explain further.....you need to drive a purely capacitive load?
This is not possible, as all device inline have an ESR.
If you are trying to characterize inrush slopes, against a turn on event....then you are making things a bit over complicated.
How fast of a rise time do you need to measure? There are some very easy ways to do this, and you don't need anything all that complex.
What is the device under test? Also what is the most common load presented to the device?
There is some info lacking here....
I know I can fully characterize inrush events with my Maynuo DC load, and the PMA software in my scope. Although even that is a bit luxurious.
Are you looking to simply characterize inrush ? or do you need to see a capacitive discharge curve, under load?
Part of the problem here is that the device/load is going to have reactive properties of it's own, that need to be characterized and subtracted from the final outcome.....
every load I have tested has at least a minimal input capacitance of it's own.....as do the test leads.....
also what kind of bandwidth attenuation (freq vs. impedance) are you expecting to measure? A 1000uF cap is going to be quite on the low side of plane impedance vs. frequency (step response), but you might be missing HF events that are sneaking through, as large inrush spikes.....those events can also couple back into the control loop and cause an oscillation. Part of the problem with an active (external) load, is that it's going to ultimately inject some feedback into the test.
Also what kind of power are we talking about here? 1 volt into a 0.01 ohm resistor is easy.....12 volts into the same is not....
There is also going to be HF ringing, that is in harmonic intervals to the primary/fundamental transient.....those artifacts are going to be delayed, in comparison to the fundamental....and they will be orders of magnitude above/below the switching frequency.....even if the PSU is linear, the turn on event will still exhibit these properties.
DC is a myth

50KHz is the sweep limit on those Chroma units.....yokogawa also makes some gear for characterizing such things....but again it's mostly LF limited. Most of this stuff is designed for testing higher voltages....where currents are far lower......but dealing with 12v 3v 5v rails (I have to assume you make ATX style PSU's?) can call for currents that most of these products can't deal with. I picked the specific Maynuo I acquired, because it was about the only DC load out there that would handle 60 amps, and still maintain usability, bench top form factor, sane pricing etc....
A 1000uF is going to have a planar impedance vs frequency of roughly 0.010 - 1 ohm at 0.01 HZ- 100 MHz.....I think you can see where a 50KHz sweep might limit fully characterizing impedance curves
The biggest problem I see here is that these loads are SIMULATING ideal and narrow tolerance ranges. The best option to actually test yield (production units) is to physically characterize the real device. For example I deal with some loads that are dynamic, varying with temperature (Ti alloy resistance wire) and the Maynuo load can get me part of the way there, but can't allow for tolerance in the production resistor. The way I do this is by sweeping the load in the Maynuo, and then trying to subtract the properties added by the load itself.
It's critical to understand the full scope of what loads might be presented to the PSU....both in a static and dynamic sense. I have to assume these are all fast step responses.....and we are talking about cyclical transients below 2ms?