give good mean values for comparisons. Now carefully remove the No Name cover and install a temporary shield. Use some art & crafts thin Cu plate and cut a shield and tack solder in place (BNC lug) at the ST Fixture shield location, careful not to disturb the wires. Make sure it's mechanically stable, you can use some hot glue if needed. Put the cover back on and reinstall the fixture, then repeat the measurement procedure. Suspect you'll see a value closer to the ST fixture measurement, and this should verify the benefit of the shield. After all, don't think Tonghui (and others) would use a shield if it provided no measurable benefit ;) Best, |
| Navigation |
| Message Index |
| Next page |
| Previous page |